Dear Andreas, 

we had a similar issue with our MicroMax 007, with a primary water/water 
chiller (ATC K3) in between the generator and a secondary inhouse water 
circuit.  

After a year of reliable service the K3 would shut off occasionally and dump 
the 007 with a cooling error. We found that the solution to the issue was to 
place an additional pump directly in front of the K3. The issue being that the 
latter would temporarily shut off unless provided with a constant water 
pressure, which the water circuit wasn't giving apparently.

This has now been running fine. Another solution would be a water-air chiller 
to cool your 007, but in our case that was not an option due to building specs. 

Hope this helps. I'm happy to provide more details if wanted. 

Han


Han Remaut, PhD
Laboratory of Structural & Molecular Microbiology
VIB / Vrije Universiteit Brussel
Building E4, Pleinlaan 2
1050 Brussel

han.rem...@vib-vub.be
tel. +32-2-629 1923 / +32-499 708050
http://www.vib.be/en/research/scientists/Pages/Han-Remaut-Lab.aspx 


On 24 Jul 2014, at 17:53, Andreas Förster wrote:

> Dear all,
> 
> I have a Rigaku MicroMax 007 HF X-ray generator that is cooled by chilled 
> water supplied rather unreliably through the building infrastructure.  I was 
> wondering what alternatives exist.
> 
> Could other MicroMax 007 users share their experiences with alternative 
> cooling solutions with me?
> 
> Thank you.
> 
> 
> Andreas
> 
> 
> 
> 
> -- 
>                  Andreas Förster
>     Crystallization and X-ray Facility Manager
>           Centre for Structural Biology
>              Imperial College London