Rob wrote:
50V across 5GOhm is 10nA. Put a standard multimeter's 10MOhm input
in series and
you have 10mV per nA reading. Anything below 100mV is pass.
Also, quite a few of the Fluke portable and handheld DMMs from the
last 35 years or so (including the faithful old 8050A and the
"80-series" DMMs) measure conductance (1/R) with a resolution down to
0.01nS (= 100G ohm). There are lots of them with this capability on
the used market for $10 and up, and several are still available new
(but not for less than Eu100, to my knowledge).
Best regards,
Charles
_______________________________________________
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.