In message <canx10hcdaxkzly__jp_py9tepsap-rorctnapvutubv-ob+...@mail.gmail.com>
, "Dr. David Kirkby (Kirkby Microwave Ltd)" writes:

>> 1. Specially cared for LTZ1000's (See: Fluke)
>
>Do you have a specific reference?  A google found lots of kits on Fluke and
>LTZ1000,  but none of the pages were from Fluke.

They get their own feed of devices from Linear.  Nobody outside Linear
and Fluke knows what is special about them.

>> 2. Lots of LTZ1000's to get sqrt(N) reductions.

>It seems to me
>that you would not gain sqrt(N) unless N is quite small. How small, is
>anyone's guess.

N=4 is commonly used.


-- 
Poul-Henning Kamp       | UNIX since Zilog Zeus 3.20
[email protected]         | TCP/IP since RFC 956
FreeBSD committer       | BSD since 4.3-tahoe    
Never attribute to malice what can adequately be explained by incompetence.