Hi,

I believe that I've been down this path before asking a similar question.

The first question relates to the following; recently the Launceston General 
Hospital (LGH) had its NICU redeveloped, the AS 3003 areas were tested by a 
private Biomedical Company on behalf of the Electrical Consultant before 
finally being handed over the Biomedical Team at the LGH checked the area and 
found that many of the Cardiac areas didn't meet the requirements of AS 
3003:2003. Their issue was that in AS 3003 4.3 Disposition of points controlled 
by RCDs "Where an RCD is used as an LPD it shall not control the supply to 
socket-outlets designated for more than one PATIENT. Rooms such as the 
Intensive Care and High dependency were wired as Cardiac Protect had a 
Non-Essential circuit in each room that supplied outlets at  different patient 
locations within that room.

The consultant argued that this was a mistake in the standards and he wasn't 
going to change and that it had been signed off by a Biomedical Engineer now 
that the defects period has finished my Electrical Trade Team is expected to 
correct the mistake however now that the Standard is AS 3003:2011 I will have 
to comply to the 5 metre rule (ridiculous rule if I might say) which now 
becomes somewhat of a major upgrade at greater cost to my budget.

My questions are, was there a mistake in the AS 3003:2003 and we have 
interpreted this incorrectly, does a NICU Unit that is only level 2 require to 
be Cardiac Protected in the latest AS 3003:2011 it only mentions Neo Natal ICU 
(Level 3) need be Cardiac, and are other hospitals wiring their NICU to Cardiac 
Protect.

Any thoughts on these questions would be useful. In the new year the 
development of new Theatres and ICU will commence at the LGH so hopefully they 
will be right.

Regards,

John

John Porter
Team Leader Electrical/Mechanical
Building & Engineering
Launceston General Hospital

Mob: 0419146979
Ph: 03 6348 7433
Fx: 03 6348 7430

Email: [email protected]<mailto:[email protected]>



________________________________

CONFIDENTIALITY NOTICE AND DISCLAIMER
The information in this transmission may be confidential and/or protected by 
legal professional privilege, and is intended only for the person or persons to 
whom it is addressed. If you are not such a person, you are warned that any 
disclosure, copying or dissemination of the information is unauthorised. If you 
have received the transmission in error, please immediately contact this office 
by telephone, fax or email, to inform us of the error and to enable 
arrangements to be made for the destruction of the transmission, or its return 
at our cost. No liability is accepted for any unauthorised use of the 
information contained in this transmission.

Reply via email to