2015-02-16 Thread Glasser, Matthew



Also to Steve: if these are used after ICAFIX, don’t they need to have the 24 motion parameters and noise ICA component timeseries confounds regressed out as well, before being used on the cleaned data?


Peace,


Matt.




From: Harms, Michael mha...@wustl.edu
Date: Monday, February 16, 2015 at 10:42 AM
To: Matt Glasser glass...@wusm.wustl.edu, Stephen Smith st...@fmrib.ox.ac.uk, Miriam Klein-Flügge miriam.klein-flu...@psy.ox.ac.uk
Cc: hcp-users@humanconnectome.org hcp-users@humanconnectome.org
Subject: Re: [HCP-Users] Incorporating physiological monitoring data









Not only that, but just because a given physiological trace exists, doesn't necessarily mean that it is a *good* trace. There is going to be considerable variability in the quality of the physiological measurements, which presents a challenge in using
 them in a large scale study. I'm sure that Greg will comment more when he has a chance.


cheers,
-MH




--
Michael Harms, Ph.D.

---
Conte Center for the Neuroscience of Mental Disorders
Washington University School of Medicine
Department of Psychiatry, Box 8134
660 South Euclid Ave. 
Tel: 314-747-6173
St. Louis, MO 63110 
Email: mha...@wustl.edu







From: Glasser, Matt Glasser glass...@wusm.wustl.edu
Date: Monday, February 16, 2015 10:10 AM
To: Stephen Smith st...@fmrib.ox.ac.uk, Miriam Klein-Flügge miriam.klein-flu...@psy.ox.ac.uk
Cc: hcp-users@humanconnectome.org hcp-users@humanconnectome.org
Subject: Re: [HCP-Users] Incorporating physiological monitoring data





Note that a major reason we didn’t use these physiological confound regressors was they don’t exist for every subject, so be sure to select a subset of subjects that have them. We’d also be interested to know if you found they were helpful.


Peace,


Matt.




From: Stephen Smith st...@fmrib.ox.ac.uk
Date: Monday, February 16, 2015 at 8:05 AM
To: Miriam Klein-Flügge miriam.klein-flu...@psy.ox.ac.uk
Cc: hcp-users@humanconnectome.org hcp-users@humanconnectome.org
Subject: Re: [HCP-Users] Incorporating physiological monitoring data





Hi - I think you would probably be best off taking the FIX-cleaned version of the data, and then apply additional confound regressors if they will help. Don't forget to apply the same highpass filter (to those regressors) that was applied already
 in the data preproc, before you use them.


Cheers.







On 16 Feb 2015, at 13:35, Miriam Klein-Flügge miriam.klein-flu...@psy.ox.ac.uk wrote:



Dear all,
Is it correct that until now the physiological monitoring data is not made use of in the preprocessing of the rfMRI data? I would like to correct for cardiac and respiratory signals and wondered how to best do that. I can see that FIX-denoising
 probably takes care of this type of noise but I am particularly interested in looking at the brain stem and in my experience, including physiological regressors in the preprocessing makes a big difference there.
Would you recommend using the spatially (minimally) pre-processed rfMRI data, performing the high-pass filtering on it myself and then incorporating the physiological regressors at that stage, or is there a better stage at which to do it?
 Also, is there a standard procedure for regressing out the physiological regressors that you can recommend?
Many thanks in advance!
Kind regards,
Miriam

-- 
Miriam Klein-Flügge
Sir Henry Wellcome Postdoctoral Fellow
Department of Experimental Psychology
University of Oxford






___
HCP-Users mailing list
HCP-Users@humanconnectome.org
http://lists.humanconnectome.org/mailman/listinfo/hcp-users
















---
Stephen M. Smith, Professor of Biomedical Engineering
Associate Director, Oxford University FMRIB Centre

FMRIB, JR Hospital, Headington, Oxford OX3 9DU, UK
44 (0) 1865 222726 (fax 222717)
st...@fmrib.ox.ac.uk http://www.fmrib.ox.ac.uk/~steve
---



