Hi Rurik,

Yes. I used the boffile ver2. I also generated bof using model file and
tried using it. I checked both the ADCs using -z option. I get the same
error.

Thanks and regards,
Sharat
On 4 Sep 2015 18:37, "Primiani, Rurik" <[email protected]> wrote:

> Hi Sharat,
>
> Are you using the revision 1 or revision 2 version of the test suite
> bitcode and does this match the version of the board that you have? By
> default it uses revision 2 which is probably what you have but just to make
> sure. To use the other bitcode you would need to use the -b flag with
> test_adc5g.py.
>
> The adc_test provided bitcode and test script *should* work out of the
> box. Have you tried running the test on ZDOK 1, you can use -z 1 for that.
>
> Best,
> Rurik
>
>
> On Fri, Sep 4, 2015 at 12:01 AM, sharat varma <[email protected]> wrote:
>
>> Hi,
>>
>> Thanks for the reply and sorry for the delayed response.
>> Yes, the x-axis represent the time and y-axis represents the signed 8 bit
>> output. The negative bias is due to the nature of the input.
>>
>> I was trying to use the files in the link you mentioned, but I keep
>> getting the error shown below. I am using the bof file provided for roach2.
>>
>> test roach connectivity ... ok
>> check if requested bof is available ... ok
>> test roach pingability ... ok
>> program the requested bof ... ok
>> estimate clock rate, should be within 1 MHz of expected ... ok
>> confirm the design has the ADC SPI controller ... ok
>> confirm the design has the needed scope ... ok
>> test if calibration finds optimal MMCM phase ... FAIL
>>
>> ======================================================================
>> FAIL: test if calibration finds optimal MMCM phase
>> ----------------------------------------------------------------------
>> Traceback (most recent call last):
>>   File "test_adc5g.py", line 107, in test_optimal_solution_found
>>     self.assertIsNotNone(self._optimal_phase)
>> AssertionError: unexpectedly None
>>
>> ----------------------------------------------------------------------
>> Ran 8 tests in 7.230s
>>
>> FAILED (failures=1)
>>
>> Please let me know if I am doing anything wrong or where could be the
>> problem.
>>
>> For your information:
>>
>> System: roach2
>>
>> ADC :  ASIAA ADC5G ADC
>>
>> Clock : 2500MHz.
>>
>> Thanks and regards,
>>
>> Sharat
>>
>>
>> On 1 September 2015 at 23:01, Primiani, Rurik <[email protected]>
>> wrote:
>>
>>> Hi Sharat,
>>>
>>> The plot you provided has no labels or units so I will assume the x-axis
>>> represents time in samples and the y-axis represents signed 8-bit sample
>>> values. I'm not sure why there is such a negative bias but perhaps that's
>>> particular to your instrument.
>>>
>>> Please, at the very least, run the MMCM calibration described at
>>> https://github.com/sma-wideband/adc_tests to reduce glitches on the
>>> interface. I believe Jack also has a more sophisticated approach which
>>> adjusts the IODELAY for each individual data line; sadly I don't have a
>>> link handy for that.
>>>
>>> Although you may not see these glitches with a sine wave, a noise-like
>>> signal will cause more transitions on each bit and thus more glitches with
>>> an uncalibrated interface.
>>>
>>> Thanks,
>>> Rurik
>>>
>>>
>>>
>>> On Tue, Sep 1, 2015 at 2:54 AM, sharat varma <[email protected]> wrote:
>>>
>>>> Hi Jack,
>>>>
>>>> Thanks for the reply.
>>>>
>>>> I did not run mmcm calibration. Actually, we checked the ADC by feeding
>>>> it a low frequency sine wave from a function generator and it works fine.
>>>>
>>>> The problem with spikes occurs when we feed the ADC with the
>>>> photo-detector output.
>>>>
>>>> Regards,
>>>> Sharat
>>>>
>>>>
>>>> On 1 September 2015 at 13:52, Jack Hickish <[email protected]>
>>>> wrote:
>>>>
>>>>> Hi Sharat,
>>>>>
>>>>> Are you running the adc mmcm calibration routine after programming
>>>>> your roach?
>>>>>
>>>>> Cheers,
>>>>> Jack
>>>>>
>>>>> On 31 August 2015 at 22:41, sharat varma <[email protected]> wrote:
>>>>>
>>>>>>
>>>>>> Hi Casper,
>>>>>>
>>>>>> I am working as a post-doc working under guidance of Dr. Hayden So at
>>>>>> The University of Hong Kong.
>>>>>>
>>>>>> We are using ROACH2 to capture data from optical cytometry. We are
>>>>>> using  ASIAA ADC5G ADC to capture data at 4 to 5 Gsps.
>>>>>>
>>>>>> We basically use the following parameters.
>>>>>>
>>>>>> Block parameter: two-channel, ZDOK0, demux 1:1 .
>>>>>> System: roach2, clock source:adc0_clk, clock rate: 300 MHz.
>>>>>>
>>>>>> We are connecting the output of a photo-detector 1544-B from Newport
>>>>>> Corp (the spec is attached) to the ADC input using SMA.
>>>>>> We find that noisy spikes are introduced when we capture the data
>>>>>> through the ADC (see attached fig). We double checked if the source had
>>>>>> problems using a oscilloscope, but on the oscilloscope we do not see any 
>>>>>> of
>>>>>> these spikes.
>>>>>>
>>>>>> We would be grateful if you could let us know if we are doing
>>>>>> anything wrong.
>>>>>>
>>>>>>  Rgards,
>>>>>> Sharat
>>>>>>
>>>>>>
>>>>>>
>>>>>
>>>>
>>>
>>
>