Re: [time-nuts] FE-5680A heat sink

2009-06-08 Thread Chuck Harris

I'm puzzled.  I admit that I don't have a lot of experience with
Rb standards, but I do have a bit of experience with the HP-5065A
Rb standard.

In the 5065A, the entire physics package is enclosed in an oven.

Assuming that your Rb is the same, and I believe it is, your plan
to force the physics package to run at room temperature is just going
to make the oven work harder in its never ending quest to maintain
stable temperature.  If you monitor the current draw of the 5680A,
you will probably see that it goes up when you put a fan on it.

Sometimes, you just have to let electronics run hot.

What did the manufacturer suggest?

-Chuck Harris

Leigh L. Klotz, Jr WA5ZNU wrote:
I've attached an SMA connector to my FE-5680A and built an external 
linear power supply with a TO-3 7815.


I ran the device today for about half an hour, and used an infrared 
sensing thermometer to measure the external case temperature.

It got up to 48 C externally in the physics package area.
I put a muffin fan on top and it brought it down to 38C but I don't 
think this is a good plan because I worry about the effects of the fan's 
magnetic field on the Rubidium system.
I found in my junk box a finned Aluminum heat sink that's exactly the 
same size as the FE-5680A and plan to tap it around the edges for 4-40 
hardware to attach to the many screw holes.  Even so, this heat sink 
will be on the bottom, so the FE-5680A will have to be operated upside 
down for this to help.


Has anybody got good thermal management solution for this device?  This 
is the one currently selling on eBay in the 25x88x125mm chassis.


Thanks,
Leigh.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

and follow the instructions there.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FE-5680A heat sink

2009-06-08 Thread Poul-Henning Kamp
In message 4a2cfce2.5020...@erols.com, Chuck Harris writes:

 I ran the device today for about half an hour, and used an infrared 
 sensing thermometer to measure the external case temperature.

Be very careful about trusting this:  you need to do some tricky
calibrations to get anywhere near precise when you measure metal
surfaces.

The easy way, is to put a piece of duc[kt]tape on the metal surface
and make sure your thermometer can see only that surface.

Unfortunately, the tape will also act as insulation, so the result
you get is not precise even then.

 It got up to 48 C externally in the physics package area.

That's quite normal.

Those small Rb's keep the internal temperature constant using
heaters, which can raise the temperature and by being able to dump
excess heat through their heat-sink to lower the temperature.

You shouldn't run your Rb too hot, as this decreases the electronics
lifetime and reduces the wiggle-room of the thermal management
inside the device.

On the other hand, cooling it too much will only increase the
power drain for the heaters and increase the thermal gradients
inside the unit, likely degrading thermal stability.

Poul-Henning

-- 
Poul-Henning Kamp   | UNIX since Zilog Zeus 3.20
p...@freebsd.org | TCP/IP since RFC 956
FreeBSD committer   | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FE-5680A heat sink

2009-06-08 Thread Chuck Harris

Hi Poul,

By replying to my reply to Leigh, and clipping out everything
that I wrote (but my name), you seem to be attributing to me,
what Leigh wrote.

You then rephrased my statement about heaters in the physics
package, restated my statement about extra cooling increasing
the power drawn by the heaters.

You then clarified things greatly by advising to not run the Rb
too hot, but also don't cool it too much.

???

If you have so much to say to the original author, and nothing to
say about my reply, wouldn't it have been be better to just reply
to his message, instead of mine?

-Chuck Harris

Poul-Henning Kamp wrote:

In message 4a2cfce2.5020...@erols.com, Chuck Harris writes:

I ran the device today for about half an hour, and used an infrared 
sensing thermometer to measure the external case temperature.


Be very careful about trusting this:  you need to do some tricky
calibrations to get anywhere near precise when you measure metal
surfaces.

The easy way, is to put a piece of duc[kt]tape on the metal surface
and make sure your thermometer can see only that surface.

Unfortunately, the tape will also act as insulation, so the result
you get is not precise even then.


It got up to 48 C externally in the physics package area.


That's quite normal.

Those small Rb's keep the internal temperature constant using
heaters, which can raise the temperature and by being able to dump
excess heat through their heat-sink to lower the temperature.

