Posted for Jon Griver:
 
 
 
 
 
 

------------------------
  From: Jon Griver <[email protected]>
  Subject: ITE Equipment installed in hospitals
  Date: Fri, 18 Jun 1999 14:30:42 +0300 
  To: [email protected]


> 
> Charles,
> 
> There should not be any problem using ITE equipment in hospitals, when the 
> equipment is used by staff and nowhere near patients, and not connected to 
> any other medical equipment.
> 
> If ITE equipment is connected to medical equipment, then its use is covered 
> by the standard IEC 601-1-1 (no, there are not too many 1's), for medical 
> SYSTEMS. Basically, if the equipment is located outside of the 'patient 
> vicinity' (see the standard for definition), then it only needs to comply 
> with IEC 950. If it is located in the 'patient vicinity', then it must comply 
> with IEC 601-1, for medical EQUIPMENT. In both cases, check IEC 601-1-1 to 
> see how the equipment is connected, particularly with regard to grounding.
> 
> Regards,
> 
> Jon Griver
> ITL (Product Testing) Ltd.
> http://www.itl.co.il
> 
> 
> >
> >Does anyone know if ITE equipment is required to
> >meet any Medical requirements if installed in
> >a hospital. (No patient connections!!)
> >Thank you
> >Charles Grasso
> >StorageTek
> >2270 Sth 88th Street
> >Louisville CO 80027
> >Tel: (303)673-2908
> >Fax(303)661-7115
> >
> >
> >---------
> >This message is coming from the emc-pstc discussion list.
> >To cancel your subscription, send mail to [email protected]
> >with the single line: "unsubscribe emc-pstc" (without the
> >quotes).  For help, send mail to [email protected],
> >[email protected], [email protected], or
> >[email protected] (the list administrators).
> >
> >
> >
> 
> 

---------------End of Original Message-----------------

--------------------------
Ed Price
[email protected]
Electromagnetic Compatibility Lab
Cubic Defense Systems
San Diego, CA.  USA
619-505-2780
List-Post: [email protected]
Date: 06/18/1999
Time: 09:12:51
Military & Avionics EMC Services Our Specialty
Also Environmental / Metrology / Reliability
--------------------------



---------
This message is coming from the emc-pstc discussion list.
To cancel your subscription, send mail to [email protected]
with the single line: "unsubscribe emc-pstc" (without the
quotes).  For help, send mail to [email protected],
[email protected], [email protected], or
[email protected] (the list administrators).

Reply via email to