Hi,

At such concentrations, it might be necessary to concentrate the medium. 
Otherwise you are likely to be working well below the KD of the affinity beads. 
At 0.5mg/L, 50KDalton gives you 10nM. The KD of Ni-NTA should be around 1uM, 
according to published SPR data  
http://www.sciencedirect.com/science/article/pii/S0003269797923265. This means 
the majority of your protein will stay in solution even at equilibrium. One of 
my colleagues does use Ni-NTA beads for their secreted mammalian protein, which 
produces at >10mg/L. And he still concentrates it at least 10x.

Zhijie




From: Jerry McCully 
Sent: Monday, April 23, 2012 9:37 PM
To: [email protected] 
Subject: [ccp4bb] minimum protein concentration for NI-NTA column


Dear All;

       This is a sort of naive question about the NI-NTA affinity purification.

        
       Is the Ni-NTA column from GE health able to capture proteins at 0.2ug/ml 
to 0.5ug/ml?

       IF not, then it is necessary to concentrate the mammalian expression 
supernatant.

      Thanks a lot and best regards,

Jerry McCully


Reply via email to