Bob,

You are bringing up good points.  Perhaps Jed or Mizuno will have an 
opportunity to pursue a cleanup effort to eliminate these noise items.  Do you 
have time to follow up on this idea to see if the numbers we are getting make 
sense?  I have never had an opportunity to go into the details like is 
happening now with regard to thermal measurements.  I have chased down plenty 
of other forms of analog noise, but not this particular type.

Could it be that we are expecting better input than is normally possible?  I am 
amazed to see how low the total temperature errors appear to be according to my 
simulation output.  I had no idea that .02 degrees of temperature change could 
even be measured with normal lab equipment.  The calculations can be done to 
many decimal places, but if the input is not accurate enough then the final 
results must be limited.

When I started this project I figured that .1 degree accuracy would be about as 
good as I could hope to determine.  I have been pleasantly surprised so far.  
Something might come up that bites me in the rear, but so far so good. :-)

Keep asking those important questions and sharing your impressive systems 
knowledge.

Dave

 

 

 

-----Original Message-----
From: Bob Cook <[email protected]>
To: vortex-l <[email protected]>
Sent: Sat, Jan 31, 2015 6:32 pm
Subject: Re: [Vo]:Alternate Calculation and Calibration Method for Mizuno Report



Dave-
 
My experience with older thermocouples is that reference t/c voltages must be 
well controlled to get accurate temperature readings based on calibrations.   
It looks to me like the reference voltages that the t/c instrumentation uses 
may have a cyclic variation that is responsible for the cyclic variation your 
model predicts.  I wonder what the instrumentation specs say about this issue.  
It would be useful to review those specifications and what the vendor says 
about stability of the readings.  Frequently T/C leads need to be electrically 
shielded to prevent small voltage pickup and disruption of the voltage 
determination.  With the small temperature changes only very voltage changes 
are produced by the t/c's--millivolts and less. 
 
Bob Cook
  
----- Original Message ----- 
  
From:   David   Roberson 
  
To: [email protected] 
  
Sent: Friday, January 30, 2015 6:07   PM
  
Subject: Re: [Vo]:Alternate Calculation   and Calibration Method for Mizuno 
Report
  


This is an   ongoing project I suppose.  I will check the Oct 20 data again, 
but I   believe I got roughly the same amount of excess.  That seems too good 
to   be true.

What do you mean by considering that the reactor vessel is   capturing 60% of 
the heat?  Are you referring to the idea that the water   and reactor have a 
combined capture of 100%?  That would seem logical if   the thermal capacity of 
each is considered.

It is good to continue with   the calibrations and I hope that the ambient is 
better controlled at this   phase.  We can make the calorimeter work well once 
it is properly   compensated to take out the signal droop.  The main issue is 
to keep the   ambient variations to a minimum.   One thought to consider.  If   
something happens to disrupt the system, you can wait a few time constants   
until the ambient average changes work their way out of the calorimeter   
system.  A good test of that condition is to measure the difference   between 
the ambient and the coolant.  Once that difference is equal to   the nominal 
value set by the product of the thermal resistance and the leakage   powers, 
you are good to go.

Of course, you must ensure that the ambient   is not varying too far since that 
immediately impacts the heat flowing into   and out of the thermal capacity.  
It is assumed that the average ambient   is constant which will prevent any 
major transients.

This system should   work well with the right precautions and compensations.

Dave
  


  


  


  
-----Original   Message-----
From: Jed Rothwell <[email protected]>
To:   vortex-l <[email protected]>
Sent: Fri, Jan 30, 2015 8:48   pm
Subject: Re: [Vo]:Alternate Calculation and Calibration Method for   Mizuno 
Report

  
  
  
  
David Roberson <[email protected]> wrote:
  
  
 
  
    
    
I do not get anywhere near     to the 3 to 1 excess power out over input power 
that is     reported.   I am seeing a 1.25 ratio instead of 3.0 or     so.

  


  
Oh, I think you did see the 3 to 1 excess, and the 6:1 as well. Keep   looking! 
Look at Table 1.
  


  
You got 1:1.25 for the water on Oct. 21. Right? I estimate 1:1.35, which   is 
close enough. Now look at Oct. 20, which got a much better ratio with the   
water, 1:2.29. NOW add in the heat captured by the stainless steel reactor   
vessel. The thermal mass of the reactor is larger than that of the water, so I  
 estimated that the vessel captures 60% of the heat, and the water 40%. So, the 
  combined ratio for Oct. 20, for the water and stainless steel, is 1:5.69
  


  
Try applying your method to the Oct. 20 data and see if you get an answer   in 
the same ballpark.
  


  
Additional calibrations are underway. I believe I will soon be able to   
confirm that the reactor vessel is, in fact, capturing 60% of the heat.
  


  
- Jed
  






Reply via email to