Jouni Valkonen <[email protected]> wrote:

Mats calibrated the thermometer and indeed he saw the 1 degree offset in
> calibration. Temperature of icy water was measured 1.0°C. Therefore this
> error is known.
>

I discussed that with him. I do not think that was a 1.0°C error. He was
seeing actual 1.0°C temperature variations. This was not an artifact. The
instrument was working fine; he was using ice water with ice cubes in it,
which has a lot of thermal gradients and streams of cold water.

I have done that sort of thing often. You have to use finely crushed ice
slurry or ice water that is poured off, with no ice, warmed up to maybe 5°C.
Plus you have to vigorously stir the water. Otherwise you will get actual 1
or 2°C variations.

An instrument of this type, of this quality will NEVER have an error as
large as 1.0°C unless there is something seriously wrong with it. When there
is something seriously wrong, it will probably display randomly -53.4, +10,
+3 . . . or more likely it will display ERROR. Or nothing; blank. These
things do not break a little bit. They are itty-bitty computers, and like
all computers they tend toward all-or-nothing performance. Also they have
internal boot up testing during turn on, and they tell you when the
thermocouple is busted.

As I mentioned, the Omega HH12B can adjust the OFFSET by +/- 4°C but at
normal temperatures up to the temperature of a roasting turkey I have never
seen the two thermouples separate by anything like that. I assume that
correction is needed at much higher temperatures when you change the
resolution from 0.1 to 1. You never need it at these temperatures.

The Delta T ability depends on precision, not absolute accuracy. In other
words, the TCs might think the temperature is 22.3°C when it is actually
20.4°C. That's inaccurate. But they will both agree later on that it has
risen by 4.8°C, and they will be right. Or if you put the second one in a
colder sample, the Delta T difference they show will be right. If they don't
agree, you can tweak the OFFSET to bring the 2 together, or press REL
(relative) to force the 2 thermocouples to agree and then start your T1-T2
measurement.

Here is the manual:

http://www.omega.com/Manuals/manualpdf/M4613.pdf

- Jed