Dear Martin,

Regarding the ROUTINE TESTS, since I assume that you are intending to test
some LABORATORY EQUIPMENT, the Annex K of EN 61010 Standard is answering to
your question: " to verify bonding of ACCESSIBLE conductive parts to the
PROTECTIVE CONDUCTOR TERMINAL a test current is at the discretion of the
manufacturer, which means that you may use any instrument (including an
ohmmeter) for this purpose, which will apply a test current and will offer
you the chance to display (calculate) the value of the impedance between the
PROTECTIVE CONDUCTOR TERMINAL and the ACCESSIBLE part for  which PROTECTIVE
BONDING that should not exceed 0.1 ohm. 

In the above conditions, since the current is at the discretion of the
manufacturer, (depending of the used instrument) the DURATION of the test
shall respect the instructions provided by the manufacturer of the used
instrument in order to obtain a "meaningful reading" (references should be
made to EN 501116 Routine tests for ITE Equipment).   

Respectfully yours,
Constantin

Constantin Bolintineanu P.Eng.
MDS SCIEX 
71 Four Valley Drive 
CONCORD, Ontario
CANADA, L4K 4V8

Tel: (905) 660 9006 x 361
E-mail: [email protected]


> ----------
> From:
> [email protected][SMTP:[email protected]]
> Reply To:     [email protected]
> Sent:         2000/09/11 5:55 PM
> To:   [email protected]
> Subject:      Production Line Continuity Testing
> 
> 
> 
> 
> What should be the duration for the production line continuity test?  How
> did
> you arrive at that duration?  Please include what standard you are using
> and any
> specific references.
> 
> Thanks for the help.
> 
> Regards
> 
> Joe Martin
> [email protected]

-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Jim Bacher:              [email protected]
     Michael Garretson:        [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]

Reply via email to