Hi Alex,

May be  we worked with the same furnace: Inel? I had also the same problem and 
solved it by making a new sample holder with a smaller thermocouple placed on 
the side of the sample holder.

Patrick 

Patrick Weisbecker
LCTS - Pessac

 
- Message d'origine 
De : Yokochi, Alexandre alexandre.yoko...@oregonstate.edu
À : Rietveld list rietveld_l@ill.fr
Envoyé le : Jeu 15 Octobre 2009, 21 h 33 min 31 s
Objet : RE: Re : HT XRD calibration

Hi all,

I've never used one of these Anton Paar furnaces, but with some of my 
instruments (externally heated, not Pt resistive foil heated) I could actually 
use the thermocouple as the sample mount.  I also had an instrument that had a 
thermocouple in contact with the powder sample, which due to thermal expansion 
mismatch would upon heating expand and split the sample surface giving rise to 
ugly split peaks so be careful... (you can imagine that implementation lasted 
as many samples as it took me to figure out what the peak splitting problem 
was).

Alex Y


Dr. Alexandre (Alex) F. T. Yokochi
Assistant Professor
Department of Chemical Engineering
Oregon State University
Corvallis, OR 97331-2702

Voice:  (541) 737-9357
Fax:  (541) 737-4600
http://oregonstate.edu/~yokochia

 -Original Message-
 From: Payzant, E. Andrew [mailto:payza...@ornl.gov]
 Sent: Thursday, October 15, 2009 6:27 AM
 To: Rietveld list
 Subject: Re: Re : HT XRD calibration
 
 Patrick,
 
 That is a good idea. Note that to actually get rid of the uncertainty you
 need to be certain that the thermocouple is either reading the temperature
 of the actual sample volume contributing to the diffraction pattern, or
 that it is reading the temperature of a location where the temperature is
 exactly the same as the sample volume.
 
 Andrew
 
 
 On 10/14/09 11:25 AM, Patrick Weisbecker weis...@yahoo.fr wrote:
 
 Hi,
 
 One way to get rid of the temperature uncertainty is to introduce a second
 thermocouple inside the furnace and to fix it as close as possible from
 the sample. If the sample is a bulk sample:Drill a hole inside the sample
 and put the thermocouple inside the hole and if the the sample is made of
 powder put it inside the powder.
 The are some miniature thermocouple (0.25-0.5mm diameter), mineral
 insulated which are unlikely to react with the samples.
 
 Patrick
 
 
 
 
 - Message d'origine 
 De : Payzant, E. Andrew payza...@ornl.gov
 À : Mikko Heikkilä mikko.j.heikk...@helsinki.fi
 Cc : Rietveld list rietveld_l@ill.fr
 Envoyé le : Mer 14 Octobre 2009, 16 h 39 min 58 s
 Objet : Re: HT XRD calibration
 
 All,
 
 The two calibration standard method was described by Andy Drews (Ford
 Research Lab) in Advances in X-ray Analysis (the proceedings of the Denver
 X-ray Conference), vol. 44, page 44 (2001).
 
 Scott Misture (Alfred Univ), Stephen Skinner (Imperial College), Christian
 Resch (Anton Paar), and I ran a workshop on non-ambient diffraction at the
 Denver X-ray Conference in 2008, where we discussed many of these issues.
 There was a lively discussion with the attendees regarding various
 experiences with HTXRD, and the difficulty in abtaining the right
 answer. Too often we are so focused on the precision and accuracy of our
 refinements that we neglect the temperature issue, to our peril.
 
 One issue with the HT16 hot stage is that you will likely need to
 recalibrate with every new heater strip - don't rely on them behaving the
 same. Even for a particular strip, use of an internal standard is best if
 you can use one that does not react with either the strip or the sample.
 If you push this furnace to temperatures above 1200degC or so, the
 platinum will creep and the tensioning will change, so that the
 height/temperature calibration may subsequently be off.
 
