I think you interchanged T1 and T2.

Harry

On Tue, Mar 1, 2016 at 2:47 PM, Jed Rothwell <jedrothw...@gmail.com> wrote:
> I printed out the graph and measured the elapsed time between events toward
> the end of the run, starting around hour 14:00. I measured some temperatures
> on the right Y axis. I assume T2 and T4 are correct. I do not trust T1.
> Times are approximate:
>
> Minute 0. T1 and T2 begin rising. T4 stable.
>
> Minute 14. T4 suddenly rises from 110°C up to around 120°C.
>
> Minute 30. T1 falls abruptly. Becomes erratic.
>
> Minute 34. Power off. T2 begins falling. T4 still rising.
>
> Minute 41. T2 begins falling much faster.
>
> Minute 68. T4 reaches a peak temperature of 167°C. This is 34 minutes after
> the power has cut off.
>
> Note that from ~9:00 to 13:15, T4 rose from ~20°C and stabilized at 110°C,
> in response to internal power levels that raised T1 and T2 up to around
> 1100°C. In other words, T4 goes up 90°C, or 1 degree for each 12 degree
> increase in T1 and T2.
>
> Then when T1 and T2 rose only about 100°C more, up to around 1200°C, T4 rose
> proportionally much more than before. It should have gone up ~8°C. Instead,
> it jumped up by around 24°C initially, then it gradually climbed to a peak
> of 167°C, a 57°C increase, even though T1 and T2 had already fallen
> drastically when it peaked.
>
> I cannot make head or tail of this behavior. If there is heat after death,
> it should show up on T2, but I don't see it. T1 is probably damaged, but T2
> seems intact.
>
> - Jed