I also agree that the headphones are a patient connection.

Hi,

Its been a while since I looked at UL 544, but a few comments.

1.      Headphones are an applied part (UL2601-1) or patient connection
(UL544).  They come into direct contact with the patient.  You can argue
that they are an ordinary patient connection (Clause 2.18 of UL544).  But,
it is still a patient connection.  The limit is 50 uA for ordinary patient
connection.

2.      How long is this product going to be in the market?  UL 544 goes
away on 1/1/2003 for new products and 1/1/2005 for all products.  You might
be better off going to UL2601-1 now.  In UL2601-1 this type of patient
connect is Type BF.  The patient leakage current limit is 100 uA in normal
condition and 500 uA in single fault condition.

3.      Call UL.  Maybe I missed some "out" in UL544.



Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax  616 574 9752
e-mail  ndev...@entela.com



-----Original Message-----
From: me...@aol.com [mailto:me...@aol.com]
Sent: Friday, March 24, 2000 3:36 PM
To: jjuh...@fiberoptions.com; m.r...@ieee.org;
emc-p...@majordomo.ieee.org
Subject: UL544 Leakage Limits Patient Equipment*



Most esteemed colleagues, we are looking for your professional opinion on a
UL 544 leakage limit (I think you will find this interesting):

1. This particular product uses a UL 544 evaluated direct plug in power
supply with outputs to the patient care equipment. In this case it is a
diagnostic unit that sends an audible tone to headphones (audiometer). The
plug in power supply Conditions of Acceptability indicate the outputs are
"not evaluated for patient leads" (i.e. applied parts").

2. Table 42.1 of UL 544 specifies leakage limits. "patient connection"
footnote "a" references testing of "patient leads (applied parts)"
connections. There is no written definition for "patient leads" or "applied
parts" in UL544.

As such NFPA 99 supplements UL 544 as it draws from the NEC and NFPA 99
(referenced in UL 544):

NFPA 99 defines the US definition of  "Patient Lead" = "A deliberate
electrical connection that can carry current between an appliance and
patient. It is not intended to include adventitious or casual contacts such
as a push button, bed surface, lamp, hand held appliance, etc."

3. As the headphones of this audiometer are clearly not "deliberate
electrical connections" we conclude these are not "patient leads (applied
parts)" which would not fall under the limits for "patient connection"
limits
per 544. The applicable limits would be as defined under "enclosure or
chassis" "grounded " or" double insulated".... Now be careful not to jump to

a conclusion yet..... You might say "enclosure or chassis?", but if you
examine this, you will find the footnotes reference UL 544's Enclosure
definition:

"Enclosure =  That external portion of an appliance that serves to house or
support component parts, or both. Enclosure of patient care equipment likely

to be contacted by a patient include, for example, bedside monitors, bed
frames, dental chairs, and examination stands."

Our conclusion: Due to the US definition of "patient leads (applied parts)",

the earphones of an audiometer (patient care equipment) are subjected to the

leakage current limits for "enclosure or chassis", and not the limits of
"patient connection". For this particular application, we conclude that
based
on the C of As, the output of the power supply has already been evaluated
for
"enclosure or chassis" leakage limits.

Your Thoughts???

Drew

PS: If you care to look, CSA supports this position in that 50uA is related
to "cardiac tissue" limits only. See Appendix A of CSA 22.2 125 (500uA).

-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
    majord...@ieee.org
with the single line:
    unsubscribe emc-pstc

For help, send mail to the list administrators:
    Jim Bacher:              jim_bac...@mail.monarch.com
    Michael Garretson:        pstc_ad...@garretson.org

For policy questions, send mail to:
    Richard Nute:           ri...@ieee.org


-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
    majord...@ieee.org
with the single line:
    unsubscribe emc-pstc

For help, send mail to the list administrators:
    Jim Bacher:              jim_bac...@mail.monarch.com
    Michael Garretson:        pstc_ad...@garretson.org

For policy questions, send mail to:
    Richard Nute:           ri...@ieee.org

______________________________________________________
Get Your Private, Free Email at http://www.hotmail.com


-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
    majord...@ieee.org
with the single line:
    unsubscribe emc-pstc

For help, send mail to the list administrators:
    Jim Bacher:              jim_bac...@mail.monarch.com
    Michael Garretson:        pstc_ad...@garretson.org

For policy questions, send mail to:
    Richard Nute:           ri...@ieee.org

Reply via email to