Yes, there is:

The monitor mode command "I" (upper case i) will write a test pattern to the 
NVM and checksum it. 

************** Any calibration constants will be lost **********************

As you don't have the manuals handy here's the long form:

Use your favourite software terminal emulator (e.g. Putty) to connect to the 
7081 using the serial port. (straight through serial cable).
The baud rate is determined by dip-switch S1 on the earthy logic board.  If all 
serial port rate switches are to the right, it's set to 9600 baud.

Issue the following commands: 
out,rs232,on
monitor
I
Ctrl-X

Dave

-----Original Message-----
From: volt-nuts [mailto:[email protected]] On Behalf Of Four Designs 
Company
Sent: 22 April 2016 17:20
To: [email protected]
Subject: Re: [volt-nuts] volt-nuts Digest, Vol 80, Issue 14; Solratron cal issue

It has been some years since I last worked on a 7081, and I don't have the 
manuals at hand..
Is there a way to clear the EAROM contents and perhaps load starting cal 
constants? I think the EAROM data is corrupt or so far off that it cannot 
compute new values.

Bob