Matthias Zebisch wrote:
Dear bb!

What is the optimal wavelength for Sulfur SAD phasing?
Is it 1.9A or should one go below that to reduce absorption/damage.

Also, would the same wavelength be appropriate to maximize anomalous
scattering to position chlorides, calcium, sulfate in already phased structures?

Thanks in advance,

Matthias
As usual, the "optimum" is a compromise, and depends on the equipment. I recommend 7 keV (1.77 A) for S-SAD to users at my beamline (8.3.1 at the Advanced Light Source). This seems to be the best compromise between the strength of the signal from Sulfur and the absorption in the air and sample. A good website for calculating the magnitude of absorption corrections (transmission) is here:
http://henke.lbl.gov/optical_constants/

Cameras that contain less air (He paths like the setups Jim Pflugrath mentioned) and perhaps thinner crystals will shift the optimum to longer wavelengths. Beamlines are even being designed to take advantage of backscattered spots, since at the S edge (5 A), your 2.5 A spots will be fired back up the beam pipe (lambda = 2*d*sin(theta)). Using the backscattered geometry is not as crazy as it sounds: the Lorentz factors are higher, and absorption corrections are both minimized and easy to calculate. In fact the "reflection geometry" is the way the Braggs did it (Bragg, James, & Bosanquet, Phil. Mag. 1921a, 1921b, 1922).
http://www.informaworld.com/10.1080/14786442108636225

On the other hand, the current standing record (to my knowledge) for measuring a weak anomalous signal was done at 0.98 A (Wang et. al. Acta D 2006): http://dx.doi.org/10.1107/S0907444906038534

In this case, the average anomalous difference was smaller than 0.5%. Z Dauter was kind enough to make these data available, and I have a copy of them here:
http://bl831.als.lbl.gov/example_data_sets/index.html#APS22ID

The authors do not claim this as an "optimum" S-SAD wavelength, but I suspect that using more penetrating radiation reduced the contribution of errors from the absorption model enough to allow this small signal to be measured. That is, it is hard to estimate the absorption corrections to better than a few percent, but if the absorption itself is small, then errors in absorption corrections are much less significant. For example, +/-5% for a 2% absorption correction is much less significant than +/- 1% for a 50% absorption correction.
The trick, as always, is getting the error to be smaller than the signal.


Basically, do this: You can approximate the noise level in your data with Rmerge or the noise-to-signal ratio (the inverse of Mn(I)/sd). This will be about 3-5% in favorable cases. Next, estimate the magnitude of the anomalous signal you are trying to measure with the Hendrickson-Teeter (Nature 1981) equation, which I paraphrased in a previous posts:
https://www.jiscmail.ac.uk/cgi-bin/webadmin?A2=ind0903&L=CCP4BB&T=0&F=&S=&P=191973
https://www.jiscmail.ac.uk/cgi-bin/webadmin?A2=ind0903&L=CCP4BB&T=0&F=&S=&P=195218

For typical S-SAD cases, the anomalous signal or "Bijvoet ratio" will be ~0.5-1%. Next, divide the noise by the signal and square it. This is the multiplicity (m) you will need:

m = (noise/Bijvoet_ratio)^2

For example, if your I/sd is 30 and you are trying to measure a 1% anomalous difference, you will need a multiplicity of not less than ( (1./30) / 0.01 )^2 = 11. Note, however, that if this I/sd = 30 was from data with a multiplicity of 5, then your "unit" I/sd is actually 30/sqrt(5) = 13.4, and you will need a multiplicity of 55 to measure a 1% anomalous difference. If your anomalous signal is 0.5% in this same situation, then you will need a multiplicity of 223. Yes, these are big multiplicities, and that is why S-SAD is usually demonstrated with large and well-diffracting crystals (less dose per scattered photon). I stress that the above is a rough estimate of the LOWER limit where the world's best crystallographers have managed to get something out of the data. If you are not one of these people, then you will need better data. If you are one of these people, you will insist on it.

-James Holton