A question to ask Keysight, if the instrument arrives without cal constants, will they consider that a "repair", or a calibration? Very different prices. I suspect that Keysight, and any other cal lab that has MetCal, will hook it up to a multifunction calibrator, computer and let it run.

If you jumper another battery with current limiting across the existing during replacement to preserve the cal, you could then ask for before and after data from the cal lab. This is typically an extra cost item.

I work on DMM's frequently, if this were in my shop, I would first make sure that the instrument is stable, regardless of the absolute accuracy, and that the relays have not become noisy. This can be checked by rapping them lightly with something like a screwdriver handle . Also a good opportunity to check and replace any electrolytic caps that have poor ESR.

I was just working on a pair of Data Precision meters, and both had noisy, unstable relays. This is a frequent problem, even with the "sealed" relays such as found in the HP. I would also make sure that any Front/Rear input switching is noise free, as these are often unused for years, if ever, and can cause problems.

Bob

On 1/31/2016 9:00 AM, [email protected] wrote:
Re: 3457A - should I preserve SRAM contents, or purposely
       corrupt them by removing power?
       (Dr. David Kirkby (Kirkby Microwave Ltd))

_______________________________________________
volt-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.