EN60601-1 (Medical Safety) Appendix C provides a list entitled "Sequence of
Testing". This list is used during the course of the safety investigation on
the device. It follows as:
C24: Cont. leakage currents and patient auxiliary currents at operating temp
C25: Dielectric strength at operating temp
C26: Humidity preconditioning treatment
C27: Dielectric strength test (cold condition after humidity conditioning)
C28: Leakage current test after humidity preconditioning treatment

In our production testing we perform a high potential dielectric strength
test first followed by a leakage current test. Our high potential tester has
a ground continuity circuit and will warn the operator if the safety ground
circuit is not in place (open).

-----Original Message-----
From: [email protected] [mailto:[email protected]]
Sent: Friday, March 17, 2000 7:37 AM
To: Bruce Touzel; Art Michael
Cc: [email protected]; Martin Rowe (TMW)
Subject: Re: SAF: Insulation resistance and hipot tests



Folks:

EN 60204-1 also specifies an insulation resistance test.
Depending on the standard, they tell you the sequence of the various safety
tests.
For example, in EN 60204-1, clause 19.1 (the test clause) states:
"When these test are performed, it is recommended that they follow the
sequence listed."

For EN 60204-1, it is Ground continuity, then Insulation resistance, then
Hipot, then cap discharge.

I hope this helps.

At 01:21 PM 3/16/00 -0800, Bruce Touzel wrote:
>
>Insulation resistance tests would apply only to IEC 950 I believe, so this
may
>not apply for you ?
>
>Art Michael wrote:
>
>> Hello Martin,
>>
>> I don't have the answer to your question, but do caution that the First
>> Test to be run (if the DUT is grounded) is the Ground Continuity Test
(for
>> test-operator protection).
>>
>> Regards, Art Michael
>>
>>                  * * * * * * * * * * * * * * * * * * * * * *
>>                 *   International Product Safety Bookshop   *
>>                 *      Check out our current offerings!     *
>>                 * <http://www.safetylink.com/bookshop.html> *
>>                 *                                           *
>>                 *     Now offering BSI's Books & Reports    *
>>                 *  including, "World Electricity Supplies"  *
>>                 *                                           *
>>                 *     Another service of the Safety Link    *
>>                 *          <www.safetylink.com>             *
>>                  * * * * * * * * * * * * * * * * * * * * * *
>> ----------------------------------------------------------------------
>>
>> On Thu, 16 Mar 2000, Martin Rowe (TMW) wrote:
>>
>> >
>> > Group,
>> >
>> > A reader asked about the order of performing safety tests.
>> > Should he perform insulation resistance tests before or after
>> > hipot tests? The reader didn't say what type of product he's
>> > testing, but I can ask.
>> >
>> > Thanks,
>> >
>> > ----------------------------    /\
>> > | Martin Rowe              |   /  \
>> > | Senior Technical Editor  |  /    \          /\
>> > | Test & Measurement World | /      \        /  \    /\  ____
>> > | voice 617-558-4426       |/        \      /    \  /  \/
>> > | fax 617-928-4426         |          \    /      \/
>> > | e-mail [email protected]   |           \  /
>> > | http://www.tmworld.com   |            \/
>> > ----------------------------
>> >
>> >
>> > -------------------------------------------
>> > This message is from the IEEE EMC Society Product Safety
>> > Technical Committee emc-pstc discussion list.
>> >
>> > To cancel your subscription, send mail to:
>> >      [email protected]
>> > with the single line:
>> >      unsubscribe emc-pstc
>> >
>> > For help, send mail to the list administrators:
>> >      Jim Bacher:              [email protected]
>> >      Michael Garretson:        [email protected]
>> >
>> > For policy questions, send mail to:
>> >      Richard Nute:           [email protected]
>> >
>> >
>> >
>>
>> -------------------------------------------
>> This message is from the IEEE EMC Society Product Safety
>> Technical Committee emc-pstc discussion list.
>>
>> To cancel your subscription, send mail to:
>>      [email protected]
>> with the single line:
>>      unsubscribe emc-pstc
>>
>> For help, send mail to the list administrators:
>>      Jim Bacher:              [email protected]
>>      Michael Garretson:        [email protected]
>>
>> For policy questions, send mail to:
>>      Richard Nute:           [email protected]
>
>
>-------------------------------------------
>This message is from the IEEE EMC Society Product Safety
>Technical Committee emc-pstc discussion list.
>
>To cancel your subscription, send mail to:
>     [email protected]
>with the single line:
>     unsubscribe emc-pstc
>
>For help, send mail to the list administrators:
>     Jim Bacher:              [email protected]
>     Michael Garretson:        [email protected]
>
>For policy questions, send mail to:
>     Richard Nute:           [email protected]
> 
Mark A. D'Agostino
Sr. Product Safety Engineer
978 282 7520
978 281 3162 (fax)

-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Jim Bacher:              [email protected]
     Michael Garretson:        [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]


-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Jim Bacher:              [email protected]
     Michael Garretson:        [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]

Reply via email to