If you want to do a comparison measurement with consideration of the actual 
measurement uncertainty, then what is relevant is not the theoretical 
resolution of the meter but what the meter data sheet tells you with respect of 
that matter. It is ususlly called the transfer accuracy. Applies when you do 
two measurements with one meter, at similar voltages. E.g. the drift of a 
reference after it has been exposed to a temperature change. It takes into 
account also things like differential non-linearity of the A/D, short term 
noise/flicker/drift issues and so on.
For a 3458A at 10V, it is 0.05+0.05ppm (range and reading), so 0,1ppm (0.2ppm 
for two measurements in total). Now for 6.5 digit DMMs it is obviously less, 
and not always specified. As an alternative, if the value is not given, one can 
take the uncertainty referenced to the range (see Fluke 'Calibration: Philosphy 
in Practice'). The 34401 has the A/D linearity specified, it is 3ppm. That 
helps as well. No way arround analyzing such transfer measurement uncertainty 
for any given meter based on the data sheet values.

cheers