I am trying to calibrate a Solartron 7071 I've been working on.

 

I set the calibrator to 800V and issued the command:
calibrate,vdc=1000,high=800
but part way through the calibrator went into overload (too much current
drawn).

 

The calibrator doesn't go into overload when feeding the 3458A.

 

The calibrate command finished, but now the meter is behaving oddly.

 

With input values up to 14V DC the correct value is displayed.   If I
increase the value above there, I get:

               Input                     Display

15V                       1134.061 

16V                       1209.664 

17V                       1285.276 

18V                       1360.867 

19V                       1436.474 OVERLOAD

calibrate,dump

OK

Vdc calibration constants

-1.78813934E-06

164.781288E-03

-2.92062759E-06

1.64774580E+00

-1.66893005E-06

16.4787127E+00

-2.14576721E-06

164.788683E+00

-2.68220901E-06

1.64800913E+03

Vac calibration constants

326.985921E-03

3.27081809E+00

32.5331218E+00

325.456807E+00

3.27348303E+03

kOhm calibration constants

-1.78813934E-06

166.021009E-03

-2.92062759E-06

1.66000890E+00

-1.66893005E-06

16.6035262E+00

-2.92062759E-06

165.948071E+00

-1.66893005E-06

1.65979766E+03

Top Ohms range calibration constants

-1.66893005E-06

10.0332974E+03

602.359830E-03

Ratio terminals calibration constant

16.4792383E+00

Zener current token value

085

 

1.      Reading that suggests the the 1000V high calibration factor appears
to be in the right "ball park" - confirmation would be good
2.      Any suggestions on what's gone awry with the autorange?
3.      How to track down the problem that's causing the calibrator
overload.

 

Has anyone been there, done that?  Any suugestions welcome.

 

Cheers, David

 

_______________________________________________
volt-nuts mailing list -- [email protected] -- To unsubscribe send an 
email to [email protected]
To unsubscribe, go to and follow the instructions there.

Reply via email to