Hi Fred,

fine slicing does not alleviate the count RATE limitation of photon counting detectors because fine slicing does not reduce the instantaneous photon flux on the detector when you cross the diffraction maximum. Fine slicing does help if you push the maximum counts per pixel per frame to the counter register limit. On integrating detectors, like CCDs, there is practically no count RATE limit. They do have a charge (~photons) per pixel per frame limit, as well, which is mostly much lower than for the photon counting detectors - about 1/10 even after taking into account the diffraction spot covers many more pixels.

Different from integrating detectors where you only have to watch the overflow, for counting detectors you have to watch for exceeding either the count rate limit or the total count limit. The former is not an easy task because you will see only the count per exposure. Divide that by the exposure time you get the AVERAGE rate, not the peak rate.

The count rate limit, very short exposure time (using high flux) and the 1-pixel point spread function work against each other. Exposure time = 0.01 s, count rate limit 1e6 /sec (Pilatus2), 1-2 pixel per spot => 10-20k counts per spot maximum for the strongest peak.

Dectris has come up with an ingenious hardware and software in the Pilatus3 pushing the rate for reasonable dead time correction to over 1e7 counts/sec so even with 10 ms exposures weak reflections can be well recorded besides strong reflections.

Gerd

On 13.07.2017 15:34, Dyda wrote:
I could be wrong here, but isn't the case that fine slicing is an option
with a CCD and a necessity on a PAD b/c of dead time and/or counter dynamic 
range
issues?

(no current and/or former financial ties to any manufacturer)

Fred
*******************************************************************************
Fred Dyda, Ph.D.                       Phone:301-402-4496
Laboratory of Molecular Biology        Fax: 301-496-0201
DHHS/NIH/NIDDK                         e-mail:fred.d...@nih.gov
Bldg. 5. Room 303
Bethesda, MD 20892-0560      URGENT message e-mail: 2022476...@mms.att.net
Google maps coords: 39.000597, -77.102102
http://www2.niddk.nih.gov/NIDDKLabs/IntramuralFaculty/DydaFred
*******************************************************************************

Reply via email to