You shouldn't run your Rb too hot, as this decreases the electronics
lifetime and reduces the wiggle-room of the thermal management
inside the device.

On the other hand, cooling it too much will only increase the
power drain for the heaters and increase the thermal gradients
inside the unit, likely degrading thermal stability.

Poul-Henning



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] Datron 4910 service manual request

2009-06-08 Thread Frans

Does anyone has a service manual for the Datron / Wavetek 4910 voltage 
standard. The output voltage is correct. I have problems with battery power.

Met vriendelijke groeten, Regards,

ing. Frans de Haan
Haarlem
the Netherlands



  

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FE-5680A heat sink

2009-06-08 Thread Poul-Henning Kamp
In message 4a2d129c.3060...@erols.com, Chuck Harris writes:

If you have so much to say to the original author, and nothing to
say about my reply, wouldn't it have been be better to just reply
to his message, instead of mine?

What can I say ?   It was early in the morning ?

See also:  my .sig 

:-)

Poul-Henning


-- 
Poul-Henning Kamp   | UNIX since Zilog Zeus 3.20
p...@freebsd.org | TCP/IP since RFC 956
FreeBSD committer   | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] FE-5680A thermal management

2009-06-08 Thread Murray Greenman
Leigh,

I agree with Chuck. I have both an FE-5650A and an FE-5680A. With the
former I was concerned about the heat, and so ran it only for short
periods, until I understood what was going on. I had the impression from
the data sheet that there were different heatsink options for different
temperature ranges, and I now believe this led me astray.

With the FE-5680A I had the opportunity to study things in more detail.
There is no temperature range specification that I could find, and no
particular advice in the manual regarding installation. I ran the unit
from regulated 15V DC, and monitored the supply current. With no extra
cooling, the steady state current was about 700mA. With air blown over
it, the current increased. With the unit placed in a poly bag, the
current decreased to about 650mA. By the way, the current also decreased
when operated from 17V DC.

This tells me that the whole structure is part of the thermal package,
and we should not attempt to force down the case temperature just
because the unit runs hot. My experience with high performance OCXOs
tells me that the thermal environment is carefully designed, and part of
the calibration process - if you modify this environment (by cooling or
extra insulation) you modify the thermal environment, and are at risk of
modifying the performance.

I'd leave well alone, and run the unit in an open, breeze-free
environment.

73,
Murray ZL1BPU


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FE-5680A thermal management

2009-06-08 Thread Poul-Henning Kamp
In message be50c3a72eba61449c804b2c8b4ae38101ebb...@neptune.rakon.net, Murra
y Greenman writes:

With the FE-5680A I had the opportunity to study things in more detail.

The PRS10 manual has some good info.

-- 
Poul-Henning Kamp   | UNIX since Zilog Zeus 3.20
p...@freebsd.org | TCP/IP since RFC 956
FreeBSD committer   | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FE-5680A heat sink

2009-06-08 Thread Chuck Harris

Hi Poul,

I have long espoused that one should never attribute to malice
that which can be more easily explained by ignorance.  So, I didn't
believe malice was involved.

I just found your post curious, that's all.

I'm pretty sure that mornings should be banned.

-Chuck Harris



Poul-Henning Kamp wrote:

In message 4a2d129c.3060...@erols.com, Chuck Harris writes:


If you have so much to say to the original author, and nothing to
say about my reply, wouldn't it have been be better to just reply
to his message, instead of mine?


What can I say ?   It was early in the morning ?

See also:  my .sig 


:-)

Poul-Henning




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FE-5680A heat sink

2009-06-08 Thread M. Warner Losh
In message: 4a2d60a2.3030...@erols.com
Chuck Harris cfhar...@erols.com writes:
: I'm pretty sure that mornings should be banned.

Perty much...  Nobody has a breathalizer to ensure that you are
sufficiently caffeinated to give a good chance of a coherent reply :)
Lord know that would have saved me much embarrassment over the
years...

However, since this is time-nuts, and we do deal with things on the
hairy edge of what is possible, I'm sure someone will point to a side
project that they've done that does just this, with schematics
available for download form their web site :)

Warner

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FE-5680A thermal management

2009-06-08 Thread bg
 In message be50c3a72eba61449c804b2c8b4ae38101ebb...@neptune.rakon.net,
 Murray Greenman writes:

With the FE-5680A I had the opportunity to study things in more detail.

