Thanks for the info. I ran all the possible combinations of phase encoding 
directions and the invert option, and their effects on the DICOM header. This 
was actually very useful for us to understand the data. It seems that the CSA 
header field #20 (PhaseEncodingDirectionPositive) does take into account the 
prescribed direction and the invert option. So in this case, I assume I should 
use the SE image with the DICOM header of PhaseEncodingDirectionPositive=1 as 
the SpinEchoPhaseEncodePositive image in the HCP setup script?

From: Harms, Michael [mailto:mha...@wustl.edu]
Sent: Monday, December 15, 2014 5:41 PM
To: Book, Gregory; Harms, Michael; Glasser, Matthew; 
hcp-users@humanconnectome.org
Subject: Re: [HCP-Users] spin echo phase encoding direction


Hi Greg,
Ultimately, getting all the signs right is a complicated combination of (1) how 
the manufacturer implements its PE gradient blips, (2) the DICOM to NIFTI 
converter that you use (and whether you 'reorient' the data, e.g., using 
'fslreorient2std'), (3) your settings for "Phase enc. dir." and "Invert RO/PE 
Polarity and (4) the plane/oriention that the EPI is acquired in.  It is very 
dangerous for us to provide you a "rule", as it is very difficult to anticipate 
all usage cases.  That's why I stressed the importance of confirming that you 
have all the conventions correct in your own data (by explicitly inverting the 
value of the "PhaseEncodeList" parameter in the pipeline, and seeing if the 
result indeed gets worse).

That said, as a sanity check for you, on a Siemen's scanner, acquiring 
transverse slices, using 'dcm2nii' as your NIFTI converter, with transverse 
slices, I would expect that you find the following:
If phase encoding direction=COL, and directionPositive=1, then PEdir="-y"
If phase encoding direction=COL, and directionPositive=0, then PEdir="+y"

As Matt noted, those "PEdir" (i.e., PhaseEncodeList values) themselves are 
dependent on a certain convention for the SpinEchoFieldMaps, which I believe is 
outlined in the comments of the HCP pipeline scripts.

cheers,
-MH

--
Michael Harms, Ph.D.
-----------------------------------------------------------
Conte Center for the Neuroscience of Mental Disorders
Washington University School of Medicine
Department of Psychiatry, Box 8134
660 South Euclid Ave.  Tel: 314-747-6173
St. Louis, MO  63110  Email: mha...@wustl.edu<mailto:mha...@wustl.edu>

From: <Book>, Gregory 
<gregory.b...@hhchealth.org<mailto:gregory.b...@hhchealth.org>>
Date: Monday, December 15, 2014 4:17 PM
To: "Harms, Michael" <mha...@wustl.edu<mailto:mha...@wustl.edu>>, Matt Glasser 
<glass...@wusm.wustl.edu<mailto:glass...@wusm.wustl.edu>>, 
"hcp-users@humanconnectome.org<mailto:hcp-users@humanconnectome.org>" 
<hcp-users@humanconnectome.org<mailto:hcp-users@humanconnectome.org>>
Subject: Re: [HCP-Users] spin echo phase encoding direction

Ok, good to know what that means as to how the data is collected relative to 
the label.

So with the original example of a console entered A>>P, what does that mean if 
the PhaseEncodingDirectionPositive is 1 or 0?
-G
________________________________
From: Harms, Michael [mha...@wustl.edu<mailto:mha...@wustl.edu>]
Sent: Monday, December 15, 2014 4:10 PM
To: Book, Gregory; Glasser, Matthew; Harms, Michael; 
hcp-users@humanconnectome.org<mailto:hcp-users@humanconnectome.org>
Subject: Re: [HCP-Users] spin echo phase encoding direction

Hi,
If you had "A>>P" with "Invert RO/PE Polarity" checked, then that would be the 
same as if you had set "P>>A" and "Invert RO/PE Polarity" was left unchecked.

