 but I recently noticed the meter reads -0.08 when the
 input is open on the 1V range and the null switch is on 0.0001
 This equates to a -8uV reading.
 I believe this is due to the input bias current of the LTC1050
       *   *   *
 However, this would imply an 8pA bias current,
 which is much higher than expected.

I am attaching the relevant portion of the LTC1050 datasheet. Note that at 25C, the 1050 has an input offset current spec of +/- 20pA (both typical -- max is ~5x worse). I would expect a total input offset of +/- 20uV (typical) in your circuit. (The 0.5uV typical input offset voltage is not a significant contributor in this context.)

Best regards