Patrick,

In my previous answer, please ignore item 2. I misread IEC 1000-3-2 for IEC 
1000-4-2.

Jon



>Patrick,
>
>1) EN 60601-1-2 just covers EMC, while the scope the Medical Devices Directive 
>includes safety, risk analysis, medical efficacy, biocompatibily, software 
>validation, etc.
>
>Medical products are classified into Classes, 1, 11a, 11b and 111. Self 
>declaration can be used for most, but not all, Class 1 products (the least 
>critical). Even this this case, self declaration is based on a technical file 
>that includes much more than EN 60601-1-2.
>
>In all other cases, approval is via a Notified Body under the Medical Devices 
>Directive. Approval covers the manufacturer as well as the product, as some 
>level of ISO 9000 approval is required. The ISO 9000 audit is carried out by 
>the Notified Body itself. There is a complex relationship between the Class of 
>product and the level of ISO 9000 approval required.
>
>2) EN 60601-1-2 does have ESD requirements, but to IEC 801-2, which is older 
>than IEC 1000-3-2.
>
>3) EMC requirements for medical equipment are covered by the Medical Devices 
>Directive, not the EMC Directive. So if harmonics are not mentioned in EN 
>60601-1-2, then it is not required.
>
>
>Jon Griver
>ITL (Product Testing) Ltd.
>http://www.itl.co.il
>
>
>
>>
>>
>>I was asked the following question, and wasn't sure about the answer:
>>
>>1) A manufacturer can claim his system complies with the Medical Device
>>Directive by testing to EN 60601-1-2:1993.
>>
>>2) EN 60601-1-2:1993 does not have any requirements to test to IEC 1000-3-2.
>>
>>3) Does that mean that the harmonic standard does not apply to this system 
>>when
>>it becomes mandatory in 2001?
>>
>>
>>I realize the Second Edition of IEC60601-1-2 is just around the corner, and
>>that edition _does_ call out the harmonic test.  However, I've been told that
>>it may not take effect until 2003.
>>--
>>Patrick Lawler
>>[email protected]
>>
>>---------
>>This message is coming from the emc-pstc discussion list.
>>To cancel your subscription, send mail to [email protected]
>>with the single line: "unsubscribe emc-pstc" (without the
>>quotes).  For help, send mail to [email protected],
>>[email protected], [email protected], or
>>[email protected] (the list administrators).
>>
>>
>>


---------
This message is coming from the emc-pstc discussion list.
To cancel your subscription, send mail to [email protected]
with the single line: "unsubscribe emc-pstc" (without the
quotes).  For help, send mail to [email protected],
[email protected], [email protected], or
[email protected] (the list administrators).

Reply via email to