Given that you are able to achieve a protein concentration of 15 mg per ml
in Tris-HCl, pH 8.0, 100mM NaCl and 1% Glycerol suggests that you have no
drastic solubility problem.

Unless you have a chromatophore as an intrinsic co-factor of one of your
component proteins, the yellow slime may be due to impurities.

The "standard" screens in most kits are fine for "standard" proteins but
not for proteins that are out of the ordinary.

The fact that precipitation is absent or delayed with PEG is coherent with
your ability to concentrate the protein to 15mg/ml.

Now for the improvement steps:
The first step is to "normalize" your protein concentration :
## The mean crystallization condition is around 16% PEG 3350. ##
The above may not be entirely correct but is a good working hypothesis
that can be used as described below:
If 16% PEG 3350, 50mM Tris-HCl, pH 8.0, 100mM NaCl and 1% Glycerol gives
you just some precipitation and 24% PEG in the same salt and buffer conditions precipitates most of the protein, then you can assume that at 15mg/ml your protein concentration is just right and now you may want to screen various ions and different pH to find suitable crystallization conditions. (If not ... reduce your protein concentration to
satisfy the test based on the hypothesis).
The next step is "basic buffer component substitution" to see what your buffer components are doing: That implies checking how other salts can be used instead of NaCl (Li2SO4, KCl Na-mono,di,tri-carboxylates, .... ) 1% glycerol replaced by 0.01-5% organic compounds such as MPD, EtOH, PrOH, xylitol, etc. Tris-HCl replaced by other basic range buffers such as imidazole, glycine, ... pH 8.5-10.6 If the protein tollerates neutral and acidic conditions you can extend the pH range to be searched.

Once you understand your protein complex better you should design a screen of your own based on your
new knowledge.


--
Enrico A. Stura D.Phil. (Oxon) ,    Tel: 33 (0)1 69 08 4302 Office
Room 19, Bat.152,                          Tel: 33 (0)1 69 08 9449  Lab
LTMB, SIMOPRO, IBiTec-S, CEA Saclay,  91191 Gif-sur-Yvette
Cedex FRANCE      http://www-dsv.cea.fr/en/ibitecs/82
    http://www.chem.gla.ac.uk/protein/mirror/stura/index2.html
e-mail: est...@cea.fr                             Fax: 33 (0)1 69 08 90 71

Reply via email to