 I agree fully with the suggestion to use parallel beam optics to eliminate
 the sample height error. You do lose some angular resolution of course, at
 least on a laboratory diffractometer, but you take one variable (sample
 height) out of the refinement.
 
 The other BIG problem with the HT16 (and many other non-ambient stages) is
 knowing the temperature of the SAMPLE (as opposed to the strip). Powders
 have hopelessly poor thermal conductivity, and the gradients in these
 furnaces can be substantial. We rarely measure the temperature of the
 sample directly, and this is another reason why internal standards are
 very useful - you can use either a known CTE to estimate the delta T over
 the full temperature range, or a known phase transformation to determine
 delta T at a single point. I personally prefer the former approach myself.
 Note that a solid-solid transformation may give a different result than a
 melting point standard, since the liquid will change the thermal
 conductivity between the strip and the sample. You can find more
 discussion of these issues in Mark Rodriguez's chapter in the book
 Industrial Applications of X-ray Diffraction, or mine

Re: Re : HT XRD calibration

2009-10-15 Thread Payzant, E. Andrew
Patrick,

That is a good idea. Note that to actually get rid of the uncertainty you need 
to be certain that the thermocouple is either reading the temperature of the 
actual sample volume contributing to the diffraction pattern, or that it is 
reading the temperature of a location where the temperature is exactly the same 
as the sample volume.

Andrew


On 10/14/09 11:25 AM, Patrick Weisbecker weis...@yahoo.fr wrote:

Hi,

One way to get rid of the temperature uncertainty is to introduce a second 
thermocouple inside the furnace and to fix it as close as possible from the 
sample. If the sample is a bulk sample:Drill a hole inside the sample and put 
the thermocouple inside the hole and if the the sample is made of powder put it 
inside the powder.
The are some miniature thermocouple (0.25-0.5mm diameter), mineral insulated 
which are unlikely to react with the samples.

Patrick




- Message d'origine 
De : Payzant, E. Andrew payza...@ornl.gov
À : Mikko Heikkilä mikko.j.heikk...@helsinki.fi
Cc : Rietveld list rietveld_l@ill.fr
Envoyé le : Mer 14 Octobre 2009, 16 h 39 min 58 s
Objet : Re: HT XRD calibration

All,

The two calibration standard method was described by Andy Drews (Ford Research 
Lab) in Advances in X-ray Analysis (the proceedings of the Denver X-ray 
Conference), vol. 44, page 44 (2001).

Scott Misture (Alfred Univ), Stephen Skinner (Imperial College), Christian 
Resch (Anton Paar), and I ran a workshop on non-ambient diffraction at the 
Denver X-ray Conference in 2008, where we discussed many of these issues. There 
was a lively discussion with the attendees regarding various experiences with 
HTXRD, and the difficulty in abtaining the right answer. Too often we are so 
focused on the precision and accuracy of our refinements that we neglect the 
temperature issue, to our peril.

One issue with the HT16 hot stage is that you will likely need to recalibrate 
with every new heater strip - don't rely on them behaving the same. Even for a 
particular strip, use of an internal standard is best if you can use one that 
does not react with either the strip or the sample. If you push this furnace to 
temperatures above 1200degC or so, the platinum will creep and the tensioning 
will change, so that the height/temperature calibration may subsequently be off.

I agree fully with the suggestion to use parallel beam optics to eliminate the 
sample height error. You do lose some angular resolution of course, at least on 
a laboratory diffractometer, but you take one variable (sample height) out of 
the refinement.

The other BIG problem with the HT16 (and many other non-ambient stages) is 
knowing the temperature of the SAMPLE (as opposed to the strip). Powders have 
hopelessly poor thermal conductivity, and the gradients in these furnaces can 
be substantial. We rarely measure the temperature of the sample directly, and 
this is another reason why internal standards are very useful - you can use 
either a known CTE to estimate the delta T over the full temperature range, or 
a known phase transformation to determine delta T at a single point. I 
personally prefer the former approach myself. Note that a solid-solid 
transformation may give a different result than a melting point standard, since 
the liquid will change the thermal conductivity between the strip and the 
sample. You can find more discussion of these issues in Mark Rodriguez's 
chapter in the book Industrial Applications of X-ray Diffraction, or mine in 
Principles and Applications of Powder Diffraction.

