-------- In message <CANX10hCVy9UPspXW1dGggb+tUEafn0mQwvnU=avjfegczdv...@mail.gmail.com> , "Dr. David Kirkby" writes:
>> > So I conclude if a 3458A needed adjusting, [...] >> >> There is nothing in the 3458A that *can* be adjusted. > >Not even by software? I thought there were calibration constants. You short the inputs, front&rear, run "CAL 0" on both of them. You connect approx 10V to front or rear, run "CAL $voltage" You connect approx 10kOhm to front or rear, run "CAL $resistance" Done. There is an AC-flatness calibration, following the same basic schedule, but unless you plan to measure or sample non-audio-range frequencies, you can ignore that. -- Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 [email protected] | TCP/IP since RFC 956 FreeBSD committer | BSD since 4.3-tahoe Never attribute to malice what can adequately be explained by incompetence. _______________________________________________ volt-nuts mailing list -- [email protected] To unsubscribe, go to http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com and follow the instructions there.