Hi Jon,

I'd like to expand on one of Dan's comments a bit...

1) If the signal source you're driving from is a perfect match and
   there are no other reflections, adding an attenuator won't help
   reduce non-flatness.  You can tolerate up to a maximum of one
   reflection on a line without causing problems other than simple
   mismatch loss.  But even if the source is perfect, the cable itself
   can contribute problems due to reflections at connectors,
   structural resonances, etc.  In these cases the attenuator may
   also fail to provide a total cure, depending on where the various
   problems originate.  But given that the ADC input is a high-
   reflection entity, I agree with Dan and re-stress the importance of
   putting the attenuator as close as possible to the ADC board's
   input, as he described.

   One other thing- you aren't using a TEE connector as a splitter
   anywhere in the line going to the ADC board, are you?  Doing
   so is almost a guarantee of flatness problems.

   The "structural resonance" I mentioned above occurs if there is a
   manufacturing problem with the cable causing spatially-periodic
   variations of characteristic impedance.  I have some cheap RG-58
   cables I once bought which do pretty well up to roughly 1 GHz,
   above which the "loss" rather abruptly becomes much higher in a
   spectacular fashion.

Dana  Whitlow  (Arecibo Observatory)



Dan Werthimer wrote:

hi jon,

there are a couple of possibilities that might explain
the frequency response you are seeing:

1)  the  ADC2x1000-8 board has poor termination.
     (the input impedance is not 50 ohms at all frequencies).
     this causes reflections, which depend on cable length
     and frequency.    you can reduce the reflections by
     placing an attenuator on the ADC input.
     (use an SMA attenuator, and connect it directly
      to the SMA input of the ADC).
     you can also reduce low frequency resonances by using a short cable.

2)   as danny pointed out,
      in your tests injecting a pure sine wave, you need to make sure
      the frequencies you select are always exactly centered in the
      middle of the PFB/FFT spectral bins.  it's best to phase lock the
      signal input synthesizer to the same synth that's driving the
      adc sample clock.   otherwise, you are sensitive to the scalloping
      (there's a 3 dB drop if you are exactly between two spectral bins).


best wishes,

dan


On 6/25/2010 6:29 AM, Jon Losh wrote:
Hi,

I'm working with a group under Professor Tegmark at MIT. We've been messing around with the roach as a spectrometer (tutorial 3) and we're trying to get the unit conversion from the strange power-like units the roach spits out to dBm. I looked at the datasheet for the ADC, and it seems like the ADC accepts inputs from [-1, 1] V and digitizes them. We then retrieve this data off the roach, and it seems like it's scaled, but we can't figure how.

Also, the input voltage at which the ADC clips seems to depend on input frequency, as seen in the attached plot. To make it, for each frequency, we ramped up the power of the input sinusoid until the ADC clipped. I guess there might be some resonances where the input voltage is low?