I chair the non-ambient diffraction subcommittee of the ICDD, and in that 
capacity I am looking for ways we can develop and standardize these methods. If 
any of you would like to participate in this exercise please let me know. 
Perhaps a special issue of Powder Diffraction could be devoted to in-situ 
high-temperature diffraction papers? Always looking for good ideas...

Andrew


On 10/8/09 1:40 AM, Mikko Heikkilä mikko.j.heikk...@helsinki.fi wrote:

Hello Rory, Ian, and the rest.

I've been planning to do a similar calibration for a long time for our
Anton Paar HTK1200N but I've had to postpone it over and over again. The
approach by John Evans is very intriguing to me, but I'm lacking a
recent Topas version and I'm not sure if very old 2.1 will do it.

However, I think that a way to circumvent the simultanious refinement of
temperature shift and wrong height is to use parallel beam if you have
the possibility. On one hand you could first determine the temperature
behaviour as the peaks shouldn't really shift because of the wrong
height when using parallel beam. When the temperature behaviour is
known, change back to focusing geometry and calibrate the height. On the
other hand it might be possible to get the correct height in different
temperatures just by changing the sample holder position until it halves
the direct beam intensity. This is possible at least for our ceramic
sample holder, I'm not sure how this would work

RE: Re : HT XRD calibration

2009-10-15 Thread Yokochi, Alexandre
Hi all,

I've never used one of these Anton Paar furnaces, but with some of my 
instruments (externally heated, not Pt resistive foil heated) I could actually 
use the thermocouple as the sample mount.  I also had an instrument that had a 
thermocouple in contact with the powder sample, which due to thermal expansion 
mismatch would upon heating expand and split the sample surface giving rise to 
ugly split peaks so be careful... (you can imagine that implementation lasted 
as many samples as it took me to figure out what the peak splitting problem 
was).

Alex Y


Dr. Alexandre (Alex) F. T. Yokochi
Assistant Professor
Department of Chemical Engineering
Oregon State University
Corvallis, OR 97331-2702

Voice:  (541) 737-9357
Fax:  (541) 737-4600
http://oregonstate.edu/~yokochia

 -Original Message-
 From: Payzant, E. Andrew [mailto:payza...@ornl.gov]
 Sent: Thursday, October 15, 2009 6:27 AM
 To: Rietveld list
 Subject: Re: Re : HT XRD calibration
 
 Patrick,
 
 That is a good idea. Note that to actually get rid of the uncertainty you
 need to be certain that the thermocouple is either reading the temperature
 of the actual sample volume contributing to the diffraction pattern, or
 that it is reading the temperature of a location where the temperature is
 exactly the same as the sample volume.
 
 Andrew
 
 
 On 10/14/09 11:25 AM, Patrick Weisbecker weis...@yahoo.fr wrote:
 
 Hi,
 
 One way to get rid of the temperature uncertainty is to introduce a second
 thermocouple inside the furnace and to fix it as close as possible from
 the sample. If the sample is a bulk sample:Drill a hole inside the sample
 and put the thermocouple inside the hole and if the the sample is made of
 powder put it inside the powder.
 The are some miniature thermocouple (0.25-0.5mm diameter), mineral
 insulated which are unlikely to react with the samples.
 
 Patrick
 
 
 
 
 - Message d'origine 
 De : Payzant, E. Andrew payza...@ornl.gov
 À : Mikko Heikkilä mikko.j.heikk...@helsinki.fi
 Cc : Rietveld list rietveld_l@ill.fr
 Envoyé le : Mer 14 Octobre 2009, 16 h 39 min 58 s
 Objet : Re: HT XRD calibration
 
 All,
 
 The two calibration standard method was described by Andy Drews (Ford
 Research Lab) in Advances in X-ray Analysis (the proceedings of the Denver
 X-ray Conference), vol. 44, page 44 (2001).
 