Stop the culturaldestructionof Tibet
















___
HCP-Users mailing list
HCP-Users@humanconnectome.org
http://lists.humanconnectome.org/mailman/listinfo/hcp-users









The materials in this message are private and may contain Protected Healthcare Information or other information of a sensitive nature. If you are not the intended
 recipient, be advised that any unauthorized use, disclosure, copying or the taking of any action in reliance on the contents of this information is strictly prohibited. If you have received this email in error, please immediately notify the sender via telephone
 or return mail.
___
HCP-Users mailing list
HCP-Users@humanconnectome.org
http://lists.humanconnectome.org/mailman/listinfo/hcp-users











The materials in this message are private and may contain Protected Healthcare

Re: [HCP-Users] Incorporating physiological monitoring data

2015-02-16 Thread Glasser, Matthew



Note that a major reason we didn’t use these physiological confound regressors was they don’t exist for every subject, so be sure to select a subset of subjects that have them. We’d also be interested to know if you found they were helpful.


Peace,


Matt.




From: Stephen Smith st...@fmrib.ox.ac.uk
Date: Monday, February 16, 2015 at 8:05 AM
To: Miriam Klein-Flügge miriam.klein-flu...@psy.ox.ac.uk
Cc: hcp-users@humanconnectome.org hcp-users@humanconnectome.org
Subject: Re: [HCP-Users] Incorporating physiological monitoring data





Hi - I think you would probably be best off taking the FIX-cleaned version of the data, and then apply additional confound regressors if they will help. Don't forget to apply the same highpass filter (to those regressors) that was applied already
 in the data preproc, before you use them.


Cheers.







On 16 Feb 2015, at 13:35, Miriam Klein-Flügge miriam.klein-flu...@psy.ox.ac.uk wrote:



Dear all,
Is it correct that until now the physiological monitoring data is not made use of in the preprocessing of the rfMRI data? I would like to correct for cardiac and respiratory signals and wondered how to best do that. I can see that FIX-denoising
 probably takes care of this type of noise but I am particularly interested in looking at the brain stem and in my experience, including physiological regressors in the preprocessing makes a big difference there.
Would you recommend using the spatially (minimally) pre-processed rfMRI data, performing the high-pass filtering on it myself and then incorporating the physiological regressors at that stage, or is there a better stage at which to do it?
 Also, is there a standard procedure for regressing out the physiological regressors that you can recommend?
Many thanks in advance!
Kind regards,
Miriam

-- 
Miriam Klein-Flügge
Sir Henry Wellcome Postdoctoral Fellow
Department of Experimental Psychology
University of Oxford






___
HCP-Users mailing list
HCP-Users@humanconnectome.org
http://lists.humanconnectome.org/mailman/listinfo/hcp-users
















---
Stephen M. Smith, Professor of Biomedical Engineering
Associate Director, Oxford University FMRIB Centre

FMRIB, JR Hospital, Headington, Oxford OX3 9DU, UK
44 (0) 1865 222726 (fax 222717)
st...@fmrib.ox.ac.uk http://www.fmrib.ox.ac.uk/~steve
---



Stop the culturaldestructionof Tibet
















___
HCP-Users mailing list
HCP-Users@humanconnectome.org
http://lists.humanconnectome.org/mailman/listinfo/hcp-users









The materials in this message are private and may contain Protected Healthcare Information or other information of a sensitive nature. If you are not the intended
 recipient, be advised that any unauthorized use, disclosure, copying or the taking of any action in reliance on the contents of this information is strictly prohibited. If you have received this email in error, please immediately notify the sender via telephone
 or return mail.
___HCP-Users mailing listHCP-Users@humanconnectome.orghttp://lists.humanconnectome.org/mailman/listinfo/hcp-users




Re: [HCP-Users] Incorporating physiological monitoring data

2015-02-16 Thread Stephen Smith
urgh - yes you're right.



