The error in thermocouple reading is far larger than suggested.  The best place 
to get an accuracy check is when Mats Lewan measured the water output flow from 
the heat exchanger at 18:57 during the October 6 test. He carefully measured 
the water out of the exchanger and calculated the flow rate as .91 
grams/second.  He then assumed it was all vapor which condensed within the 
exchanger and this is the results.

.91 grams/second x 2260 joules/gram = 2056.6 watts.  Water all assumed to be 
vapor which is not being conservative.


(29.8 C - 24.5 C + .8 C) = 6.1 C  Measured at time of water collection with 
correction factor included.

6.1 C x 4.18 joules/C-grams x 178 grams/second = 4538.644 watts According to 
thermocouple data.

The maximum power available is approximately 2056.6 watts versus a reading of 
4538.644 watts.  This is far too much to neglect.  Please explain how the small 
error expected could allow this?

Dave



-----Original Message-----
From: Alan J Fletcher <[email protected]>
To: vortex-l <[email protected]>
Sent: Thu, Nov 10, 2011 7:28 pm
Subject: [Vo]:Oct 6 Heat Exchanger Manifold Thermocouple Placement.


My latest Spice Simulation results, using Bob Higgins' diagram, gives 
n error of  +3.8 %
his is probably LESS than the modelling errors.
http://lenr.qumbu.com/rossi_ecat_oct11_spice.php
Main screenshot :
ttp://lenr.qumbu.com/lenr_spicepics/111109_spice_007.png
Primary Input  (HOT): 100 C water at 15 liters/hour.
econdary Output (COLD) : 30C water at 600 liters/hour (40x ratio)
mbient : 30 C
hermocouple  : 31.96 C
econdary Outlet :  30.78 C (raised slightly from 30C by conduction 
hrough the manifold).
ifference : 1.19 C (or 3.8 %)
This simulation is for Primary Input water at 100C, and is UNCALLIBRATED.
The main unknown in the simulation is the thermal resistance through 
he insulation, and to air where it is uninsulated.
But removing ALL connections to ambient temperature -- which yields 
he highest thermocouple placement error --  gives only a +4.4% error.
here are also possible errors in the placement of the Secondary 
nput thermocouple,  which would give a negative offset, and partly 
ancelling the primary.
I therefore accept the published Heat Exchanger Secondary results as 
alid, and don't plan to study this any further.
It would still be advisable to separate both the thermocouples from 
he heat exchanger assembly through a section of insulated hose.