Illya,

That is a great idea.  I will give it a try.

Thanks

On Sat, Nov 18, 2017 at 4:47 AM, Illya Tsemenko <il...@xdevs.com> wrote:

> Since you have 732A, testing should be easy enough. Calibrate faulty meter
> for zero and DCV 10V to 732A, record CAL? 2,1 value. This is your LTZ
> output. Then leave it running for few days to drift away and calibrate
> again to same 732A. Check CAL? 2,1 again. Calculate the difference and if
> it matches output drift (that 1.1ppm/day you mention) - you can be 80% sure
> that A9 is a problem. Other 19% go to A1 and A3 circuits, as 7V is not used
> directly in the meter, and there are still gain parts to get +12 and
> -12VREF on A3 and 10Vish bipolar levels on A1. If your CAL? 2,1 stays same
> (within 0.3ppm) then A9 is fine.
>
>
> On November 18, 2017 12:59:53 PM GMT+08:00, Randy Evans <
> randyevans2...@gmail.com> wrote:
>>
>> I just received an Agilent 3458A that has a problem with noise and a
>> drifting voltage measurements.  I am using two Fluke 732As to compare
>> absolute voltage measurements over time against the Agilent and an HP
>> 3458A.  The HP unit has a new A3 ADC card and seems to be very stable and
>> low noise, so is being used for comparison.  I have been doing simultaneous
>> absolute voltage measurements and DC Cal Constant measurements several
>> times a day and then calculating the drift rates of the two units using the
>> HP Service Note 18 procedure.
>>
>>
>>
>> The results indicate the Cal Constant drift rate of both units is very
>> similar and within spec per Service Note 18.  However, the absolute value
>> measurements show the Agilent unit changing 1.1 ppm over a day whereas the
>> HP unit is within a tenth of a ppm over a day.  In my way of thinking the
>> Cal Constant procedure assumes the voltage reference board in the 3458A is
>> stable, hence the absolute value reading should remain essentially constant
>> after each ACAL DCV, which is the case with the HP unit.  Since the Agilent
>> unit shows a steady drift in the absolute reading, this would indicate to
>> me that the voltage reference board is likely the cause of the problem, and
>> is also likely the cause of the noisy readings.  If so, this is a
>> “relatively” easy fix (I have several 3458A voltage reference boards, one
>> of which has been continuously powered up for several years).
>>
>>
>>
>> The issue is that I have to make a decision to keep or return the Agilent.
>> It has a cal seal on it and if I open the unit up to change the voltage
>> reference board, I own it and can’t return it.  I would appreciate an
>> opinion from the members of the group as to what they think the odds are
>> that the voltage reference board is the source of the problems with the
>> Agilent 3458A.
>>
>>
>>
>> Thanks,
>>
>>
>>
>> Randy Evans
>>
>>
_______________________________________________