 Scott Misture (Alfred Univ), Stephen Skinner (Imperial College), Christian
 Resch (Anton Paar), and I ran a workshop on non-ambient diffraction at the
 Denver X-ray Conference in 2008, where we discussed many of these issues.
 There was a lively discussion with the attendees regarding various
 experiences with HTXRD, and the difficulty in abtaining the right
 answer. Too often we are so focused on the precision and accuracy of our
 refinements that we neglect the temperature issue, to our peril.
 
 One issue with the HT16 hot stage is that you will likely need to
 recalibrate with every new heater strip - don't rely on them behaving the
 same. Even for a particular strip, use of an internal standard is best if
 you can use one that does not react with either the strip or the sample.
 If you push this furnace to temperatures above 1200degC or so, the
 platinum will creep and the tensioning will change, so that the
 height/temperature calibration may subsequently be off.
 
 I agree fully with the suggestion to use parallel beam optics to eliminate
 the sample height error. You do lose some angular resolution of course, at
 least on a laboratory diffractometer, but you take one variable (sample
 height) out of the refinement.
 
 The other BIG problem with the HT16 (and many other non-ambient stages) is
 knowing the temperature of the SAMPLE (as opposed to the strip). Powders
 have hopelessly poor thermal conductivity, and the gradients in these
 furnaces can be substantial. We rarely measure the temperature of the
 sample directly, and this is another reason why internal standards are
 very useful - you can use either a known CTE to estimate the delta T over
 the full temperature range, or a known phase transformation to determine
 delta T at a single point. I personally prefer the former approach myself.
 Note that a solid-solid transformation may give a different result than a
 melting point standard, since the liquid will change the thermal
 conductivity between the strip and the sample. You can find more
 discussion of these issues in Mark Rodriguez's chapter in the book
 Industrial Applications of X-ray Diffraction, or mine in Principles and
 Applications of Powder Diffraction.
 
 I chair the non-ambient diffraction subcommittee of the ICDD, and in
 that capacity I am looking for ways we can develop and standardize these
 methods. If any of you would like to participate in this exercise please
 let me know. Perhaps a special issue of Powder Diffraction could be
 devoted to in-situ high-temperature diffraction papers? Always looking for
 good ideas...
 
 Andrew
 
 
 On 10/8

Re: HT XRD calibration

2009-10-14 Thread Payzant, E. Andrew
All,

The two calibration standard method was described by Andy Drews (Ford Research 
Lab) in Advances in X-ray Analysis (the proceedings of the Denver X-ray 
Conference), vol. 44, page 44 (2001).

Scott Misture (Alfred Univ), Stephen Skinner (Imperial College), Christian 
Resch (Anton Paar), and I ran a workshop on non-ambient diffraction at the 
Denver X-ray Conference in 2008, where we discussed many of these issues. There 
was a lively discussion with the attendees regarding various experiences with 
HTXRD, and the difficulty in abtaining the right answer. Too often we are so 
focused on the precision and accuracy of our refinements that we neglect the 
temperature issue, to our peril.

One issue with the HT16 hot stage is that you will likely need to recalibrate 
with every new heater strip - don't rely on them behaving the same. Even for a 
particular strip, use of an internal standard is best if you can use one that 
does not react with either the strip or the sample. If you push this furnace to 
temperatures above 1200degC or so, the platinum will creep and the tensioning 
will change, so that the height/temperature calibration may subsequently be off.

I agree fully with the suggestion to use parallel beam optics to eliminate the 
sample height error. You do lose some angular resolution of course, at least on 
a laboratory diffractometer, but you take one variable (sample height) out of 
the refinement.