 On 16 Feb 2015, at 16:47, Glasser, Matthew glass...@wusm.wustl.edu wrote:
 
 Also to Steve: if these are used after ICA+FIX, don’t they need to have the 
 24 motion parameters and noise ICA component timeseries confounds regressed 
 out as well, before being used on the cleaned data?
 
 Peace,
 
 Matt.
 
 From: Harms, Michael mha...@wustl.edu mailto:mha...@wustl.edu
 Date: Monday, February 16, 2015 at 10:42 AM
 To: Matt Glasser glass...@wusm.wustl.edu mailto:glass...@wusm.wustl.edu, 
 Stephen Smith st...@fmrib.ox.ac.uk mailto:st...@fmrib.ox.ac.uk, Miriam 
 Klein-Flügge miriam.klein-flu...@psy.ox.ac.uk 
 mailto:miriam.klein-flu...@psy.ox.ac.uk
 Cc: hcp-users@humanconnectome.org mailto:hcp-users@humanconnectome.org 
 hcp-users@humanconnectome.org mailto:hcp-users@humanconnectome.org
 Subject: Re: [HCP-Users] Incorporating physiological monitoring data
 
 
 Not only that, but just because a given physiological trace exists, doesn't 
 necessarily mean that it is a *good* trace.  There is going to be 
 considerable variability in the quality of the physiological measurements, 
 which presents a challenge in using them in a large scale study.  I'm sure 
 that Greg will comment more when he has a chance.
 
 cheers,
 -MH
 
 -- 
 Michael Harms, Ph.D.
 ---
 Conte Center for the Neuroscience of Mental Disorders
 Washington University School of Medicine
 Department of Psychiatry, Box 8134
 660 South Euclid Ave.  Tel: 314-747-6173
 St. Louis, MO  63110  Email: mha...@wustl.edu mailto:mha...@wustl.edu
 
 From: Glasser, Matt Glasser glass...@wusm.wustl.edu 
 mailto:glass...@wusm.wustl.edu
 Date: Monday, February 16, 2015 10:10 AM
 To: Stephen Smith st...@fmrib.ox.ac.uk mailto:st...@fmrib.ox.ac.uk, 
 Miriam Klein-Flügge miriam.klein-flu...@psy.ox.ac.uk 
 mailto:miriam.klein-flu...@psy.ox.ac.uk
 Cc: hcp-users@humanconnectome.org mailto:hcp-users@humanconnectome.org 
 hcp-users@humanconnectome.org mailto:hcp-users@humanconnectome.org
 Subject: Re: [HCP-Users] Incorporating physiological monitoring data
 
 Note that a major reason we didn’t use these physiological confound 
 regressors was they don’t exist for every subject, so be sure to select a 
 subset of subjects that have them.  We’d also be interested to know if you 
 found they were helpful.
 
 Peace,
 
 Matt.
 
 From: Stephen Smith st...@fmrib.ox.ac.uk mailto:st...@fmrib.ox.ac.uk
 Date: Monday, February 16, 2015 at 8:05 AM
 To: Miriam Klein-Flügge miriam.klein-flu...@psy.ox.ac.uk 
 mailto:miriam.klein-flu...@psy.ox.ac.uk
 Cc: hcp-users@humanconnectome.org mailto:hcp-users@humanconnectome.org 
 hcp-users@humanconnectome.org mailto:hcp-users@humanconnectome.org
 Subject: Re: [HCP-Users] Incorporating physiological monitoring data
 
 Hi - I think you would probably be best off taking the FIX-cleaned version of 
 the data, and then apply additional confound regressors if they will help.  
 Don't forget to apply the same highpass filter (to those regressors) that was 
 applied already in the data preproc, before you use them.
 
 Cheers.
 
