Todd, Since the KE5FX NVRAM download software works, that would seem to indicate the GPIB is working, wouldn't it (download completes and checksums OK)?
Randy On Sat, Dec 30, 2017 at 9:52 PM, Randy Evans <[email protected]> wrote: > Todd, > > The minimal quick test would be greatly appreciated if you have the time. > I know this is above and beyond what I would expect but it would be really > useful to see if there is any hope for the GPIB on this unit. I will do > the service note 13 procedure in the morning. > > Thanks again, > > Randy > > On Sat, Dec 30, 2017 at 8:58 PM, Todd Micallef <[email protected]> > wrote: > >> Randy, >> >> Short of swapping out the A5 digital board I have no other ideas. If you >> don't have a NI or Agilent adapter, I am sure I can find a spare for you >> to >> test. >> >> There is a Service Note 3458A-13A that deals with GPIB communication >> failures. There is a test within the document to see if a shorting jumper >> needs to be installed. >> I can try reducing the code to something minimal for a quick test if you >> want. It will take some time in the morning. >> >> Todd >> >> On Sat, Dec 30, 2017 at 11:18 PM, Randy Evans <[email protected]> >> wrote: >> >> > Todd, >> > >> > Same as before. Still not working. All it does is print out "HP3458A" >> on >> > the output console and then just sits there. >> > >> > Randy >> > >> > On Sat, Dec 30, 2017 at 7:43 PM, Todd Micallef <[email protected]> >> > wrote: >> > >> > > Randy, >> > > >> > > I just remembered that I changed the duration from 30 minutes to 2 >> > minutes >> > > to speed up the test. I forgot to change it back. >> > > >> > > Todd >> > > >> > > On Sat, Dec 30, 2017 at 10:27 PM, Todd Micallef <[email protected]> >> > > wrote: >> > > >> > > > Randy, >> > > > >> > > > I just sent it. I haven't modified Martin's code yet. That can be >> done >> > > > once you confirm it works for you. >> > > > >> > > > Todd >> > > > >> > > > On Sat, Dec 30, 2017 at 10:12 PM, randyevans2688 < >> > > [email protected] >> > > > > wrote: >> > > > >> > > >> Todd, >> > > >> If you don't mind, just send the code directly to me. >> > > >> Randy >> > > >> >> > > >> Sent from my Verizon, Samsung Galaxy smartphone >> > > >> -------- Original message --------From: Todd Micallef < >> > > >> [email protected]> Date: 12/30/17 6:40 PM (GMT-08:00) To: >> > > Discussion >> > > >> of precise voltage measurement <[email protected]> Subject: Re: >> > > >> [volt-nuts] GPIB Issue with 3458A >> > > >> Randy, >> > > >> >> > > >> I have not written anything else for the 3458A using EzGPIB. If >> you >> > > have >> > > >> a >> > > >> National Instruments or Agilent adapters, I have a handful of >> Python >> > > code >> > > >> I >> > > >> use to measure some of my standards. >> > > >> >> > > >> If you are reading the temperature from the HP meter, I would think >> > that >> > > >> the problem lies in how the triggering is configured. The >> triggering >> > has >> > > >> been changed in this script so it may work for you. There was an >> > > >> additional >> > > >> command added for disabling the buffer. >> > > >> The code can be posted here, if possible or I can email it to you >> for >> > > >> testing. >> > > >> >> > > >> Todd >> > > >> >> > > >> On Sat, Dec 30, 2017 at 9:26 PM, randyevans2688 < >> > > [email protected] >> > > >> > >> > > >> wrote: >> > > >> >> > > >> > Todd, thanks for the help. Where will you post the new code? >> > > >> > BTW, do you have any very simple code that is likely to work. I >> am >> > > >> > interested in seeing if anything other than KE5FX's memory dump >> > > program >> > > >> > will work on my HP3458A. >> > > >> > Thanks again, >> > > >> > Randy >> > > >> > >> > > >> > Sent from my Verizon, Samsung Galaxy smartphone >> > > >> > -------- Original message --------From: Todd Micallef < >> > > >> [email protected]> >> > > >> > Date: 12/30/17 6:13 PM (GMT-08:00) To: Discussion of precise >> > voltage >> > > >> > measurement <[email protected]> Subject: