Guys,

The Fluke DVMP Care Plan will give you the ability to know the value of your 
best voltage standard, and it's drift rate, at FAR less expense than buying 
your own JJA:

http://download.flukecal.com/pub/literature/2001531_A_w.pdf

It will give you results with about HALF the uncertainty compared to sending 
your standard to NIST.

How the program works:

"When you purchase a Fluke DVMP, you will be grouped with two other customers 
in the “Round Robin.” A Fluke-owned 732, with all necessary connecting cables 
and clear operating instructions, will be sent to your site for comparison with 
your reference standard. You make and record a series of readings over a period 
of three days and forward the Fluke-owned 732 to the next company in the “Round 
Robin.” You send your data to Fluke for evaluation at the Fluke Primary 
Standards Laboratory. After the
Fluke-owned 732 is received at Fluke, it is once again compared to the 
J-Junction. A value is then assigned to your 10 volt standard, relative to the 
volt as maintained by the Fluke Primary Standards Laboratory using its 
J-Junction, and a report of calibration is sent to you."

They ship their 732B to you "hot" and it is guaranteed to arrive in calibration.

Anyway, to me it makes a lot more sense than building your own JJA.


Best regards,

mitch


-----Original Message-----
From: volt-nuts [mailto:[email protected]] On Behalf Of Tom Knox
Sent: Wednesday, October 19, 2016 8:43 PM
To: Discussion of precise voltage measurement
Subject: [volt-nuts] Sorry: Low-cost Josephson Junction Array Corrected link

Hi Ken;

I corrected the link. There is a lot more to a voltage standard then the Zener, 
a very reasonably priced alternative to a JJA may be a Fluke 732A.

These boxes have been in use for years and their characteristics have been 
extensively documented.

This link may also be of interest.


http://assets.fluke.com/appnotes/calibration/deaver_msc01.pdf

PREDICTABILITY OF SOLID STATE ZENER 
REFERENCES<http://assets.fluke.com/appnotes/calibration/deaver_msc01.pdf>
assets.fluke.com
PREDICTABILITY OF SOLID STATE ZENER REFERENCES David Deaver Fluke Corporation 
PO Box 9090 Everett, AW 98206 425-446-6434 [email protected] Abstract - 
With the ...

Cheers;

Thomas Knox



________________________________
From: volt-nuts <[email protected]> on behalf of Ken Peek 
<[email protected]>
Sent: Wednesday, October 19, 2016 4:35 PM
To: volt-nuts
Subject: Re: [volt-nuts] Low-cost Josephson Junction Array

@Vince:

Thank you for the link!  Very informative!  There is also a nice video showing 
some of the lab techniques, and some cautions working with cryogenic liquids.

I would also like to explore miniature cryocoolers-- as these might be able to 
support a small lower power array (1V) if it doesn't dissipate too much power...

There is already some progress in this area with a QHR made from graphene (at 
the NPL in the UK).  So, maybe the same cryocooler could be used also for the 
low-cost JJA ?

Having a 10V (fixed output) JJA and a 12K9 QHR would be the basis to calibrate 
all other electrical standards in a lab.  It would be nice to have these 
sitting in their cryocoolers cranking out volts and ohms practically 
indefinitely (or as long as you want)-- and if one has even a rubidium atomic 
clock, then no external signals or standards of any kind would be needed.  
Well, that and a triple-point of water cell (which I
have) for temperature calibrations.

-Ken
_______________________________________________
volt-nuts mailing list -- [email protected] To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
volt-nuts Info Page - American Febo 
Enterprises<https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts>
www.febo.com
volt-nuts is a low volume, high SNR list for the discussion of precise voltage 
measurement and related topics. To see the collection of prior postings to the 
list ...


and follow the instructions there.
_______________________________________________
volt-nuts mailing list -- [email protected] To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

_______________________________________________
volt-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Reply via email to