Hi, In general, the answer is no. You need to look at IEC 60601-1-1 for Medical Electrical Systems, Annex BBB. In particular the Table in BBB.7 and Figure BBB.201.
Depending on the solution you chose, you will have to either add an additional protective ground, extra isolating transformer, floating power supply, or separation. You will also have to keep the total enclosure leakage current to less than 500 micro amps. Ned Devine Entela, Inc. Program Manager III Phone 616 248 9671 Fax 616 574 9752 e-mail [email protected] -----Original Message----- From: Grasso, Charles (Chaz) [mailto:[email protected]] Sent: Thursday, June 17, 1999 9:58 PM To: 'EMC Group' Subject: ITE Equipment installed in hospitals Does anyone know if ITE equipment is required to meet any Medical requirements if installed in a hospital. (No patient connections!!) Thank you Charles Grasso StorageTek 2270 Sth 88th Street Louisville CO 80027 Tel: (303)673-2908 Fax(303)661-7115 --------- This message is coming from the emc-pstc discussion list. To cancel your subscription, send mail to [email protected] with the single line: "unsubscribe emc-pstc" (without the quotes). For help, send mail to [email protected], [email protected], [email protected], or [email protected] (the list administrators). --------- This message is coming from the emc-pstc discussion list. To cancel your subscription, send mail to [email protected] with the single line: "unsubscribe emc-pstc" (without the quotes). For help, send mail to [email protected], [email protected], [email protected], or [email protected] (the list administrators).