Re: [Elphel-support] Extension of stereobase and elongation sensor's cable length

2015-04-16 Thread Oleg
 We got a good results with 0x1ff80 value for MULTI_PHASE[n] in both
 channels without changing SENSOR_PHASE.

 I would measure the range around 0x1ff80 while the image is still good,
then set the phase to the middle. For a 450mm cable the working range is
~60-100 steps.

At first we get a broken images at same manner. By the way, if we use 450
 mm cable with MULTI_PHASE = 0x1ff80 we get same problem, so MULTI_PHASE is
 strongly depends on cable length.

Yes - signal propagation delay.

If the experiment will give us a good results we intend to expand a
 stereobase for second setup at the same manner. In with case we will order
 to you addition cables.
 Do you offer now a sensor board (compatible with our 10359) with
 monochrome sensor? If so what type of sensor installed into such board?


We do have some monochrome sensor boards.
You can buy them but using a monochrome sensor reduces overall resolution
of the sensor+lens combination. And most of nowadays the applications deal
with color sensors.
So, you can:
a. Keep using a color sensor - just work with R, G and B channels
separately, then combine.
b. Get a monochrome sensor - but use a green color band filter in front of
the lens to a have better resolution.

Best regards,
Oleg Dzhimiev
Electronics Engineer
phone: +1 801 783  x124
Elphel, Inc.
___
Support-list mailing list
Support-list@support.elphel.com
http://support.elphel.com/mailman/listinfo/support-list_support.elphel.com


Re: [Elphel-support] Extension of stereobase and elongation sensor's cable length

2015-04-16 Thread Oleg
Forgot to send a copy back to the mailing-list:


  I found that next values are:

SENSOR_PHASE - 0x9
 MULTI_PHASE1 - 0x100b0.


1.
I don't know.
With SENSOR_PHASE unchanged try several points within each quarter
for MULTI_PHASE1, e.g.:
0x8ff70, 0x8, 0x80070,
0x9ff70, ...
0xaff70, ...
0xbff70, 0xb, 0xb0070.

Sometimes a phase shifter (digital clock manager in the fpga) can get stuck
- so it's safer to set 15th bit to 1 to reset it.
You can try with a 450mm cable if the long one can be easily disassembled.

2.
How exactly do you connect 2 cables? picture?
Any difference in the voltage levels on the ends between 450mm and 900mm
setups?

3.
If both images are broken in the same manner then try to change
the SENSOR_PHASE.

Best regards,
Oleg Dzhimiev
Electronics Engineer
phone: +1 801 783  x124
Elphel, Inc.
___
Support-list mailing list
Support-list@support.elphel.com
http://support.elphel.com/mailman/listinfo/support-list_support.elphel.com


Re: [Elphel-support] Extension of stereobase and elongation sensor's cable length

2015-04-16 Thread Клещенков Анатолий


17.04.2015 0:05, Oleg пишет:


We do have some monochrome sensor boards.
You can buy them but using a monochrome sensor reduces overall 
resolution of the sensor+lens combination. And most of nowadays the 
applications deal with color sensors.

So, you can:
a. Keep using a color sensor - just work with R, G and B channels 
separately, then combine.
b. Get a monochrome sensor - but use a green color band filter in 
front of the lens to a have better resolution.


I am not familiar with sensor's details but I know that in color sensor 
actual resolution for every color channel  is lower than resolution 
provided in specification.
For example if we consider color sensor 200 x 200 px then actual 
resolution for Red and Blue channel is 100 x 100 because of Bayer 
filter. Final image resolution 200 x 200 is obtained by interpolation.
Most of mine applications are based on stereovision and processing of 
monochrome images. As you know most of the modern stereo-matching 
techniques operate with grayscale images too.
My another knowledge is that monochrome sensor has a much higher 
sensitivity than color one at the same resolution. That is why I guess 
that using the monochrome sensor let me obtain higher ratio of 
signal/noise at the same resolution than color sensor. What do you think 
about that?
From the other hand you mentioned about reducing of overall resolution 
of the sensor+lens combination. Do you mean reason of light dispersion 
in the lenses?  Actually we have done a lot of experiments to find out a 
acceptable quality of the lenses for 5PMx resolution. We use now high 
quality computar M0918FIC-MPlenses. These lenses gives the best results 
of the all we try but anyway I suppose these lenses do not utilize a 
full resolution of the sensor. I think so because images analysis shows 
that high contrast contour line appears at the images like smooth 
contour with several pixels width. At the same time at the half (about 
1MP) resolution we obtain much more contrast contour.





Best regards,
Oleg Dzhimiev
Electronics Engineer
phone: +1 801 783  x124
Elphel, Inc.


--
С уважением, Клещенков Анатолий.
г. Ростов-на-Дону
теп.: +7(863)275-95-05
e-mail: akt...@inbox.ru

___
Support-list mailing list
Support-list@support.elphel.com
http://support.elphel.com/mailman/listinfo/support-list_support.elphel.com


Re: [Elphel-support] Extension of stereobase and elongation sensor's cable length

2015-04-16 Thread Клещенков Анатолий


Hello!

Thank you for fast response!

16.04.2015 21:48, Oleg пишет:

Hi,

To do so we have connected two cables in serial, so we obtained
total cable length 450mm + 450 mm = 900 mm between 359 board and
each sensor board.

There's no guarantee this length will work.

We try use control page http://192.168.0.9/359/10359_controls.html
but it is not clear which values I have to change.

I'd need to test if it still works fine.

A better way to adjust phases is this one:
Go to: http://192.168.0.9/   Parameter Editor  Check Unsafe  
View/Edit Current
Or here's a link 
http://192.168.0.9/parsedit.php?title=Parameters+for+groups:+unsafe+CLK_FPGACLK_SENSORSENSOR_PHASEI2C_QPERIODI2C_BYTESMULTI_PHASE_SDRAMMULTI_PHASE1MULTI_PHASE2MULTI_PHASE3.


The phases are changed with:
SENSOR_PHASE - system board - MUX (in stereo setup)
MULTI_PHASE1 - MUX - sensor 1
MULTI_PHASE2 - MUX - sensor 2
MULTI_PHASE3...

bits[18:17] - 90 degree shift
bits[15:0]  - fine steps (with sign) 0xff01...0x00ff


I found that next values are:
SENSOR_PHASE - 0x9
MULTI_PHASE1 - 0x100b0.

Which new values need I to try setup here?




Best regards,
Oleg Dzhimiev
Electronics Engineer
phone: +1 801 783  x124
Elphel, Inc.


--
С уважением, Клещенков Анатолий.
г. Ростов-на-Дону
теп.: +7(863)275-95-05
e-mail: akt...@inbox.ru

___
Support-list mailing list
Support-list@support.elphel.com
http://support.elphel.com/mailman/listinfo/support-list_support.elphel.com


Re: [Elphel-support] Extension of stereobase and elongation sensor's cable length

2015-04-16 Thread Oleg
Hi,

To do so we have connected two cables in serial, so we obtained total cable
 length 450mm + 450 mm = 900 mm between 359 board and each sensor board.


There's no guarantee this length will work.

We try use control page http://192.168.0.9/359/10359_controls.html but it
 is not clear which values I have to change.

I'd need to test if it still works fine.

A better way to adjust phases is this one:
Go to: http://192.168.0.9/   Parameter Editor  Check Unsafe 
View/Edit Current
Or here's a link
http://192.168.0.9/parsedit.php?title=Parameters+for+groups:+unsafe+CLK_FPGACLK_SENSORSENSOR_PHASEI2C_QPERIODI2C_BYTESMULTI_PHASE_SDRAMMULTI_PHASE1MULTI_PHASE2MULTI_PHASE3
.

The phases are changed with:
SENSOR_PHASE - system board - MUX (in stereo setup)
MULTI_PHASE1 - MUX - sensor 1
MULTI_PHASE2 - MUX - sensor 2
MULTI_PHASE3...

bits[18:17] - 90 degree shift
bits[15:0]  - fine steps (with sign) 0xff01...0x00ff

Best regards,
Oleg Dzhimiev
Electronics Engineer
phone: +1 801 783  x124
Elphel, Inc.
___
Support-list mailing list
Support-list@support.elphel.com
http://support.elphel.com/mailman/listinfo/support-list_support.elphel.com