Dear Chen,

how sure are you that your crystals contain the protein of interest? Repeatedly 
washing harversted crystals in crystal stabilization solution followed by 
SDS-PAGE coupled to appropriate staining protocols (e.g. Coomassie, Silver 
staining) or western blotting can give a pretty conclusive answer.
In our group, SDS-PAGE followed by Silver-staining is done routinely and in 
most cases leads to conclusive results. Here is a summary of the protocol:

--------
- select a drop containing substantial crystalline material. The crystals can 
be many and small (crystal shower) or few and large.
- prepare a PCR-tube with crystal stabilization buffer (e.g. 50 uL of 
motherliquor containing a 10% higher concentration of precipitant).
- transfer all the crystalline material from the drop into the PCR-tube using a 
pipet (You can use stabilization buffer from the PCR tube to collect all 
crystals). One can also use a cryo-loop to harvest the crystals if they are 
large enough to allow efficient harvesting.
- centrifuge the PCR-tube at low speed for 30-60 sec and observe the crystals 
under the microscope to make sure they are at the bottom of the PCR-tube.
- remove as much of the supernatant as you can using a pipet making sure not to 
remove the crystals. Then add crystal stabilization buffer to wash the 
crystals, and centrifuge again.
- repeat this washing procedure a few times (typicaly 3-4 times).
- after the final washing step, centrifugation and removal of the supernatant, 
add Laemmli-buffer to the crystals and use this sample for loading onto the 
SDS-PAGE gel.
- include a positive control (e.g. solubilize another drop directly in 
Laemmli-buffer) and a negative control (final washing buffer). Including a 
pre-crystallization sample of your protein as a control is also recommended, to 
control for the integrity of the protein under crystallization conditions.
- use silver staining to visualize the protein.
-------


best regards
Savvas



On 31 Mar 2014, at 23:48, Chen Zhao <[email protected]> wrote:

> BTW, I forgot to mention that phenix.autosol also gave similar result.
> 
> 
> On Mon, Mar 31, 2014 at 5:46 PM, Chen Zhao <[email protected]> wrote:
> Dear all,
> 
> I am now trying to phase a structure in C2 using anomalous scattering at 5-6 
> A. It is hard to improve the derivative resolution at the moment. Shelxd is 
> able to locate 6 sites with a distinct CC and FOM. After density modification 
> in shelxe, the contrast of the two enantiomers is 0.59/0.38 for SAD and 
> 0.7/0.3 for MAD. When I looked at the electron density, the maps from SAD and 
> MAD are similar, and the solvent boundary is quite clear. However, the 
> problem is that the electron density blob passes through the 2-fold rotation 
> axis, even at 3 rmsd contour level. Also, the unit cell seems to be too small 
> for the molecule. I am afraid that the space group assignment is wrong, but I 
> am a beginner so I nearly have no clue. I did reprocess the data in P1 and 
> looked at the self-rotation function with a radius at 200 A. From the list it 
> seems that there is only one 2-fold rotation axis. I am quite confused. Could 
> anybody give me some hint of this problem?
> 
> Thanks a lot in advance!
> 
> Sincerely,
> Chen
> 