The other BIG problem with the HT16 (and many other non-ambient stages) is 
knowing the temperature of the SAMPLE (as opposed to the strip). Powders have 
hopelessly poor thermal conductivity, and the gradients in these furnaces can 
be substantial. We rarely measure the temperature of the sample directly, and 
this is another reason why internal standards are very useful - you can use 
either a known CTE to estimate the delta T over the full temperature range, or 
a known phase transformation to determine delta T at a single point. I 
personally prefer the former approach myself. Note that a solid-solid 
transformation may give a different result than a melting point standard, since 
the liquid will change the thermal conductivity between the strip and the 
sample. You can find more discussion of these issues in Mark Rodriguez's 
chapter in the book Industrial Applications of X-ray Diffraction, or mine in 
Principles and Applications of Powder Diffraction.

I chair the non-ambient diffraction subcommittee of the ICDD, and in that 
capacity I am looking for ways we can develop and standardize these methods. If 
any of you would like to participate in this exercise please let me know. 
Perhaps a special issue of Powder Diffraction could be devoted to in-situ 
high-temperature diffraction papers? Always looking for good ideas...

Andrew


On 10/8/09 1:40 AM, Mikko Heikkilä mikko.j.heikk...@helsinki.fi wrote:

Hello Rory, Ian, and the rest.

I've been planning to do a similar calibration for a long time for our
Anton Paar HTK1200N but I've had to postpone it over and over again. The
approach by John Evans is very intriguing to me, but I'm lacking a
recent Topas version and I'm not sure if very old 2.1 will do it.

However, I think that a way to circumvent the simultanious refinement of
temperature shift and wrong height is to use parallel beam if you have
the possibility. On one hand you could first determine the temperature
behaviour as the peaks shouldn't really shift because of the wrong
height when using parallel beam. When the temperature behaviour is
known, change back to focusing geometry and calibrate the height. On the
other hand it might be possible to get the correct height in different
temperatures just by changing the sample holder position until it halves
the direct beam intensity. This is possible at least for our ceramic
sample holder, I'm not sure how this would work with Pt-strip in other
ovens.


I'm very interested about suitable test materials. I was planning to go
with corundum and parallel beam first, but now that this thread is open,
I'll wait a bit longer (again). Regarding the calibration with phase
change temperature, Robert Dinnebier has some materials listed on his
webpage: http://www.fkf.mpg.de/xray/html/temperature_calibration.html

Please correct me if I've made much mistakes above, I'm still at rather
beginning of my learning curve :-)

Cheers,
Mikko





ian.mad...@csiro.au wrote:
 Hi Rory,

 At the moment we are in the same position i.e. attempting to calibrate the 
 temperature and systematic offsets in an HTK10 furnace with Pt strip heater. 
 One approach is to take materials which have a well defined solid-solid phase 
 transition or melting point and observe this as a function of set 
 temperature. Unfortunately, most materials only provide one or two 
 calibration points and hence you will need to heat several samples in order 
 to cover a wide temperature range.

 At the moment we (and by we I mean my PhD student Mark Styles is doing all 
 the work and I 

Re : HT XRD calibration

2009-10-14 Thread Patrick Weisbecker
Hi,

One way to get rid of the temperature uncertainty is to introduce a second 
thermocouple inside the furnace and to fix it as close as possible from the 
sample. If the sample is a bulk sample:Drill a hole inside the sample and put 
the thermocouple inside the hole and if the the sample is made of powder put it 
inside the powder. 
The are some miniature thermocouple (0.25-0.5mm diameter), mineral insulated 
which are unlikely to react with the samples.

Patrick

 


- Message d'origine 
De : Payzant, E. Andrew payza...@ornl.gov
À : Mikko Heikkilä mikko.j.heikk...@helsinki.fi
Cc : Rietveld list rietveld_l@ill.fr
Envoyé le : Mer 14 Octobre 2009, 16 h 39 min 58 s
Objet : Re: HT XRD calibration

All,

The two calibration standard method was described by Andy Drews (Ford Research 
Lab) in Advances in X-ray Analysis (the proceedings of the Denver X-ray 
Conference), vol. 44, page 44 (2001).

