Dear Xinghua,

Other have already commented on the fact that if the spots are overlapping, even if you adjust some parameters to be able to integrate the data and get a reasonably complete data set, the resulting intensities will have some systematic errors that may cause problems in downstream structure solution. One should therefore make use of the various strategy programs that are available to try to avoid overlap when the data are collected.

One particularly common problem is the use of a one degree oscillation angle, regardless of the unit cell parameters, mosaic spread and resolution. In this case, even a moderately long cell dimension of (say) 150A that is not approximately along the rotation axis can give rise to a large number of overlaps. There are other very good reasons not to use a one degree oscillation angle (signal to noise) even if there are no overlaps, so it would be to everyone's advantage if this habit could be abandoned (Historically there were good reasons for using one degree, but they are no longer valid). Using 0.25 degrees as a standard would be greatly preferable.

As far as processing the images is concerned, as Harry Powell has pointed out, increasing the "Profile Tolerance" parameters (there are two of them, the first for the central part of the image, the second for the outer regions) is a very good way to start. These default to 0.01, 0.01 for lab data and 0.02,0.03 for synchrotron data. It is worth trying values of up to 0.05-0.08 instead. This will have the effect of "shrinking" the size of the spots as far as the program is concerned, which will automatically reduce the minimum spot separation. It is more effective with lab data (with larger spots) and may not work very well if you have very small spots from a synchrotron source. I prefer this option to changing the SEPARATION parameters, if it works.

The second thing that you can do is to reduce the mosaic spread (and fix it at the input value). This can have a big effect on the number of overlaps, but inevitably it will introduce systematic errors into the intensities and this will be apparent in the FRCBIAS column in SCALA or AIMLESS when you scale and merge the data, giving rise to significant -ve values (ie bigger than 2-3% and negative). I do not have any formula to decide how best to do this, but can only suggest that you try integration with different fixed values and checking the resulting scaling statistics (Mn(I)/sd, FRCBIAS and completeness) and if possible try these different data sets in the downstream structure determination to see which gives the best maps. A further option here is to reduce the "Mosaic block size", which will effectively increase the mosaic spread at low resolution and help offset the fact that you have set the mosaic spread to a value that is lower than it should be, as far as the low resolution data is concerned. You can try values between 1 and 10, you will see immediately from the predicted pattern what the effect is. On scaling the data, you should find that it will help to reduce the FRCBIAS at low resolution. Again, there is no formula for success (well, I don't have one anyway), you will need to try different combinations of mosaic spread and mosaic block size and see how the merging statistics (or maps) look.

Best wishes,

Andrew Leslie




On 14 May 2012, at 03:22, Xinghua Qin wrote:

Dear CCP4ers,

We collected a diffraction dataset with high percentage of spot overlaps, It would be so kind to tell me how to ignore spot overlap in imosflm and explain the hazard of high percentage of spot overlaps.
Thanks in advance.

Best wishes

Xinghua Qin
--
Xinghua Qin
State Key Laboratory of Plant Physiology and biochemistry
College of Biological Sciences
China Agricultural University
No.2, Yuan Ming Yuan West Road
Haidian District, Beijing, China 100193
Tel: +86-10-62732672
E-mail: [email protected]



Reply via email to