     We manufacture a piece of laboratory equipment that has a control unit 
     and a furnace for burning samples and analyzing the gases to determine 
     the compounds in the sample. The control unit and furnace units are in 
     separate enclosures.
     
     In IEC 1010 part 2, for equipment for heating of materials, Sec 6.8.2 
     allows that equipment for which a drying period is specified, shall 
     not be subjected to humidity preconditioning.
     
     It appears that we just specify a drying out period in the manual if 
     the equipment has been exposed to humid conditions. This makes sense 
     to me because the equipment will not be operated in such a high humid 
     condition. 
     
     My question is, can this same type of approach be taken with both the 
     furnace and control unit? If so, than can it be applied to other lab 
     equipment?
     
     My real problem is trying to justify a $30k - $50k thermal/humidity 
     test chamber, when I do not see that this type of test will improve 
     the safety of our type of equipment.
     
     Thank you for your assistance.
     