On Thu, Jul 12, 2012 at 10:53 PM, Zhijie Li <zhijie...@utoronto.ca> wrote:

> Hi,
>
> Yes, it can be done by yourself. I repacked our 26/60 column a few years
> ago with homemade apparatus. The column has been working for us since then.
> Basically the loading apparatus was a tubing connecting the top of the
> column and bottom of a reservoir (a flask with an opening near the bottom)
> that contained 30%-50% suspended beads. The length of the tubing was about
> 2m, and I left the reservoir on the highest shelf I could find in our lab,
> so that the gravity of the buffer in the column and tubing will drive the
> flow in a constant even manner. It took overnight to get the column nearly
> filled.
> The gel filtration columns need to be tightly packed with no
> discontinuity. As I remember, the GE gel filtration packing manual said:
> push the top filter 2mm further down, after the column has been fully
> compressed by flowing buffer at 1cm/min flow rate.
> The beads need to be gently suspended in the beginning, but constant
> stirring is not necessary. The flow of the buffer will carry beads into to
> column (because my reservoir has an opening at the bottom). The difficult
> part is at the end, when the column is filled to the top. The beads will be
> compressed quite a lot when you start pumping buffer, then you need to fill
> in more beads. As I remember, I removed the upper filter on the inlet
> connector so that I could load beads through the upper inlet tubing with a
> 10mL or 30mL syringe, which generated enough pressure while loading the
> beads - so that I didn't have to take the top off again and again.
>
> There is a manual from GE on how to pack superdex columns. You should be
> able to find it online. The Superdex beads used in different sized columns
> from GE have different diameters. The bigger the column is, the larger the
> beads are. And the larger the beads are, the easier the packing will be.
> 16/60 should still be fairly easy to pack. I checked the theoretical plate
> number of our 26/60 after repacking, it was >11000, not as high as the
> 13000 stated in the 26/60 new column manual, but it was good enough for us.
>
> Zhijie
>
> ------------------------------**--------------------
> From: "Peter Hsu" <hsuu...@u.washington.edu>
> Sent: Thursday, July 12, 2012 9:51 PM
> To: <CCP4BB@JISCMAIL.AC.UK>
> Subject: [ccp4bb] offtopic: packing gel filtration columns
>
>
>  Hi all,
>>
>> Sorry for the slew of offtopic posts, but does anyone here have any
>> experience repacking the large 120mL Superdex75/200 columns? Any
>> advice/tips on doing it? I've got an older column that's gotten clogged
>> while washing w/NaOH (can't go over 0.1mL/min w/o getting overpressure
>> alarm), and not sure the bossman would be thrilled w/buying another column.
>>
>> Thanks for any ideas.
>>
>> Peter
>>
>


-- 
Pius S Padayatti,PhD,
Phone: 216-658-4528

Reply via email to