Re: [volt-nuts] GPIB >> > Issue >> > > >> with >> > > >> > 3458A >> > > >> > Randy, >> > > >> > >> > > >> > I found the Prologix adapters and have modified the code in the >> > > original >> > > >> > script. I am doing a test run now and will post the script when >> I am >> > > >> > certain I didn't make it worse. >> > > >> > >> > > >> > Todd >> > > >> > >> > > >> > On Sat, Dec 30, 2017 at 7:15 PM, Randy Evans < >> > > [email protected]> >> > > >> > wrote: >> > > >> > >> > > >> > > Todd, >> > > >> > > >> > > >> > > I have swapped cables, moved the Prologix adapter to both >> meters >> > > >> (Agilent >> > > >> > > works, HP doesn't), tried Martins code as well as Xdevs's. I >> have >> > > not >> > > >> > > tried another PC yet, maybe the next step. >> > > >> > > >> > > >> > > Here is the code that Martin modified for two meters, as well >> as >> > the >> > > >> > > modified Xdev's code per your directions (V1.02). >> > > >> > > >> > > >> > > Thanks, >> > > >> > > >> > > >> > > Randy >> > > >> > > >> > > >> > > On Sat, Dec 30, 2017 at 3:55 PM, Todd Micallef < >> > [email protected] >> > > > >> > > >> > > wrote: >> > > >> > > >> > > >> > > > Randy, >> > > >> > > > >> > > >> > > > I will continue to look for my adapter. I haven't used it for >> > > quite >> > > >> a >> > > >> > > > while. >> > > >> > > > >> > > >> > > > I have seen similar issues and swapped cables, moved the >> adapter >> > > to >> > > >> the >> > > >> > > > other meter, etc... One of my cables was bad from one side of >> > the >> > > >> > > connector >> > > >> > > > to the other (on the same end of the cable) and caused me >> lots >> > of >> > > >> > wasted >> > > >> > > > time. Weird reading errors were happening. >> > > >> > > > >> > > >> > > > Would you email me Martin's code so I can test it here and I >> > will >> > > >> see >> > > >> > if >> > > >> > > > anything can be done to replicate your issue. >> > > >> > > > >> > > >> > > > Todd >> > > >> > > > >> > > >> > > > >> > > >> > > > On Sat, Dec 30, 2017 at 6:42 PM, Randy Evans < >> > > >> [email protected] >> > > >> > > >> > > >> > > > wrote: >> > > >> > > > >> > > >> > > > > Todd, >> > > >> > > > > >> > > >> > > > > I tried adding the line but it made no difference. >> > > >> > > > > >> > > >> > > > > Randy >> > > >> > > > > >> > > >> > > > > On Sat, Dec 30, 2017 at 12:19 PM, Randy Evans < >> > > >> > > [email protected]> >> > > >> > > > > wrote: >> > > >> > > > > >> > > >> > > > > > Todd, >> > > >> > > > > > >> > > >> > > > > > Does it matter where I put the line in the Init_Device >> > > >> procedure? >> > > >> > > > > > >> > > >> > > > > > >> > > >> > > > > > procedure Init_Device; >> > > >> > > > > > begin; >> > > >> > > > > > EZGPIB_BusWriteData(gpib_address,GET_ID); >> // >> > > Get >> > > >> ID >> > > >> > of >> > > >> > > > > meter >> > > >> > > > > > if EZGPIB_BusWaitForData(gpib_add >> ress,meter_id,Timeout) >> > > then >> > > >> > > > > > EZGPIB_ScreenWriteLn(meter_id); >> > > >> > > > > > >> > > >> > > > > > EZGPIB_BusWriteData(gpib_address,TRIGGER_SOURCE); >> // >> > > Hold >> > > >> > > > trigger >> > > >> > > > > > until meter can be setup >> > > >> > > > > > EZGPIB_BusWriteData(gpib_address,'INBUF ON'); >> > > >> > > > > > EZGPIB_BusWriteData(gpib_address,AUTO_ZERO); >> > > >> > > > > > EZGPIB_BusWriteData(gpib_address,OUTPUT_FORMAT); >> > > >> > > > > > end; >> > > >> > > > > > >> > > >> > > > > > >> > > >> > > > > > Randy >> > > >> > > > > > >> > > >> > > > > > On Sat, Dec 30, 2017 at 11:20 AM, Todd Micallef < >> > > >> > [email protected] >> > > >> > > > >> > > >> > > > > > wrote: >> > > >> > > > > > >> > > >> > > > > >> Randy, >> > > >> > > > > >> >> > > >> > > > > >> I am the person that wrote that EzGPIB script. It was my >> > > first >> > > >> > > before >> > > >> > > > I >> > > >> > > > > >> moved on to Python and I have only written a couple for >> the >> > > >> > > different >> > > >> > > > > >> meters I have on hand. Version 1.01 is the most current >> > > >> version