 The PRS10 manual has some good info.

 --
 Poul-Henning Kamp   | UNIX since Zilog Zeus 3.20
 p...@freebsd.org | TCP/IP since RFC 956
 FreeBSD committer   | BSD since 4.3-tahoe
 Never attribute to malice what can adequately be explained by
 incompetence.

Ok, hope everyone can see what above was written by Poul and Murray
respectively. ;-)

For the FRS-C read page 7 (chapter 2.4 Mounting in the below manual.

http://www.to-way.com/frs.pdf

It says the base plate should never go above 65 deg C.

--

Björn


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Datron 4910 service manual request

2009-06-08 Thread Steve Krull
I'd like to find a copy, too. Same problem, and there doesn't seem to be 
replacement batteries that are fit-and-form equivalents.


Thanks,

Steve



On Mon, Jun 8, 2009 at 3:35 PM , Frans wrote:

Does anyone has a service manual for the Datron / Wavetek 4910 voltage 
standard. The output voltage is correct. I have problems with battery 
power.


Met vriendelijke groeten, Regards,

ing. Frans de Haan
Haarlem
the Netherlands




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FE-5680A thermal management

2009-06-08 Thread Dave Brown


- Original Message - 
From: Murray Greenman murray.green...@rakon.com

To: time-nuts@febo.com
Sent: Tuesday, June 09, 2009 6:42 AM
Subject: [time-nuts] FE-5680A thermal management



Leigh,

I agree with Chuck. I have both an FE-5650A and an FE-5680A. With 
the

former I was concerned about the heat, and so ran it only for short
periods, until I understood what was going on. I had the impression 
from
the data sheet that there were different heatsink options for 
different

temperature ranges, and I now believe this led me astray.

With the FE-5680A I had the opportunity to study things in more 
detail.
There is no temperature range specification that I could find, and 
no
particular advice in the manual regarding installation. I ran the 
unit
from regulated 15V DC, and monitored the supply current. With no 
extra
cooling, the steady state current was about 700mA. With air blown 
over

it, the current increased. With the unit placed in a poly bag, the
current decreased to about 650mA. By the way, the current also 
decreased

when operated from 17V DC.

This tells me that the whole structure is part of the thermal 
package,

and we should not attempt to force down the case temperature just
because the unit runs hot. My experience with high performance OCXOs
tells me that the thermal environment is carefully designed, and 
part of
the calibration process - if you modify this environment (by cooling 
or
extra insulation) you modify the thermal environment, and are at 
risk of

modifying the performance.

I'd leave well alone, and run the unit in an open, breeze-free
environment.

73,
Murray ZL1BPU




Murray et al,
The package of both units is clearly meant to be mounted on 
something-the real question is how much heat sinking would that 
something have provided?  Changing the external heatsinking to achieve 
the nominal supply current at the nominal supply voltage would appear 
to be the only simple way to operate the units as intended.


I've not gone through this exercise yet with the 5650, but I suspect 
the amount of heatsinking used is not that critical. Using a variable 
speed fan to determine the 'correct' mounting plate temperature 
(corresponding to nominal supply current/voltage) might be a good way 
to start. The design of a heatsink to achieve the same base plate 
temperature should be a trivial exercise.
Constraining the 'ambient' air temperature to an appropriate range in 
the vicinity of the unit should certainly help as well.


Regards
DaveB, NZ


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Datron 4910 service manual request

2009-06-08 Thread wje
I have a manual, but it's hard-copy and LARGE. If you can narrow down 
your problem area, I'd be happy to scan some schematics.
I also have two 4910's; I can probe some test points if you need some 
values.


Bill Ezell
--
They said 'Windows or better'
so I used Linux.



Steve Krull wrote:
div class=moz-text-flowed style=font-family: -moz-fixedI'd like 
to find a copy, too. Same problem, and there doesn't seem to be 
replacement batteries that are fit-and-form equivalents.


Thanks,

Steve



On Mon, Jun 8, 2009 at 3:35 PM , Frans wrote:

Does anyone has a service manual for the Datron / Wavetek 4910 
voltage standard. The output voltage is correct. I have problems with 
battery power.