Scott Misture (Alfred Univ), Stephen Skinner (Imperial College), Christian 
Resch (Anton Paar), and I ran a workshop on non-ambient diffraction at the 
Denver X-ray Conference in 2008, where we discussed many of these issues. There 
was a lively discussion with the attendees regarding various experiences with 
HTXRD, and the difficulty in abtaining the right answer. Too often we are so 
focused on the precision and accuracy of our refinements that we neglect the 
temperature issue, to our peril.

One issue with the HT16 hot stage is that you will likely need to recalibrate 
with every new heater strip - don't rely on them behaving the same. Even for a 
particular strip, use of an internal standard is best if you can use one that 
does not react with either the strip or the sample. If you push this furnace to 
temperatures above 1200degC or so, the platinum will creep and the tensioning 
will change, so that the height/temperature calibration may subsequently be off.

I agree fully with the suggestion to use parallel beam optics to eliminate the 
sample height error. You do lose some angular resolution of course, at least on 
a laboratory diffractometer, but you take one variable (sample height) out of 
the refinement.

The other BIG problem with the HT16 (and many other non-ambient stages) is 
knowing the temperature of the SAMPLE (as opposed to the strip). Powders have 
hopelessly poor thermal conductivity, and the gradients in these furnaces can 
be substantial. We rarely measure the temperature of the sample directly, and 
this is another reason why internal standards are very useful - you can use 
either a known CTE to estimate the delta T over the full temperature range, or 
a known phase transformation to determine delta T at a single point. I 
personally prefer the former approach myself. Note that a solid-solid 
transformation may give a different result than a melting point standard, since 
the liquid will change the thermal conductivity between the strip and the 
sample. You can find more discussion of these issues in Mark Rodriguez's 
chapter in the book Industrial Applications of X-ray Diffraction, or mine in 
Principles and Applications of Powder Diffraction.

I chair the non-ambient diffraction subcommittee of the ICDD, and in that 
capacity I am looking for ways we can develop and standardize these methods. If 
any of you would like to participate in this exercise please let me know. 
Perhaps a special issue of Powder Diffraction could be devoted to in-situ 
high-temperature diffraction papers? Always looking for good ideas...

Andrew


On 10/8/09 1:40 AM, Mikko Heikkilä mikko.j.heikk...@helsinki.fi wrote:

Hello Rory, Ian, and the rest.

I've been planning to do a similar calibration for a long time for our
Anton Paar HTK1200N but I've had to postpone it over and over again. The
approach by John Evans is very intriguing to me, but I'm lacking a
recent Topas version and I'm not sure if very old 2.1 will do it.

However, I think that a way to circumvent the simultanious refinement of
temperature shift and wrong height is to use parallel beam if you have
the possibility. On one hand you could first determine the temperature
behaviour as the peaks shouldn't really shift because of the wrong
height when using parallel beam. When the temperature behaviour is
known, change back to focusing geometry and calibrate the height. On the
other hand it might be possible to get the correct height in different
temperatures just by changing the sample holder position until it halves
the direct beam intensity. This is possible at least for our ceramic
sample holder, I'm not sure how this would work with Pt-strip in other
ovens.


I'm very interested about suitable test materials. I was planning to go
with corundum and parallel beam first, but now that this thread is open,
I'll wait a bit longer (again). Regarding the calibration with phase
change temperature, Robert Dinnebier has some materials listed on his
webpage: http://www.fkf.mpg.de/xray/html/temperature_calibration.html

Please correct me if I've made much

Re: HT XRD calibration

2009-10-07 Thread Mikko Heikkilä
Hello Rory, Ian, and the rest.

I've been planning to do a similar calibration for a long time for our
Anton Paar HTK1200N but I've had to postpone it over and over again. The
approach by John Evans is very intriguing to me, but I'm lacking a
recent Topas version and I'm not sure if very old 2.1 will do it.

