Moshe, You might check out Dale Safety Analyzers. They have a variety of
models. Contact Dale Technology Corp.  PO Box 196, Thornwood, NY 10594 
phone 800-544-3253.

****Added note:
We've had problems with DNA Nevada. Their older products had inadequate
bandwidth, and we found several testers with identical model numbers had
different load circuits, with no way to tell which was which.
****

An additional factor to consider is the "body model" circuit used in the
leakage tester. Different load circuits are specified by IEC 601-1 and
AAMI ES-1. The old UL 544 was different from both. All give similar
results for current sources at frequencies below 100 kHz. The UL load
gave significantly different response for voltage sources. (Most
real-life leakage sources can be considered high-impedance current
sources.)
 ----------
>I'm looking for a leakage tester which can measure the low currents
>of the medical standards (e.g. 601)

Reply via email to