I >> > > >> > > have >> > > >> > > > > for >> > > >> > > > > >> the 3458A. >> > > >> > > > > >> I see that I should have added at least one more >> command to >> > > the >> > > >> > > meter >> > > >> > > > > >> initialization. The additional init command is used in >> the >> > > >> > scripts >> > > >> > > on >> > > >> > > > > >> xDevs. >> > > >> > > > > >> >> > > >> > > > > >> Add the following line in the Init_Device procedure >> > > >> > > > > >> EZGPIB_BusWriteData(gpib_address,'END ALWAYS'); >> > > >> > > > > >> >> > > >> > > > > >> This tells the meter to send an EOI on every read. >> > > >> > > > > >> >> > > >> > > > > >> I am looking for one of my Prologix adapters or I would >> > test >> > > >> this. >> > > >> > > > When >> > > >> > > > > I >> > > >> > > > > >> find it, I will also test it. I want to make sure there >> are >> > > no >> > > >> > > further >> > > >> > > > > >> issues. >> > > >> > > > > >> Also, do both meters run the same firmware versions? I >> > wonder >> > > >> if >> > > >> > the >> > > >> > > > HP >> > > >> > > > > >> meter is using a much older version. >> > > >> > > > > >> >> > > >> > > > > >> Todd >> > > >> > > > > >> >> > > >> > > > > >> On Sat, Dec 30, 2017 at 10:42 AM, Randy Evans < >> > > >> > > > [email protected] >> > > >> > > > > > >> > > >> > > > > >> wrote: >> > > >> > > > > >> >> > > >> > > > > >> > Todd, >> > > >> > > > > >> > >> > > >> > > > > >> > I have version 1.01. Yes, please e-mail me your >> modified >> > > >> > version. >> > > >> > > > > >> Maybe it >> > > >> > > > > >> > will help. >> > > >> > > > > >> > >> > > >> > > > > >> > Thanks, >> > > >> > > > > >> > >> > > >> > > > > >> > Randy >> > > >> > > > > >> > >> > > >> > > > > >> > On Sat, Dec 30, 2017 at 4:24 AM, Todd Micallef < >> > > >> > > [email protected] >> > > >> > > > > >> > > >> > > > > >> > wrote: >> > > >> > > > > >> > >> > > >> > > > > >> > > Randy, >> > > >> > > > > >> > > >> > > >> > > > > >> > > Is there a version number at the top of the file? I >> had >> > > two >> > > >> > > > versions >> > > >> > > > > >> > (1.00 >> > > >> > > > > >> > > and 1.01). I can't remember all that changed between >> > the >> > > >> two >> > > >> > > since >> > > >> > > > > it >> > > >> > > > > >> has >> > > >> > > > > >> > > been a while but I did find an additional timeout >> code >> > > that >> > > >> > > > limited >> > > >> > > > > >> the >> > > >> > > > > >> > > wait time for reading data to 1 minute. I am >> guessing I >> > > >> had a >> > > >> > > > > problem >> > > >> > > > > >> > with >> > > >> > > > > >> > > mine and added it. >> > > >> > > > > >> > > >> > > >> > > > > >> > > I can email it to you if you want to try it. If it >> > works >> > > >> for >> > > >> > > you, >> > > >> > > > I >> > > >> > > > > >> will >> > > >> > > > > >> > > have Illya update it on xDevs. >> > > >> > > > > >> > > >> > > >> > > > > >> > > Todd >> > > >> > > > > >> > > >> > > >> > > > > >> > > On Sat, Dec 30, 2017 at 12:39 AM, Randy Evans < >> > > >> > > > > >> [email protected]> >> > > >> > > > > >> > > wrote: >> > > >> > > > > >> > > >> > > >> > > > > >> > > > I am still unable to get my HP-3458A to work with >> the >> > > >> xdevs >> > > >> > > > > program >> > > >> > > > > >> > > > "HP3458A_Measure_DMM_Noise.488" ( >> > > >> > > https://xdevs.com/article/dmm >> > > >> > > > > >> _noise/ >> > > >> > > > > >> > ). >> > > >> > > > > >> > > > It >> > > >> > > > > >> > > > works fine with my Agilent 3458A and with a >> friend's >> > > >> > HP-3458A >> > > >> > > > and >> > > >> > > > > >> his >> > > >> > > > > >> > > > Agilent 3458A using the EZGPIB program. The SW >> > > versions >> > > >> are >> > > >> > > > > >> identical >> > > >> > > > > >> > > with >> > > >> > > > > >> > > > my friend's DMMs, the HP units are (8.2) and with >> the >> > > >> > Agilent >> > > >> > > > > units >> > > >> > > > > >> are >> > > >> > > > > >> > > > (9.2). However, both of my units work fine with >> the >> > > >> KE5FX >> > > >> > > GPIB >> > > >> > > > > >> Toolkit >> > > >> > > > > >> > > > HP3458.EXE program to download the NVRAM CAL data. >> > > This >> > > >> > would >> > > >> > > > > seem >> > > >> > > > > >> to >> > > >> > > > > >> > > > indicate that my HP-3458 GPIB works at least >> > partially. >> > > >> > > > > >> > > > >> > > >> > > > > >> > > > Does anyone have a EZGPIB program that I could >> try to >> > > run >> > > >> > > (that >> > > >> > > > is >> > > >> > > > > >> > > > different from the Xdev program)? It would be >> nice >> > to >> > > >> see >> > > >> > if >> > > >> > > I >> > > >> > > > > >> could >> > > >> > > > > >> > get >> > > >> > > > > >> > > > something to work other than the HP3458.EXE >> program. >> > > Or >> > > >> if >> > > >> > > any >> > > >> > > > > one >> > > >> > > > > >> has >> > > >> > > > > >> > > > some ideas on what to look for. >> > > >> > > > > >> > > > >> > > >> > > > > >> > > > Thanks, >> > > >> > > > > >> > > > >> > > >> > > > > >> > > > Randy Evans >> > > >> > > > > >> > > > >> > > >> > > > > >> > > > On Tue, Dec 26, 2017 at 7:24 PM, Randy Evans < >> > > >> > > > > >> [email protected] >> > > >> > > > > >> > > >> > > >> > > > > >> > > > wrote: >> > > >> > > > > >> > > > >> > > >> > > > > >> > > > > I am hoping some of you experts might be able to >> > > assist >> > > >> > > with a >> > > >> > > > > >> > strange >> > > >> > > > > >> > > > > problem i am having trying to make automated >> > > >> measurements >> > > >> > > > using >> > > >> > > > > >> > EZGPIB >> > > >> > > > > >> > > > with >> > > >> > > > > >> > > > > an Agilent and an HP 3458A. I have a program >> given >> > > to >> > > >> me >> > > >> > by >> > > >> > > > > >> Martin >> > > >> > > > > >> > > > > Reynolds, which is a variant on the Xdevs noise >> > > >> > measurement >> > > >> > > > > >> program. >> > > >> > > > > >> > > > > Martin's program continuously reads the voltage >> > > >> > measurements >> > > >> > > > > >> > > sequentially >> > > >> > > > > >> > > > > from the two meters and then stores then in a >> > > >> measurements >> > > >> > > > file, >> > > >> > > > > >> > along >> > > >> > > > > >> > > > with >> > > >> > > > > >> > > > > the temperature measurements from both meters. >> The >> > > >> > program >> > > >> > > > > works >> > > >> > > > > >> > fine >> > > >> > > > > >> > > on >> > > >> > > > > >> > > > > his Agilent and HP 3458As. >> > > >> > > > > >> > > > > >> > > >> > > > > >> > > > > On my two 3458As, it takes the temperature >> > > measurements >> > > >> > just >> > > >> > > > > fine >> > > >> > > > > >> but >> > > >> > > > > >> > > > only >> > > >> > > > > >> > > > > takes a voltage measurement from the Agilent >> unit. >> > > >> The HP >> > > >> > > > unit >> > > >> > > > > >> will >> > > >> > > > > >> > > not >> > > >> > > > > >> > > > > make any voltage measurements although I can see >> > both >> > > >> the >> > > >> > > > 3458As >> > > >> > > > > >> in >> > > >> > > > > >> > > talk >> > > >> > > > > >> > > > > states and, of course, it does sent the >> temperature >> > > >> > > > > measurements. >> > > >> > > > > >> > > > However, >> > > >> > > > > >> > > > > if I turn the HP unit off and then on again, >> the HP >> > > >> will >> > > >> > > start >> > > >> > > > > to >> > > >> > > > > >> > take >> > > >> > > > > >> > > > > voltage and temperature measurements and send >> them >> > to >> > > >> the >> > > >> > > PC. >> > > >> > > > > >> > > > > >> > > >> > > > > >> > > > > Any ideas on what might be causing the strange >> > > >> behavior? >> > > >> > > > > >> > > > > >> > > >> > > > > >> > > > > Thanks, >> > > >> > > > > >> > > > > >> > > >> > > > > >> > > > > Randy Evans >> > > >> > > > > >> > > > > >> > > >> > > > > >> > > > _______________________________________________ >> > > >> > > > > >> > > > volt-nuts mailing list -- [email protected] >> > > >> > > > > >> > > > To unsubscribe, go to >> https://www.febo.com/cgi-bin/ >> > > >> > > > > >> > > > mailman/listinfo/volt-nuts >> > > >> > > > > >> > > > and follow