Met vriendelijke groeten, Regards,

ing. Frans de Haan
Haarlem
the Netherlands




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

and follow the instructions there.



/div


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] FE-5680A thermal management

2009-06-08 Thread Mark Sims

Almost all rubidium standards DO specify the use of some form of a heat sink.  
For the FRK and M100 units this can be a heat sink with around 1 fins or just 
bolted to a metal plate or chassis.

The military freq standard that had M100's in them had the unit mounted to a 
5x5x.2 aluminum plate that was in turn bolted to the chassis.   My Efratom 
PTB-100 time bases for the Tektronix  TM500 mainframes have a large heat sink 
mounted to a FRK style oscillator.  

LPROs are supposed to be mounted to a metal chassis (they usually come with a 
thermal pad attached to them).  I have seen them lose lock in free air.  I saw 
one mounted in a piece of cell phone equipment.  It was bolted to a large heat 
sink that formed most of the front of the enclosure.

I have also seen FE-5680A's in their native habitat (again,  cell phone 
equipment).  They were mounted to a thick (1/8?) PCB around 6x16.  The side 
of the PCB that the 5680 was bolted to had a solid ground plane.  The 5650A's 
had one side bolted to a metal chassis.




_
Hotmail® has ever-growing storage! Don’t worry about storage limits. 
http://windowslive.com/Tutorial/Hotmail/Storage?ocid=TXT_TAGLM_WL_HM_Tutorial_Storage_062009
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FE-5680A heat sink

2009-06-08 Thread Leigh L. Klotz, Jr WA5ZNU

Chuck,
This device is a pull from a larger system, probably a cell site.  It's 
clearly designed to be mounted on something, as it has lots and lots of 
holes around the edge.  I suspect there's some thermal management that's 
missing.  The FEI sheet gives typical data for the 0-50C range, though 
presumably that's ambient temperature. 

I don't want to cool the physics package per se, but I do want to at 
least approximate what kind of thermal solution ought to be supplied.


Quite a few of these have been sold, and I've gotten good advice from 
others on this list about calibration.  I'd hoped that someone would 
have experience with the thermal management.


So far I've seen 
http://www.dd1us.de/Downloads/precise%20reference%20frequency%20rev%200_4.pdf

which shows (but doesn't describe) a heat sink on the bottom.

Leigh.


I'm puzzled.  I admit that I don't have a lot of experience with
Rb standards, but I do have a bit of experience with the HP-5065A
Rb standard.

In the 5065A, the entire physics package is enclosed in an oven.

Assuming that your Rb is the same, and I believe it is, your plan
to force the physics package to run at room temperature is just going
to make the oven work harder in its never ending quest to maintain
stable temperature.  If you monitor the current draw of the 5680A,
you will probably see that it goes up when you put a fan on it.

Sometimes, you just have to let electronics run hot.

What did the manufacturer suggest?

-Chuck Harris

Leigh L. Klotz, Jr WA5ZNU wrote:
I've attached an SMA connector to my FE-5680A and built an external 
linear power supply with a TO-3 7815.


I ran the device today for about half an hour, and used an infrared 
sensing thermometer to measure the external case temperature.

It got up to 48 C externally in the physics package area.
I put a muffin fan on top and it brought it down to 38C but I don't 
think this is a good plan because I worry about the effects of the 
fan's magnetic field on the Rubidium system.
I found in my junk box a finned Aluminum heat sink that's exactly the 
same size as the FE-5680A and plan to tap it around the edges for 
4-40 hardware to attach to the many screw holes.  Even so, this heat 
sink will be on the bottom, so the FE-5680A will have to be operated 
upside down for this to help.


Has anybody got good thermal management solution for this device?  
This is the one currently selling on eBay in the 25x88x125mm chassis.


Thanks,
Leigh.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

and follow the instructions there.


the instructions there.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FE-5680A heat sink

2009-06-08 Thread Steve Rooke
Calm down Chuck, your doing it again.

2009/6/9 Chuck Harris cfhar...@erols.com:
 Hi Poul,

 By replying to my reply to Leigh, and clipping out everything
 that I wrote (but my name), you seem to be attributing to me,
 what Leigh wrote.

 You then rephrased my statement about heaters in the physics
 package, restated my statement about extra cooling increasing
 the power drawn by the heaters.

