Hi All,

Thanks a lot for the advice from Mark.

1. For the comment to use a attenuated beam, the major concern here is
resolution, since at full beam,  the resolution is already somewhere
between 3.5-4.2, I assume that significant attenuation will bring down the
resolution dramatically, and also decrease the I/Sigma which might  lead to
more errors in |F|?

2.  We used TaBr because as a large electron dense cluster, it is good for
low resolution phasing. This derivative, in the best scenario, give a 4.8A
data set (a resolution that is supposed to be enough to find the huge
cluster) which only has 0.1~0.2A unit cell difference to a native data set
collected on the same beam, theoretically, this dataset should give
prominent peaks on hacker section (the crystal is green, and fluoresence
scan give moderate signal), but what I find is only two small peaks (6-7
sigma, boarder line) (maybe occupancy too low?), and using Hyss to find the
HA only result in a FOM of 0.20. This is what triggers me to wonder if a
high scaling error model (0.05-0.1) is also detrimental for isomorphous
phasing?

3. speaking of Se. Is it possible to use Se as a second derivative for MIR
in my case (500 amino acid, 18 Met)? my concern is that Se might be too
“light” for my protein, and 4.5A resolution is NOT much better than the 7A
Mark has mentioned.

4. Has anybody ever used xenon? Can anyone share the starting pressure and
time for optimization?

Thank you again,

Bei