Yes, there is: The monitor mode command "I" (upper case i) will write a test pattern to the NVM and checksum it.
************** Any calibration constants will be lost ********************** As you don't have the manuals handy here's the long form: Use your favourite software terminal emulator (e.g. Putty) to connect to the 7081 using the serial port. (straight through serial cable). The baud rate is determined by dip-switch S1 on the earthy logic board. If all serial port rate switches are to the right, it's set to 9600 baud. Issue the following commands: out,rs232,on monitor I Ctrl-X Dave -----Original Message----- From: volt-nuts [mailto:[email protected]] On Behalf Of Four Designs Company Sent: 22 April 2016 17:20 To: [email protected] Subject: Re: [volt-nuts] volt-nuts Digest, Vol 80, Issue 14; Solratron cal issue It has been some years since I last worked on a 7081, and I don't have the manuals at hand.. Is there a way to clear the EAROM contents and perhaps load starting cal constants? I think the EAROM data is corrupt or so far off that it cannot compute new values. Bob _______________________________________________ volt-nuts mailing list -- [email protected] To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.