 You then clarified things greatly by advising to not run the Rb
 too hot, but also don't cool it too much.

 ???

 If you have so much to say to the original author, and nothing to
 say about my reply, wouldn't it have been be better to just reply
 to his message, instead of mine?

 -Chuck Harris

 Poul-Henning Kamp wrote:

 In message 4a2cfce2.5020...@erols.com, Chuck Harris writes:

 I ran the device today for about half an hour, and used an infrared
 sensing thermometer to measure the external case temperature.

 Be very careful about trusting this:  you need to do some tricky
 calibrations to get anywhere near precise when you measure metal
 surfaces.

 The easy way, is to put a piece of duc[kt]tape on the metal surface
 and make sure your thermometer can see only that surface.

 Unfortunately, the tape will also act as insulation, so the result
 you get is not precise even then.

 It got up to 48 C externally in the physics package area.

 That's quite normal.

 Those small Rb's keep the internal temperature constant using
 heaters, which can raise the temperature and by being able to dump
 excess heat through their heat-sink to lower the temperature.

 You shouldn't run your Rb too hot, as this decreases the electronics
 lifetime and reduces the wiggle-room of the thermal management
 inside the device.

 On the other hand, cooling it too much will only increase the
 power drain for the heaters and increase the thermal gradients
 inside the unit, likely degrading thermal stability.

 Poul-Henning


 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.




-- 
Steve Rooke - ZL3TUV  G8KVD  JAKDTTNW
A man with one clock knows what time it is;
A man with two clocks is never quite sure.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FE-5680A heat sink

2009-06-08 Thread Chuck Harris

Hi Leigh,

I just looked at a manual for an EGG Rb standard module, and their spec
is for -55C ambient to +68C baseplate.  They had another spec that said
MTBF 90,000 hours at 40C baseplate.  Your FEI unit probably has a similarly
worded spec.

To me this means that the ideal (expected) temperature for the baseplate is 40C.

If you are intending to operate your Rb block in a home environment, that
has conditioned air, you can assume that the ambient temperature will be
nominally 25C.  I would start from that point, simply attach the
block to the chassis, and measure the operating temperature of the baseplate.
If it is around 40C, great!  If it is much above 40C, you might want to add a
heatsink.

I wouldn't go out of my way to get below 40C.

You can calculate the heatsink size knowing only the ambient temperature,
and the power demand of the Rb block.  But for operation in a civilized
area, it is easier just to try a simple experiment.

-Chuck Harris

Leigh L. Klotz, Jr WA5ZNU wrote:

Chuck,
This device is a pull from a larger system, probably a cell site.  It's 
clearly designed to be mounted on something, as it has lots and lots of 
holes around the edge.  I suspect there's some thermal management that's 
missing.  The FEI sheet gives typical data for the 0-50C range, though 
presumably that's ambient temperature.
I don't want to cool the physics package per se, but I do want to at 
least approximate what kind of thermal solution ought to be supplied.


Quite a few of these have been sold, and I've gotten good advice from 
others on this list about calibration.  I'd hoped that someone would 
have experience with the thermal management.


So far I've seen 
http://www.dd1us.de/Downloads/precise%20reference%20frequency%20rev%200_4.pdf 


which shows (but doesn't describe) a heat sink on the bottom.

Leigh.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FE-5680A heat sink

2009-06-08 Thread Rex

Leigh L. Klotz, Jr WA5ZNU wrote:


Chuck,
This device is a pull from a larger system, probably a cell site.  
It's clearly designed to be mounted on something, as it has lots and 
lots of holes around the edge.  I suspect there's some thermal 
management that's missing.  The FEI sheet gives typical data for the 
0-50C range, though presumably that's ambient temperature.
I don't want to cool the physics package per se, but I do want to at 
least approximate what kind of thermal solution ought to be supplied.



Several years back I bought a 5680A and mine came still mounted on the 
original (Lucent?) circuit board. There is a small amount of circuitry 
on one end, but most of the large board is just a solid plated area 
where the FE-5680A mounts. This is almost 100 in^2 plated on both sides, 
so an equivalent would be a sheet of aluminum about 7.25 x 13.5 inches. 
I'll let someone else translate that to an appropriate finned heatsink.