The *dInPlaneRot field does NOT include the impact of the "Invert RO/PE 
Polarity" flag.  Sorry, I directed you to the wrong shadow field previously.  
"PhaseEncodingDirectionPositive" is (according to my notes) subfield #20 of 
(0029, 1010).  If it doesn't have a value, that means it value was 0 for that 
scan.  (Its value is either 0 or 1).

cheers,
-MH

--
Michael Harms, Ph.D.
-----------------------------------------------------------
Conte Center for the Neuroscience of Mental Disorders
Washington University School of Medicine
Department of Psychiatry, Box 8134
660 South Euclid Ave.  Tel: 314-747-6173
St. Louis, MO  63110  Email: mha...@wustl.edu<mailto:mha...@wustl.edu>

From: <Book>, Gregory 
<gregory.b...@hhchealth.org<mailto:gregory.b...@hhchealth.org>>
Date: Monday, December 15, 2014 2:59 PM
To: Matt Glasser <glass...@wusm.wustl.edu<mailto:glass...@wusm.wustl.edu>>, 
"Harms, Michael" <mha...@wustl.edu<mailto:mha...@wustl.edu>>, 
"hcp-users@humanconnectome.org<mailto:hcp-users@humanconnectome.org>" 
<hcp-users@humanconnectome.org<mailto:hcp-users@humanconnectome.org>>
Subject: Re: [HCP-Users] spin echo phase encoding direction

I'm slightly more confused now. So, using the CMRR sequences, if I set the 
phase encoding direction on my spin echo to A>>P, and "Invert RO/PE Polarity" 
option is checked... a) what is the actual phase direction of the data b) what 
is recorded in the DICOM header in the *.dInPlaneRot field?



From: Glasser, Matthew [mailto:glass...@wusm.wustl.edu]
Sent: Monday, December 15, 2014 2:33 PM
To: Book, Gregory; Harms, Michael; 
hcp-users@humanconnectome.org<mailto:hcp-users@humanconnectome.org>
Subject: Re: [HCP-Users] spin echo phase encoding direction

If your phase encoding directions are inverted, I would recommend swapping the 
SpinEchoPositive and NegativeImage values.  This would make the computed field 
map have the same sign as a typical field map and then the PhaseEncodeList 
would not have to change.

Peace,

Matt.

From: <Book>, Gregory 
<gregory.b...@hhchealth.org<mailto:gregory.b...@hhchealth.org>>
Date: Monday, December 15, 2014 at 12:16 PM
To: "Harms, Michael" <mha...@wustl.edu<mailto:mha...@wustl.edu>>, 
"hcp-users@humanconnectome.org<mailto:hcp-users@humanconnectome.org>" 
<hcp-users@humanconnectome.org<mailto:hcp-users@humanconnectome.org>>
Subject: Re: [HCP-Users] spin echo phase encoding direction

Would that be the *.dInPlaneRot value for each slice in the CSA header?

From: Harms, Michael [mailto:mha...@wustl.edu]
Sent: Monday, December 15, 2014 10:11 AM
To: Book, Gregory; 
hcp-users@humanconnectome.org<mailto:hcp-users@humanconnectome.org>
Subject: Re: [HCP-Users] spin echo phase encoding direction


The value of "PhaseEncodingDirectionPositive" in the shadow field (0029,1020) 
reflects the combination of both "Phase enc. dir." on the Routine tab (e.g., 
"A>>P" vs. "P>>A") AND the setting of "Invert RO/PE Polarity" on the 
Sequence/Special tab.

As long as you have one of each polarity, you can use the spin echo field map 
correction.   I highly recommend that you purposely try both possible 
assignments, and visually confirm which one is giving the correct results.  
That's because the correct setting for PhaseEncodeList (e.g., "y-" vs. "y+") 
can in theory even be dependent on which DICOM to NIFTI converter you use.  
(e.g., if one converter returns a +LAS orientation, while another returns a 
+LPS orientation, then the proper settings for PhaseEncodeList may be swapped).

cheers,
-MH

--
Michael Harms, Ph.D.
-----------------------------------------------------------
Conte Center for the Neuroscience of Mental Disorders
Washington University School of Medicine
Department of Psychiatry, Box 8134
660 South Euclid Ave. Tel: 314-747-6173
St. Louis, MO  63110 Email: mha...@wustl.edu<mailto:mha...@wustl.edu>

From: <Book>, Gregory 
<gregory.b...@hhchealth.org<mailto:gregory.b...@hhchealth.org>>
Date: Monday, December 15, 2014 8:38 AM
To: "hcp-users@humanconnectome.org<mailto:hcp-users@humanconnectome.org>" 
<hcp-users@humanconnectome.org<mailto:hcp-users@humanconnectome.org>>
Subject: [HCP-Users] spin echo phase encoding direction

I have another question about the spin echo images, hopefully this will be my 
last question!

