Moshe,
I have used a couple different Educated Design Leakage current
testers, and have been very satisfied. Although I have not used the
model for IEC 601, they have one in their brochure. The contact is:
Rita Fischer
ED&D
521E Uwharrie Court
Raleigh NC 27606 USA
ph 919 821-7088
Fx 919 821-1939
Email [email protected]
www.productsafet.com/prodsafety
Good Luck,
Ray Russell
[email protected]
______________________________ Reply Separator _________________________________
Subject: RE: looking for a leakage tester for IEC 601-1
Author: "Farnsworth,Heber" <[email protected]> at Internet
List-Post: [email protected]
Date: 8/13/97 6:37 PM
Moshe, You might check out Dale Safety Analyzers. They have a variety of
models. Contact Dale Technology Corp. PO Box 196, Thornwood, NY 10594
phone 800-544-3253.
An additional factor to consider is the "body model" circuit used in the
leakage tester. Different load circuits are specified by IEC 601-1 and
AAMI ES-1. The old UL 544 was different from both. All give similar
results for current sources at frequencies below 100 kHz. The UL load
gave significantly different response for voltage sources. (Most
real-life leakage sources can be considered high-impedance current
sources.)
----------
>I'm looking for a leakage tester which can measure the low currents
>of the medical standards (e.g. 601)