It may not be important whether you call it an applied part or not. You are not
making a direct electrical connection to the patient. In any case you should
meet the leakage requirement. This should be easy with properly insulated
earphones.
IEC 60950 also addresses this issue and throws in more considerations. For the
condition of telephone type headsets they are worried about transients from
telecom or power lines, and therefore have additional isolation requirements.
Clause 6.2.2 requires an impulse test of 2.5 KV or hipot of 1.5 KV (in Australia
this is increased to both 7KV impulse and 3KV hipot) applied to foil wrapped
around the headset.
Bob Johnson

Kenneth McCormick wrote:

> I also agree that the headphones are a patient connection.
>
> Hi,
>
> Its been a while since I looked at UL 544, but a few comments.
>
> 1.      Headphones are an applied part (UL2601-1) or patient connection
> (UL544).  They come into direct contact with the patient.  You can argue
> that they are an ordinary patient connection (Clause 2.18 of UL544).  But,
> it is still a patient connection.  The limit is 50 uA for ordinary patient
> connection.
>
> 2.      How long is this product going to be in the market?  UL 544 goes
> away on 1/1/2003 for new products and 1/1/2005 for all products.  You might
> be better off going to UL2601-1 now.  In UL2601-1 this type of patient
> connect is Type BF.  The patient leakage current limit is 100 uA in normal
> condition and 500 uA in single fault condition.
>
> 3.      Call UL.  Maybe I missed some "out" in UL544.
>
> Ned Devine
> Entela, Inc.
> Program Manager III
> Phone 616 248 9671
> Fax  616 574 9752
> e-mail  ndev...@entela.com
>
> -----Original Message-----
> From: me...@aol.com [mailto:me...@aol.com]
> Sent: Friday, March 24, 2000 3:36 PM
> To: jjuh...@fiberoptions.com; m.r...@ieee.org;
> emc-p...@majordomo.ieee.org
> Subject: UL544 Leakage Limits Patient Equipment*
>
> Most esteemed colleagues, we are looking for your professional opinion on a
> UL 544 leakage limit (I think you will find this interesting):
>
> 1. This particular product uses a UL 544 evaluated direct plug in power
> supply with outputs to the patient care equipment. In this case it is a
> diagnostic unit that sends an audible tone to headphones (audiometer). The
> plug in power supply Conditions of Acceptability indicate the outputs are
> "not evaluated for patient leads" (i.e. applied parts").
>
> 2. Table 42.1 of UL 544 specifies leakage limits. "patient connection"
> footnote "a" references testing of "patient leads (applied parts)"
> connections. There is no written definition for "patient leads" or "applied
> parts" in UL544.
>
> As such NFPA 99 supplements UL 544 as it draws from the NEC and NFPA 99
> (referenced in UL 544):
>
> NFPA 99 defines the US definition of  "Patient Lead" = "A deliberate
> electrical connection that can carry current between an appliance and
> patient. It is not intended to include adventitious or casual contacts such
> as a push button, bed surface, lamp, hand held appliance, etc."
>
> 3. As the headphones of this audiometer are clearly not "deliberate
> electrical connections" we conclude these are not "patient leads (applied
> parts)" which would not fall under the limits for "patient connection"
> limits
> per 544. The applicable limits would be as defined under "enclosure or
> chassis" "grounded " or" double insulated".... Now be careful not to jump to
>
> a conclusion yet..... You might say "enclosure or chassis?", but if you
> examine this, you will find the footnotes reference UL 544's Enclosure
> definition:
>
> "Enclosure =  That external portion of an appliance that serves to house or
> support component parts, or both. Enclosure of patient care equipment likely
>
> to be contacted by a patient include, for example, bedside monitors, bed
> frames, dental chairs, and examination stands."
>
> Our conclusion: Due to the US definition of "patient leads (applied parts)",
>
> the earphones of an audiometer (patient care equipment) are subjected to the
>
> leakage current limits for "enclosure or chassis", and not the limits of
> "patient connection". For this particular application, we conclude that
> based
> on the C of As, the output of the power supply has already been evaluated
> for
> "enclosure or chassis" leakage limits.
>
> Your Thoughts???
>
> Drew
>
> PS: If you care to look, CSA supports this position in that 50uA is related
> to "cardiac tissue" limits only. See Appendix A of CSA 22.2 125 (500uA).
>
> -------------------------------------------
> This message is from the IEEE EMC Society Product Safety
> Technical Committee emc-pstc discussion list.
>
> To cancel your subscription, send mail to:
>      majord...@ieee.org
> with the single line:
>      unsubscribe emc-pstc
>
> For help, send mail to the list administrators:
>      Jim Bacher:              jim_bac...@mail.monarch.com
>      Michael Garretson:        pstc_ad...@garretson.org
>
> For policy questions, send mail to:
>      Richard Nute:           ri...@ieee.org
>
> -------------------------------------------
> This message is from the IEEE EMC Society Product Safety
> Technical Committee emc-pstc discussion list.
>
> To cancel your subscription, send mail to:
>      majord...@ieee.org
> with the single line:
>      unsubscribe emc-pstc
>
> For help, send mail to the list administrators:
>      Jim Bacher:              jim_bac...@mail.monarch.com
>      Michael Garretson:        pstc_ad...@garretson.org
>
> For policy questions, send mail to:
>      Richard Nute:           ri...@ieee.org
>
> ______________________________________________________
> Get Your Private, Free Email at http://www.hotmail.com
>
> -------------------------------------------
> This message is from the IEEE EMC Society Product Safety
> Technical Committee emc-pstc discussion list.
>
> To cancel your subscription, send mail to:
>      majord...@ieee.org
> with the single line:
>      unsubscribe emc-pstc
>
> For help, send mail to the list administrators:
>      Jim Bacher:              jim_bac...@mail.monarch.com
>      Michael Garretson:        pstc_ad...@garretson.org
>
> For policy questions, send mail to:
>      Richard Nute:           ri...@ieee.org


-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to: