I did it time ago.
Warmed up ~3 hours / Measure value with my 3548A.
Next trimmed to mid of trimmable range (pot to center).
Than trimmed (by shorting step by step) as close as possible to 10.000V.
/PeLuLe
 

> Gesendet: Sonntag, 19. November 2017 um 06:51 Uhr
> Von: "Jeremy Nichols" <[email protected]>
> An: [email protected]
> Betreff: [volt-nuts] HP-3455A calibration
>
> My new-to-me 3455A DVM (s/n 1622A06699) does not want to calibrate to the 10 
> Volt level although 1 Volt is fine. The 10V adjustment pot in the 11177B 
> reference assembly simply doesn't adjust far enough—I can get the reading 
> down to 10.00020 Volts but no lower.
> 
> Just for the heck of it, I picked up another reference assembly from an eBay 
> seller. It's an "A" assembly rather than the "B" the 3455 came with but it 
> works fine and I am able to calibrate both 1V and 10V. 
> 
> The reference assemblies have a set of jumpers for "coarse adjust" so it 
> should be possible to tweak the "B" version back into calibration. Does 
> anyone have experience doing this? The service manual just says "return to 
> the factory." It would be nice to hear from someone who has worked with these 
> things before I start messing with the jumpers.
> 
> Jeremy
> 
> Sent from my iPad 4
> _______________________________________________
> volt-nuts mailing list -- [email protected]
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
_______________________________________________
volt-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Reply via email to