Quite a few of these have been sold, and I've gotten good advice from 
others on this list about calibration.  I'd hoped that someone would 
have experience with the thermal management.


So far I've seen 
http://www.dd1us.de/Downloads/precise%20reference%20frequency%20rev%200_4.pdf 


which shows (but doesn't describe) a heat sink on the bottom.

Leigh.



I never saw that particular pdf article before. It looks interesting. 
The 5680A I got was a bit different from most and I never found a way to 
program mine. I'll have to revisit the unit and see if the header 
connector he used works. I though I looked at that, but it has been a 
long time. Worth another attempt, I guess.


-Rex, kk6mk


I'm puzzled.  I admit that I don't have a lot of experience with
Rb standards, but I do have a bit of experience with the HP-5065A
Rb standard.

In the 5065A, the entire physics package is enclosed in an oven.

Assuming that your Rb is the same, and I believe it is, your plan
to force the physics package to run at room temperature is just going
to make the oven work harder in its never ending quest to maintain
stable temperature.  If you monitor the current draw of the 5680A,
you will probably see that it goes up when you put a fan on it.

Sometimes, you just have to let electronics run hot.

What did the manufacturer suggest?

-Chuck Harris

Leigh L. Klotz, Jr WA5ZNU wrote:

I've attached an SMA connector to my FE-5680A and built an external 
linear power supply with a TO-3 7815.


I ran the device today for about half an hour, and used an infrared 
sensing thermometer to measure the external case temperature.

It got up to 48 C externally in the physics package area.
I put a muffin fan on top and it brought it down to 38C but I don't 
think this is a good plan because I worry about the effects of the 
fan's magnetic field on the Rubidium system.
I found in my junk box a finned Aluminum heat sink that's exactly 
the same size as the FE-5680A and plan to tap it around the edges 
for 4-40 hardware to attach to the many screw holes.  Even so, this 
heat sink will be on the bottom, so the FE-5680A will have to be 
operated upside down for this to help.


Has anybody got good thermal management solution for this device?  
This is the one currently selling on eBay in the 25x88x125mm chassis.


Thanks,
Leigh.





___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] FE-5680A heat sink

2009-06-08 Thread Mark Sims


Also see my post from June of last year where I had measured the heat sink rise 
over ambient of several rubidium oscillators in free air...  the FEI-5650 was 
very close to it's operating limit if operated in free air and not attached to 
something:
http://www.febo.com/pipermail/time-nuts/2008-June/031695.html

_
Lauren found her dream laptop. Find the PC that’s right for you.
http://www.microsoft.com/windows/choosepc/?ocid=ftp_val_wl_290
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FE-5680A heat sink

2009-06-08 Thread Neville Michie

Hi,
some light might be shed on the heatsink problem by reading the data  
sheet for the LPRO 101 (Symmetricom)

The units will have similar issues to cope with.
The LPRO has temperature controlled lamp and filter, at different  
temperatures above 80C.
Its power consumption is shown in a chart and is reduced as the  
temperature rises because the heating power needed becomes less.
It must have a 17V internal series regulator, because total power  
consumption falls down to the minimum of 18V.
MTBF is listed as 380kh at 20C, 320kh at 30C, 253kh at 40C, 189k at  
50C and 134k at 60C.

These are values that you would expect for most solid state boards.
My solution was to put the device onto a finned heat sink in an  
aluminum box with a tiny 12V brushless fan, (40mm)
which runs on 8 volts and is switched by a thermistor in the heatsink  
to keep the baseplate at 40C.
This seems to be a good trade-off with MTBF, and the 10 MHZ xtal will  
run at a constant temperature of 40C + or - 0.05.
I have not yet assembled the gear to measure how much its performance  
is improved.

cheers, Neville Michie


On 09/06/2009, at 2:18 PM, Mark Sims wrote:




Also see my post from June of last year where I had measured the  
heat sink rise over ambient of several rubidium oscillators in free  
air...  the FEI-5650 was very close to it's operating limit if  
operated in free air and not attached to something:

http://www.febo.com/pipermail/time-nuts/2008-June/031695.html

_
Lauren found her dream laptop. Find the PC that’s right for you.
http://www.microsoft.com/windows/choosepc/?ocid=ftp_val_wl_290
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/ 
time-nuts

and follow the instructions there.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.