Hi Kees,

The 8502A manual on KO4BB's site does contain a section on calibration adjustments.
http://www.ko4bb.com/getsimple/index.php?id=download&file=Fluke/Fluke_8502A_6_12_Digit_Multimeter_Service_Manual.pdf

Hope that helps you a bit.


Met vriendelijke groet,
Rob.


Op 5-11-2018 om 22:00 schreef cgm de bruijn:
Hi, all

I have not been able to find a service and/or user manual for a vintage
(1978-1982) DMM Fluke 7600A that I picked up and want to check/calibrate
against my calibrated Keithley 2001 multimeter.

The instrument is still in good shape and up and running:  offset and gain
need adjustment, but how?

Does anyone know of a copy of the user manual/schematic?  A helpful Fluke
technician came up with the 8502A manual that comes close, but gives no
calibration procedure. Further surfing on Google had no success.

Thanks in advance,




Kees de Bruijn

[email protected]