Richard,
     
     Thanks for the reply.
     
     Sorry, I forgot to mention I'm using EN61010-1 "Safety requirements 
     for electrical equipment for measurement, control, and laboratory use" 
     as my low voltage directive test standard.  (We make cable testers).
     
     It says "symbols must be in compliance with IEC-27." but then there is 
     a table that says the caution symbol is "black on yellow."
     
     No other equipment that I can see (scopes, meters, spectrum analyzers) 
     use black on yellow.  They all use whatever label colors they have 
     available.
     
     
     Thanks,
     
     Jon Bertrand
     [email protected]
     


______________________________ Reply Separator _________________________________
Subject: RE: Symbols and IEC 27
Author:  "WOODS, RICHARD" <[email protected]> at Address-InternetPO
List-Post: [email protected]
Date:    12/19/96 3:01 PM


You may be starting in the wrong place. Look at the standard under which 
you are certifying the product. Does it, or the other standards that it 
references, specifiy the color of warnings? For example, EN60950 does 
not, but  the component specification for lasers, EN 60825-1, does call 
out for specific colors for some warnings. As far as EN60950 is 
concerned, I am unaware of any other need for colored warnings.
     
Richard Woods
Sensormatic Electronics
     
These are my ideas and not necessarily those of my company.