 
 
 On 16 Feb 2015, at 13:35, Miriam Klein-Flügge 
 miriam.klein-flu...@psy.ox.ac.uk mailto:miriam.klein-flu...@psy.ox.ac.uk 
 wrote:
 
 Dear all,
 Is it correct that until now the physiological monitoring data is not made 
 use of in the preprocessing of the rfMRI data? I would like to correct for 
 cardiac and respiratory signals and wondered how to best do that. I can see 
 that FIX-denoising probably takes care of this type of noise but I am 
 particularly interested in looking at the brain stem and in my experience, 
 including physiological regressors in the preprocessing makes a big 
 difference there.
 Would you recommend using the spatially (minimally) pre-processed rfMRI 
 data, performing the high-pass filtering on it myself and then incorporating 
 the physiological regressors at that stage, or is there a better stage at 
 which to do it? Also, is there a standard procedure for regressing out the 
 physiological regressors that you can recommend?
 Many thanks in advance!
 Kind regards,
 Miriam
 -- 
 Miriam Klein-Flügge
 Sir Henry Wellcome Postdoctoral Fellow
 Department of Experimental Psychology
 University of Oxford
 
 ___
 HCP-Users mailing list
 HCP-Users@humanconnectome.org mailto:HCP-Users@humanconnectome.org
 http://lists.humanconnectome.org/mailman/listinfo/hcp-users 
 http://lists.humanconnectome.org/mailman/listinfo/hcp-users
 
 ---
 Stephen M. Smith, Professor of Biomedical Engineering
 Associate Director,  Oxford University FMRIB Centre
 
 FMRIB, JR Hospital, Headington, Oxford  OX3 9DU, UK
 +44 (0) 1865 222726  (fax 222717)
 st...@fmrib.ox.ac.uk mailto:st...@fmrib.ox.ac.uk
 http://www.fmrib.ox.ac.uk/~steve http://www.fmrib.ox.ac.uk/~steve

Re: [HCP-Users] Incorporating physiological monitoring data

2015-02-16 Thread Harms, Michael







Not only that, but just because a given physiological trace exists, doesn't necessarily mean that it is a *good* trace. There is going to be considerable variability in the quality of the physiological measurements, which presents a challenge in using
 them in a large scale study. I'm sure that Greg will comment more when he has a chance.


cheers,
-MH




--
Michael Harms, Ph.D.

---
Conte Center for the Neuroscience of Mental Disorders
Washington University School of Medicine
Department of Psychiatry, Box 8134
660 South Euclid Ave. 
Tel: 314-747-6173
St. Louis, MO 63110 
Email: mha...@wustl.edu







From: Glasser, Matt Glasser glass...@wusm.wustl.edu
Date: Monday, February 16, 2015 10:10 AM
To: Stephen Smith st...@fmrib.ox.ac.uk, Miriam Klein-Flügge miriam.klein-flu...@psy.ox.ac.uk
Cc: hcp-users@humanconnectome.org hcp-users@humanconnectome.org
Subject: Re: [HCP-Users] Incorporating physiological monitoring data





Note that a major reason we didn’t use these physiological confound regressors was they don’t exist for every subject, so be sure to select a subset of subjects that have them. We’d also be interested to know if you found they were helpful.


Peace,


Matt.




From: Stephen Smith st...@fmrib.ox.ac.uk
Date: Monday, February 16, 2015 at 8:05 AM
To: Miriam Klein-Flügge miriam.klein-flu...@psy.ox.ac.uk
Cc: hcp-users@humanconnectome.org hcp-users@humanconnectome.org
Subject: Re: [HCP-Users] Incorporating physiological monitoring data





Hi - I think you would probably be best off taking the FIX-cleaned version of the data, and then apply additional confound regressors if they will help. Don't forget to apply the same highpass filter (to those regressors) that was applied already
 in the data preproc, before you use them.


Cheers.







