Thank you. Now that I know this is not the
normal behaviour I can better search for the
cause.

> Hello,
>
> This behavior smacks of a genuine hardware failure, not "drift"
> in the usual analog sense.  I'd be looking for bad solder joints,
> loose or defective cable connectors, that sort of thing.  If you have
> series SMT capacitor(s) in your signal path, that's another
> possibility.  These parts are made of very brittle dielectric
> materials and can develop cracks and other failure modes due
> to board flexing associated with tightening nearby SMA
> connectors, etc.  If you can arrange to monitor the levels in
> near-real-time, try wiggling the signal cable, prodding parts
> on the acquisition board, and that kind of thing.
>
> Or perhaps there's a failure in your signal generator itself.
>
> Ordinary component drift is of the character of one (or a few)
> percent change in value occurring slowly and fairly smoothly
> over a span of several years.
>
> Dana Whitlow
> Arecibo Observatory
>
>
> On 3/22/2013 6:00 PM, [email protected] wrote:
>> If other people see the drift, then I
>> suspect the gain of the input stages
>> of the ADC boards drift. If so, then
>> switches will be needed in some of my
>> spectrometer designs which currently
>> lack them. What do you see or think?
>>
>>> I had reason to return to using the old
>>> spectrometer of CASPER Tutorial 3. This
>>> spectrometer is old enough to be useful
>>> for diagnostics with new designs, but I
>>> noticed a problem when viewing the FFT
>>> spectra. With an input of a 1Vpp sine
>>> wave at 5, 10, 15, 20, 25 or 30 MHz,
>>> the amplitude of the peak in the FFT
>>> spectrum changes in two ways: during a
>>> single run and between runs. I was
>>> trying to determine the source of this
>>> change because I am trying to calibrate
>>> the vertical (amplitude) axis of the
>>> spectrum obtained in the tutorial.
>>>
>>> For example, during a run the amplitude
>>> either rises to almost four times its
>>> original value or drops to a quarter of
>>> its original value. If I wait for a
>>> long time, the amplitude seems to rise
>>> slowly. If I shut down the spectrometer
>>> and all the instruments except the
>>> ROACH board, and then repeat the run at
>>> a later time, the readings may be
>>> different by a factor of five or even
>>> ten. This is making calibration very
>>> difficult. If this is normal behaviour
>>> for the spectrometer of Tutorial 3,
>>> then I may be forced to give up an easy
>>> calibration and use a switch for all
>>> measurements so I can monitor the drift
>>> during a run.
>>>
>>> The apparatus consists of the ROACH 1
>>> board, an old Bee2 ADC board 1GHz
>>> oscillator for the ADC clock, an old
>>> Wavetek oscillator for the input sine
>>> wave and a blocking capacitor at the
>>> input of the ADC board. The oscillator
>>> does not give a perfect sine wave. Is
>>> the drift normal behaviour for the FFT
>>> spectra?
>>>
>>>
>>>
>>
>>
>>
>
>
>