However, I think that a way to circumvent the simultanious refinement of
temperature shift and wrong height is to use parallel beam if you have
the possibility. On one hand you could first determine the temperature
behaviour as the peaks shouldn't really shift because of the wrong
height when using parallel beam. When the temperature behaviour is
known, change back to focusing geometry and calibrate the height. On the
other hand it might be possible to get the correct height in different
temperatures just by changing the sample holder position until it halves
the direct beam intensity. This is possible at least for our ceramic
sample holder, I'm not sure how this would work with Pt-strip in other
ovens.


I'm very interested about suitable test materials. I was planning to go
with corundum and parallel beam first, but now that this thread is open,
I'll wait a bit longer (again). Regarding the calibration with phase
change temperature, Robert Dinnebier has some materials listed on his
webpage: http://www.fkf.mpg.de/xray/html/temperature_calibration.html

Please correct me if I've made much mistakes above, I'm still at rather
beginning of my learning curve :-)

Cheers,
Mikko





ian.mad...@csiro.au wrote:
 Hi Rory,
 
 At the moment we are in the same position i.e. attempting to calibrate the 
 temperature and systematic offsets in an HTK10 furnace with Pt strip heater. 
 One approach is to take materials which have a well defined solid-solid phase 
 transition or melting point and observe this as a function of set 
 temperature. Unfortunately, most materials only provide one or two 
 calibration points and hence you will need to heat several samples in order 
 to cover a wide temperature range.
 
 At the moment we (and by we I mean my PhD student Mark Styles is doing all 
 the work and I stick my head in occasionally and say how's it going?) have 
 taken an 'external standard' approach (we cannot add a standard to our sample 
 and risk reaction). In this we heat a material with known thermal coefficient 
 of expansion (TCE), measure the unit cell edge a function of temperature and 
 then determine a calibration curve. However, the flaw in this approach will 
 be the impact of other systematic errors (e.g. sample displacement) on the 
 observed peak positions and hence cell edge.
 
 John Evans (Durham Uni) uses a modified TCE approach in which he adds two 
 standards with different TCE's (I haven't got the references to hand, but 
 this is described in one of his papers) and normalises observed cell edge (or 
 cell volume) to that at the starting temperature (To). Instead of using the 
 absolute cell edge (volume) for each phase, he then uses the difference 
 between the two compounds as a way of eliminating the systematic errors which 
 should affect both equally.
 
 If you pardon the pun, this is a hot topic for us at the moment - we are 
 coming up to some synchrotron time where T calibration will be important - we 
 have not completely sorted out the fine details but intend to:-
 
 (1)   Use a mixture of NIST silicon and corundum
 (2)   Heat to various temperatures over the range of interest
 (3)   Use the 'surface analysis' approach (Stinton and Evans) which relies on 
 use of the Topas Rietveld software. In this, the cell edge is not refined, 
 but defined as a function of applied temperature, the temperature offset (a 
 refinable parameter) and the published TCE equations for each phase. 
 
 What we hope to end up with is an analytical method which is quick and easy 
 to apply to produce a T calibration for every new heater - i.e. it should not 
 be an onerous task so that it can be carried out frequently.
 
 I hope this helps - we still have much fine detail to sort out.
   
 Cheers
  
 o--oo0oo---o
  Ian Madsen
  Team Leader - Diffraction Science
  CSIRO Minerals
  Box 312
  Clayton South 3169
  Victoria
  AUSTRALIA
  Phone +61 3 9545 8785 direct
  +61 3 9545 8500 switch
  +61 (0) 417 554 935 mobile
  FAX+61 3 9562 8919
  Email ian.mad...@csiro.au 
 o--oo0oo---o
 
 -Original Message-
 From: r.m.wil...@qmul.ac.uk [mailto:r.m.wil...@qmul.ac.uk] 
 Sent: Thursday, 8 October 2009 3:14 AM
 To: rietveld_l@ill.fr
 Subject: HT XRD calibration
 
 
  Dear All
   We have an Anton Paar HTK16 furnace with a Pt strip. I am  
 hoping to calibrate it both thermally and mechanically (height shift)  
 using either alpha alumina or ceria so as to better identify HT  
 phases. Have any of you experience of this, or better suggestions