I have Qty 2 of 731B, Both battery packs were NiCd and were dead when I
received them. I built two packs with the largest amp-hour NiMH  D cells
that I cold find. I drove all day with the references to a cal lab two
years ago. Even with the battery back up, the references required more than
a day on the cal lab bench to settle.

cheers

Brian Smith

On Tue, Aug 6, 2019 at 11:30 AM Tom Knox <[email protected]> wrote:

> An interesting side note, there are 9AH batteries with the exact same
> dimensions as the 7AH. In addition Power-Sonic (and perhaps other
> manufactures) have a DOT exemption allowing them to be shipped on all
> aircraft which can save time and trouble when shipping the 732B for
> calibration or other reasons. I think that the big concern affecting
> calibration is if the unit actually powers down and the extra battery life
> and DOT exemption can really help avoid the battery dying during shipping.
> Cheers;
>
> Tom Knox
>
> 303-554-0307
>
> [email protected]
>
> "Peace is not the absence of violence, but the presence of Justice" Both
> MLK and Albert Einstein
>
> ________________________________
> From: volt-nuts <[email protected]> on behalf of Illya
> Tsemenko <[email protected]>
> Sent: Tuesday, August 6, 2019 9:35 AM
> To: [email protected] <[email protected]>
> Subject: Re: [volt-nuts] Fluke voltage standard
>
> Because any power cycle or temperature cycle expected to cause output
> shift due to hysteresis and these standards designed to be always
> powered on without any interruptions, otherwise uniform constant drift
> characteristics cannot be guaranteed. IN CAL LED circuit monitors power
> and thermistor value, so when your battery depleted below threshold, or
> temperature outside of the normal window LED will be turned off. Fluke
> recommends to perform recalibration after standard lost IN CAL LED
> status. Typical battery life of 732B is ~3 days and that can be extended
> once you have external chargers/+12V DC supply for backup power.
>
> This concept applies not just for 732 series standards, but for any
> voltage standard which try to claim ppm annual stability. That also
> includes 8.5-digit DMMs like 3458A and multi-function calibrators like
> 5720A/5730A. Less stress changes in temperature/power - longer life for
> instrument. Drift might be bit more, but it would be predicable once
> characterized properly.
>
> On 8/6/2019 11:03 PM, Victor Silva via volt-nuts wrote:
> > Can anyone explain why the Fluke Voltage standards, like 732B, which I
> use
> > as my primary standard, turn off the IN-CAL LED when they lose power
> > (including battery)? I believe the same is true for the 732A, but I do
> not
> > have one so I'm not sure on that model.
> >
> >
> >
> > My 732B lost power during the hurricane in NC last year and the CAL LED
> is
> > off. I realize I can press the reset button to turn it on, but that
> doesn't
> > really make it in cal if it was out.
> >
> >
> >
> > Is the CAL LED really just an indicator to warn that the unit has lost
> power
> > or does the unit actually go out of spec after a power loss?
> >
> > I can see a scenario where there was a power loss then power came back,
> > without that LED you could use the standard before it stabilized.
> >
> > But if you allow it to stabilize for some time, my question is will it
> be in
> > CAL still [but with the LED off]?
> >
> >
> >
> > Thanks,
> >
> > Victor Silva
> >
> > _______________________________________________
> > volt-nuts mailing list -- [email protected]
> > To unsubscribe, go to
> http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
> > and follow the instructions there.
>
>
>
>
> _______________________________________________
> volt-nuts mailing list -- [email protected]
> To unsubscribe, go to
> http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
> and follow the instructions there.
> _______________________________________________
> volt-nuts mailing list -- [email protected]
> To unsubscribe, go to
> http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
> and follow the instructions there.
>
_______________________________________________
volt-nuts mailing list -- [email protected]
To unsubscribe, go to 
http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
and follow the instructions there.

Reply via email to