Patrick Lawler wrote:
> 
> My copy of IEC601-1-2 (Medical EMC compatibility, First edition) says
> there are no requirements for harmonic distortion (paragraph 36.201.2.1).
> Is this still correct today?  Are there plans to incorporate this
> requirement in the future?

Dear Patrick,
The copy you have is still right. In Europe this standard has been 
harmonized  as EN60601-1-2 (1993) and is used for conformity with 
89/336/EEC EMC Directive requirements.
On July 1996 a Committe Draft was circulated, referenced as 62A/206/CD, 
in order to produce a second edition of the standard.
In this CD one can read:
"36.201.3.1 Harmonic distortion
a) Requirement:
Equipment and/or systems with a rated input current up to and including 
16A per phase and intended to be connected to the public mains network 
shall comply with the applicable limits contained in IEC 1000-3-2.
b) Test Methid:
The test methods and test equipment described in IEC 1000-3-2 applay."

This CD is still under study: it should not be used for reference 
purposes. Just to make you feel the situation.

Ciao
M.P.
-- 
================================================================