On 16 Feb 2015, at 13:35, Miriam Klein-Flügge miriam.klein-flu...@psy.ox.ac.uk wrote:



Dear all,
Is it correct that until now the physiological monitoring data is not made use of in the preprocessing of the rfMRI data? I would like to correct for cardiac and respiratory signals and wondered how to best do that. I can see that FIX-denoising
 probably takes care of this type of noise but I am particularly interested in looking at the brain stem and in my experience, including physiological regressors in the preprocessing makes a big difference there.
Would you recommend using the spatially (minimally) pre-processed rfMRI data, performing the high-pass filtering on it myself and then incorporating the physiological regressors at that stage, or is there a better stage at which to do it?
 Also, is there a standard procedure for regressing out the physiological regressors that you can recommend?
Many thanks in advance!
Kind regards,
Miriam

-- 
Miriam Klein-Flügge
Sir Henry Wellcome Postdoctoral Fellow
Department of Experimental Psychology
University of Oxford






___
HCP-Users mailing list
HCP-Users@humanconnectome.org
http://lists.humanconnectome.org/mailman/listinfo/hcp-users
















---
Stephen M. Smith, Professor of Biomedical Engineering
Associate Director, Oxford University FMRIB Centre

FMRIB, JR Hospital, Headington, Oxford OX3 9DU, UK
44 (0) 1865 222726 (fax 222717)
st...@fmrib.ox.ac.uk http://www.fmrib.ox.ac.uk/~steve
---



Stop the culturaldestructionof Tibet
















___
HCP-Users mailing list
HCP-Users@humanconnectome.org
http://lists.humanconnectome.org/mailman/listinfo/hcp-users









The materials in this message are private and may contain Protected Healthcare Information or other information of a sensitive nature. If you are not the intended
 recipient, be advised that any unauthorized use, disclosure, copying or the taking of any action in reliance on the contents of this information is strictly prohibited. If you have received this email in error, please immediately notify the sender via telephone
 or return mail.
___
HCP-Users mailing list
HCP-Users@humanconnectome.org
http://lists.humanconnectome.org/mailman/listinfo/hcp-users









The materials in this message are private and may contain Protected Healthcare Information or other information of a sensitive nature. If you are not the intended
 recipient, be advised that any unauthorized use, disclosure, copying or the taking of any action in reliance on the contents of this information is strictly prohibited. If you have received this email in error, please immediately notify the sender via telephone
 or return mail.
___HCP-Users mailing listHCP-Users@humanconnectome.orghttp://lists.humanconnectome.org/mailman/listinfo/hcp-users




Re: [HCP-Users] Incorporating physiological monitoring data

2015-02-16 Thread Stephen Smith
Hi - I think you would probably be best off taking the FIX-cleaned version of 
the data, and then apply additional confound regressors if they will help.  
Don't forget to apply the same highpass filter (to those regressors) that was 
applied already in the data preproc, before you use them.

Cheers.



 On 16 Feb 2015, at 13:35, Miriam Klein-Flügge 
 miriam.klein-flu...@psy.ox.ac.uk wrote:
 
 Dear all,
 
 Is it correct that until now the physiological monitoring data is not made 
 use of in the preprocessing of the rfMRI data? I would like to correct for 
 cardiac and respiratory signals and wondered how to best do that. I can see 
 that FIX-denoising probably takes care of this type of noise but I am 
 particularly interested in looking at the brain stem and in my experience, 
 including physiological regressors in the preprocessing makes a big 
 difference there.
 
 Would you recommend using the spatially (minimally) pre-processed rfMRI data, 
 performing the high-pass filtering on it myself and then incorporating the 
 physiological regressors at that stage, or is there a better stage at which 
 to do it? Also, is there a standard procedure for regressing out the 
 physiological regressors that you can recommend?
 
 Many thanks in advance!
 
 Kind regards,
 
 Miriam
 
 -- 
 Miriam Klein-Flügge
 Sir Henry Wellcome Postdoctoral Fellow
 Department of Experimental Psychology
 University of Oxford
 
 
 ___
 HCP-Users mailing list
 HCP-Users@humanconnectome.org
 http://lists.humanconnectome.org/mailman/listinfo/hcp-users
 


---
Stephen M. Smith, Professor of Biomedical Engineering
Associate Director,  Oxford University FMRIB Centre

FMRIB, JR Hospital, Headington, Oxford  OX3 9DU, UK