the instructions there. >> > > >> > > > > >> > > > >> > > >> > > > > >> > > _______________________________________________ >> > > >> > > > > >> > > volt-nuts mailing list -- [email protected] >> > > >> > > > > >> > > To unsubscribe, go to https://www.febo.com/cgi-bin/ >> > > >> > > > > >> > > mailman/listinfo/volt-nuts >> > > >> > > > > >> > > and follow the instructions there. >> > > >> > > > > >> > > >> > > >> > > > > >> > _______________________________________________ >> > > >> > > > > >> > volt-nuts mailing list -- [email protected] >> > > >> > > > > >> > To unsubscribe, go to https://www.febo.com/cgi-bin/ >> > > >> > > > > >> > mailman/listinfo/volt-nuts >> > > >> > > > > >> > and follow the instructions there. >> > > >> > > > > >> > >> > > >> > > > > >> _______________________________________________ >> > > >> > > > > >> volt-nuts mailing list -- [email protected] >> > > >> > > > > >> To unsubscribe, go to https://www.febo.com/cgi-bin/m >> > > >> > > > > >> ailman/listinfo/volt-nuts >> > > >> > > > > >> and follow the instructions there. >> > > >> > > > > >> >> > > >> > > > > > >> > > >> > > > > > >> > > >> > > > > _______________________________________________ >> > > >> > > > > volt-nuts mailing list -- [email protected] >> > > >> > > > > To unsubscribe, go to https://www.febo.com/cgi-bin/ >> > > >> > > > > mailman/listinfo/volt-nuts >> > > >> > > > > and follow the instructions there. >> > > >> > > > > >> > > >> > > > _______________________________________________ >> > > >> > > > volt-nuts mailing list -- [email protected] >> > > >> > > > To unsubscribe, go to https://www.febo.com/cgi-bin/ >> > > >> > > > mailman/listinfo/volt-nuts >> > > >> > > > and follow the instructions there. >> > > >> > > > >> > > >> > > >> > > >> > > _______________________________________________ >> > > >> > > volt-nuts mailing list -- [email protected] >> > > >> > > To unsubscribe, go to https://www.febo.com/cgi-bin/ >> > > >> > > mailman/listinfo/volt-nuts >> > > >> > > and follow the instructions there. >> > > >> > > >> > > >> > _______________________________________________ >> > > >> > volt-nuts mailing list -- [email protected] >> > > >> > To unsubscribe, go to https://www.febo.com/cgi-bin/ >> > > >> > mailman/listinfo/volt-nuts >> > > >> > and follow the instructions there. >> > > >> > _______________________________________________ >> > > >> > volt-nuts mailing list -- [email protected] >> > > >> > To unsubscribe, go to https://www.febo.com/cgi-bin/ >> > > >> > mailman/listinfo/volt-nuts >> > > >> > and follow the instructions there. >> > > >> > >> > > >> _______________________________________________ >> > > >> volt-nuts mailing list -- [email protected] >> > > >> To unsubscribe, go to https://www.febo.com/cgi-bin/m >> > > >> ailman/listinfo/volt-nuts >> > > >> and follow the instructions there. >> > > >> _______________________________________________ >> > > >> volt-nuts mailing list -- [email protected] >> > > >> To unsubscribe, go to https://www.febo.com/cgi-bin/m >> > > >> ailman/listinfo/volt-nuts >> > > >> and follow the instructions there. >> > > >> >> > > > >> > > > >> > > _______________________________________________ >> > > volt-nuts mailing list -- [email protected] >> > > To unsubscribe, go to https://www.febo.com/cgi-bin/ >> > > mailman/listinfo/volt-nuts >> > > and follow the instructions there. >> > > >> > _______________________________________________ >> > volt-nuts mailing list -- [email protected] >> > To unsubscribe, go to https://www.febo.com/cgi-bin/ >> > mailman/listinfo/volt-nuts >> > and follow the instructions there. >> > >> _______________________________________________ >> volt-nuts mailing list -- [email protected] >> To unsubscribe, go to https://www.febo.com/cgi-bin/m >> ailman/listinfo/volt-nuts >> and follow the instructions there. >> > > _______________________________________________ volt-nuts mailing list -- [email protected] To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.