Weikai,

We did it using NMR but you asked for a simple way so I guess I'm out of topic.

Anyway, since I believe it is the most accurate method, here it is: using a 
high detergent concentration stock solution you can assign resonance peaks to 
your detergent molecule bonds.

Then you can set up a standard curve using different known detergent 
concentrations (for example from 10% down to 0.1%) by calculating the surface 
of your peak(s) which is directly related to your detergent concentration.

Each time you need to know the concentration of a new sample, you just need to 
record the peaks, and use the three-click rule to deduct the unknown value.

As a colleague answered you earlier, we noticed that a 50kDa cutoff withheld a 
lot of detergent during concentration process and consequently your final 
concentration might increase significantly. For example we started with 0.25% 
DES and noticed increases of above 1%. Of course this will depend on the 
concentration factor.

This did not happen when using a 100kDa cutoff, and DES concentration remain 
pretty much constant.

Now, it will depend on your system: what detergent you are using, since micel 
size and CMC are obviously the critical parameters here -- but also what 
maximal cutoff you can use w/o loosing your membrane protein in the flow 
through...

Good luck,
Michael
  ----- Original Message ----- 
  From: Patrick Loll 
  To: CCP4BB@jiscmail.ac.uk 
  Sent: Friday, October 23, 2009 1:12 PM
  Subject: Re: [ccp4bb] measure detergent concentration


  I'll second this.  We've done this as an exercise in NSLS Membrane Protein 
Crystallization workshop for a few years, and it works like a charm. You can 
stain in a warm iodine chamber and visualize by scanning the TLC plate on a 
garden variety scanner (we use an inexpensive Canon LIDE that probably cost 
less than USD 60 five years ago). We quantify the spot intensity with NIH Image 
or equivalent, and get lovely linearity down to the CMC, spotting only 1 uL of 
sample--so we haven't seen any need to concentrate.


  On 23 Oct 2009, at 3:41 PM, Edward A. Berry wrote:


    Only easy if you happen to have silica gel TLC plates and
    a chromatography jar lying around, perhaps from some
    phospholipid analysis:

    A strategy for identification and quantification of
    detergents frequently used in the purification of membrane proteins
    Laura R. Eriks, June A. Mayor, and Ronald S. Kaplan
    Analytical Biochemistry 323 (2003) 234–241

    This paper recommends spotting on a TLC plate and running
    beside standard amounts of the same detergent. From intensity/size
    of the detergent spot after developing you can bracket the detergent
    concentration. (And by the way they found that detergents are concentrated 
by ultrafiltration). To increase sensitivity, speedvac a volume too large to
    spot on the plate, dissolve the residue in Me0H.

    Ed
    wei...@crystal.harvard.edu wrote:

      Hi Folks:

      After concentrating a membrane protein, is there a (easy) way of measuring

      the detergent concentration in the sample?

      Regards,

      Weikai



  
---------------------------------------------------------------------------------------