Amateur Radio Operator N5GE wrote:
> 
> Where do I find the instructions for using this feature? 
> 

Al must have forgotten to include the instructions.  The following is
PRELIMINARY ONLY but should get you going:

Amplitude Calibration

The P3 comes with a default amplitude calibration that works well with a K3
transceiver that either was manufactured after approximately 9/1/2009 or, if
an earlier unit, has the IF output buffer gain modification described in the
application note on the Elecraft web site.  If you have an earlier unit
without the modification, or you are using the P3 with a different model
transceiver, or if you just wish to obtain better amplitude accuracy on the
P3 display, you can do the following calibration.

The recommended test frequency is7040 kHz at a signal level of 50 microvolts
(-73 dBm) using a calibrated signal generator such as the Elecraft XG2. 
However, any frequency and power level within the range of the transceiver
and P3 display will also work.

* Make sure the the P3 IF IN is connected to the transceiver IF output.  If
the transceiver is a K3, connect the RS-232 cable as well.  Turn off the
preamplifier and attenuator on the K3.  If you have the KRX3 installed, make
sure SUB is disabled.

* Connect the signal generator to the transceiver antenna input.  If it is a
K3 that includes the KXV3 or KXV3A option, it is best to use the RX ANT IN
connector to eliminate the possibility of damaging the signal generator if
the K3 transmits accidentally.

* Set the signal generator for 7040 kHz at a signal level of 50 mV (-73
dBm).

* Set the SPAN on the P3 to 100 kHz, and tune the transceiver to
approximately 7015 kHz so that the signal is centered on the right half of
the display.  Set the SCALE to maximum expansion (10 dB) and adjust the REF
LVL to about -78 dBm, in order to place the peak of the signal about half
way up the screen.  You may notice a slight variation (~1.5 dB) in the peak
signal level with small adjustments in the transceiver frequency, as the
signal is tuned from pixel to pixel across the display.  Adjust the
frequency for a maximum reading.

* Tap MENU, select Lvl Cal, tap the knob, and adjust the level calibration
until the signal is at -73 dBm on the display.

You can also calibrate a receiving or transceiving converter, either the
K3's optional internal K144XV or an external device.  If you plan to do the
above calibration on a standard frequency, do it first since that
calibration is added to the converter calibration.

* Set up a signal generator, the transceiver and the P3 in a manner similar
to the above to obtain a signal near the center of the right side of the
display and approximately centered vertically, with the P3's SCAN at 100 kHz
and the SCALE set to 10 dB.

* Tap MENU, select XV Gain, tap the knob, and adjust the knob to select the
desired converter band, 1-9.  Tap the knob again and adjust the gain on the
selected band so that the signal is at the the correct level on the display.



-- 
View this message in context: 
http://elecraft.365791.n2.nabble.com/P3-New-LVL-Cal-Function-Beta-0-41-tp5727222p5728318.html
Sent from the Elecraft mailing list archive at Nabble.com.
______________________________________________________________
Elecraft mailing list
Home: http://mailman.qth.net/mailman/listinfo/elecraft
Help: http://mailman.qth.net/mmfaq.htm
Post: mailto:Elecraft@mailman.qth.net

This list hosted by: http://www.qsl.net
Please help support this email list: http://www.qsl.net/donate.html

Reply via email to