It turns out that we've been collecting our spin echo images with the "Invert 
RO/PE Polarity" option selected, which apparently inverts the phase encoding 
direction 180deg, but does not record this to the DICOM header. We've been 
collecting the EPI images without this inversion.
So, for our data:
EPI is A>>P
"A>>P" SE is actually P>>A
"P>>A" SE is actually A>>P

Would we be able to use the spin echo correction if we simply switched the 
positive and negative images? So, would this work for our data?
TaskList="task1"
PhaseEncodeList="y-"
...
SpinEchoPositiveImage="the P>>A labeled image (actually acquired A>>P)"
SpinEchoNegativeImage="the A>>P labeled image (actually acquired P>>A)"

-G

_________________________________________________
Gregory Book
Senior Technology Manager
Olin Neuropsychiatry Research Center, Institute of Living, Hartford Hospital
200 Retreat Avenue
Hartford, CT 06106
Tel: 860-545-7267 Fax: 860-545-7797
gregory.b...@hhchealth.org<mailto:gregory.b...@hhchealth.org>
http://nidb.sourceforge.net


This e-mail message, including any attachments, is for the sole use of the 
intended recipient(s) and may contain confidential and privileged information. 
Any unauthorized review, use, disclosure, or distribution is prohibited. If you 
are not the intended recipient, or an employee or agent responsible for 
delivering the message to the intended recipient, please contact the sender by 
reply e-mail and destroy all copies of the original message, including any 
attachments.

_______________________________________________
HCP-Users mailing list
HCP-Users@humanconnectome.org<mailto:HCP-Users@humanconnectome.org>
http://lists.humanconnectome.org/mailman/listinfo/hcp-users

________________________________
The materials in this message are private and may contain Protected Healthcare 
Information or other information of a sensitive nature. If you are not the 
intended recipient, be advised that any unauthorized use, disclosure, copying 
or the taking of any action in reliance on the contents of this information is 
strictly prohibited. If you have received this email in error, please 
immediately notify the sender via telephone or return mail.

_______________________________________________
HCP-Users mailing list
HCP-Users@humanconnectome.org<mailto:HCP-Users@humanconnectome.org>
http://lists.humanconnectome.org/mailman/listinfo/hcp-users

________________________________
The materials in this message are private and may contain Protected Healthcare 
Information or other information of a sensitive nature. If you are not the 
intended recipient, be advised that any unauthorized use, disclosure, copying 
or the taking of any action in reliance on the contents of this information is 
strictly prohibited. If you have received this email in error, please 
immediately notify the sender via telephone or return mail.

_______________________________________________
HCP-Users mailing list
HCP-Users@humanconnectome.org<mailto:HCP-Users@humanconnectome.org>
http://lists.humanconnectome.org/mailman/listinfo/hcp-users

________________________________
The materials in this message are private and may contain Protected Healthcare 
Information or other information of a sensitive nature. If you are not the 
intended recipient, be advised that any unauthorized use, disclosure, copying 
or the taking of any action in reliance on the contents of this information is 
strictly prohibited. If you have received this email in error, please 
immediately notify the sender via telephone or return mail.

_______________________________________________
HCP-Users mailing list
HCP-Users@humanconnectome.org<mailto:HCP-Users@humanconnectome.org>
http://lists.humanconnectome.org/mailman/listinfo/hcp-users

________________________________
The materials in this message are private and may contain Protected Healthcare 
Information or other information of a sensitive nature. If you are not the 
intended recipient, be advised that any unauthorized use, disclosure, copying 
or the taking of any action in reliance on the contents of this information is 
strictly prohibited. If you have received this email in error, please 
immediately notify the sender via telephone or return mail.

_______________________________________________
HCP-Users mailing list
HCP-Users@humanconnectome.org
http://lists.humanconnectome.org/mailman/listinfo/hcp-users

Reply via email to